Power Supply/ Circuit Breaker; Figure 6-12. Power Distribution Block Diagram - Teledyne T400 User Manual

Photometric ozone analyzer with numaview software
Table of Contents

Advertisement

6.3.4 P
OWER
The analyzer operates on 100 VAC, 115 VAC or 230 VAC power at either 50 Hz or
60Hz. Individual instruments are set up at the factory to accept any combination of these
five attributes. Power enters the analyzer through a standard IEC 320 power receptacle
located on the rear panel of the instrument. From there it is routed through the ON/OFF
Switch located on the front panel. A 6.75 Amp circuit breaker is built into the ON/OFF
Switch.
AC Line power is stepped down and converted to DC power by two DC Power Supplies.
One supplies +12 VDC, for various valves and valve options, while a second supply
provides +5 VDC and ±15 VDC for logic and analog circuitry as well as the power
supplies for the Photometer and IZS UV Lamps.
All AC and DC voltages are distributed via the Relay PCA.
SENSOR SUITES
ANALOG
SENSORS
(e.g. UV sensors,
Temp Sensors,
Flow Sensors,
UV Detectors,
etc.)
Photometer UV
Lamp DC Heater
DC Heater for the
optional IZS O
Generator UV Lamp
AC Heater for the
optional
Metal Wool Scrubber
PUMP
Photometer UV
Lamp P/S
Optional IZS O
3
Generator UV
Lamp P/S
156
S
/ C
B
UPPLY
IRCUIT
REAKER
Sensor Control
& I/O Logic
LOGIC DEVICES
Pre-Amplifiers
(e.g. CPU and its
& Amplifiers
peripheral devices,
2
I
C bus, MotherBoard,
RELAY PCA
3
Configuration
Jumpers
Configuration
Jumpers
Solenoid
Drivers
OPTION
MEASURE /
VALVES
REFERENCE
(IZS, Sample/Cal,
VALVES
Zero/Span)

Figure 6-12. Power Distribution Block Diagram

Teledyne API – Model T400 O
etc.)
PS 1
+5 VDC
±15 VDC
PS 2
(+12 VDC)
COOLING
FAN(S)
Analyzer
3
KEY
AC POWER
DC POWER
ON / OFF
SWITCH
AC
POWER IN
083730400A DCN8130