Figure 2-49. Alternative Setup Using 250Ω Resistor for Checking Current Output Signal Levels
In this case, follow the procedure above but adjust the output for the following values:
Table 2-13. Current Loop Output Check
% FS
0
100
2.5.9 S
>I
ETUP
NSTRUMENT
As presented in Table 2-14, view product and system information and network settings,
edit network settings, and perform certain maintenance tasks.
Table 2-14. Setup>Instrument Menu
Product Info
System Info
Network Settings
Date/Time Settings
NTP Time Settings
Language
Remote Update
083730400A DCN8130
V OUT +
V OUT -
ANALYZER
Voltage across Resistor for 2-20 mA
500 mVDC
5000 mVDC
MENU
View Model, Part, and Serial Numbers and Package and Driver Versions,
and options information.
View Windows and RAM information.
View the network settings (configurable through the
Setup>Comm>Network Settings menu).
Adjust date, hour, and minutes, select a time zone*, and set the system
clock to automatically adjust for Daylight Savings Time or not. (Also see
Setup>Vars>Daylight Savings Enable). *Time Zone change requires a
special procedure; see Maintenance Section 5.5.
Configure Network Time Protocol settings for clock synchronization.
Select an available language.
When an instrument is connected to a network that is connected to the
Internet, follow the instructions on this Remote Update page to check for
and activate software/firmware updates. (Also refer to Section 5.3).
Teledyne API – Model T400 O
V
+ D C
G n d
Volt
Meter
V IN +
250 Ω
V IN -
Recording
Device
Voltage across Resistor for 4-20 mA
DESCRIPTION
Analyzer
3
1000 mVDC
5000 mVDC
75