6.3.3.5
THERMOCOUPLE INPUTS AND CONFIGURATION JUMPER (JP5)
In its base configuration, the T400 analyzer does not include any thermocouple sensors;
however, in instruments where the optional metal wool scrubber (OPT-68) is installed,
one thermocouple is used to sense the temperature of the scrubber. By default, this single
thermocouple input is plugged into the TC1 input (J15) on the relay PCA. TC2 (J16) is
currently not used.
Table 6-2. Thermocouple Configuration Jumper (JP5) Pin-Outs
JUMPER
TC INPUT
PAIR
1 – 11
2 – 12
TC1
3 – 13
4 – 14
5 – 15
TC2
083730400A DCN8130
DESCRIPTION
Selects preamp gain factor for J or K TC
Gain Selector
OUT = K TC gain factor
Selects preamp gain factor for J or K TC
Output Scale Selector
OUT = 10 mV / °C;
When present, sets Cold Junction Compensation for J
Type J Compensation
type Thermocouple
When present, sets Cold Junction Compensation for K
Type K Compensation
type Thermocouple
Selects between Isolated and grounded TC
Termination Selector
IN = Isolate TC; OUT = Grounded TC
NOT USED
Teledyne API – Model T400 O
FUNCTION
; IN = J TC gain factor
IN = 5 mV / °C
Analyzer
3
155