Table 4-1. AUTO CAL States

MODE NAME
Enabled
Calibrate
Zero
Low
High
083730400A DCN8130

Figure 4-3. Auto Cal Page

enables the sequence;
disables the sequence.
enables an actual calibration when the Enabled box is also
allows a calibration check when the Enabled box is also
causes the sequence to perform a Zero calibration when both the
Calibrate and Enabled boxes are also
causes a Zero check when the Enabled box is also
Calibrate box is unchecked (
disables Zero calibration and check
causes the sequence to perform a Low Span calibration when both
the Calibrate and Enabled boxes are also
causes a Low Span check when the Enabled box is also
the Calibrate box is unchecked (
disables Low Span calibration and check
causes the sequence to perform a High Span concentration
calibration when both the Calibrate and Enabled boxes are also
causes a High Span check when the Enabled box is also
the Calibrate box is unchecked (
disables the High Span calibration and check.