2.5.8 S
>A
ETUP
NALOG
Map the four user-configurable Analog Outputs to any of a wide variety of "Signals"
present in the instrument and customize their respective configurations.
Refer to Figure 2-44 for the following:
•
Signal Out: select a Signal for the output.
Min/Max: edit minimum and maximum values associated with the selected Signal.
•
Calibration Type:
•
AUTO for group calibration (Figure 2-45) of the analog outputs (cannot be
•
selected when Current is selected for the Range)
MANUAL for individual calibration (Figure 2-46) of analog outputs where
•
manual adjustments can be made (the only calibration type allowed when
Current is selected for the Range). See Sections 2.5.8.1 and 2.5.8.2 .
Range: assign a voltage or select Current (refer to Table 2-11).
•
Recorder Offset: add a zero offset for recording slightly negative readings from
•
noise around the zero point.
Allow Overrange: check to allow a ± 5% over-range; uncheck to disable over-range
•
if the recording device is sensitive to excess voltage or current.
083730400A DCN8130
O
UTPUTS
Figure 2-44. Analog Output Configuration Example
Teledyne API – Model T400 O
Analyzer
3
71