Adjusting Or Replacing The Izs Ozone Generator Option Suv Lamp; Figure 5-9. O 3 Generator Uv Lamp - Teledyne T400 User Manual

Photometric ozone analyzer with numaview software
Table of Contents

Advertisement

5.6.10
A
UV L
This procedure details the steps for replacing and calibrating the UV lamp of the O
Generator that is included in the Internal Zero Span (IZS) option. If adjusting an existing
lamp, skip to Step 8.
1. Power off the analyzer.
2. Remove the cover from the analyzer.
3. Locate the O
4. Unscrew the UV lamp holder and gently pull out the old lamp.
5. Inspect the o-ring beneath the lamp holder and replace if damaged.
6. Install the new lamp in O
7. Power on the analyzer and allow it to stabilize for at least 20 minutes.
112
R
DJUSTING OR
EPLACING THE
AMP
CAUTION – Mercury Exposure Risk
The UV lamp contains mercury (Hg), which is considered hazardous
waste. The lamp should be disposed of in accordance with local
regulations regarding waste containing mercury.
generator (see Figure 2-3).
3
Figure 5-9. O
3
holder.
The lamp should be free enough to be rotated in the assembly by grasping the
lamp cable.
Teledyne API – Model T400 O
IZS O
ZONE
Generator UV Lamp
generator housing, and partially tighten the UV lamp
3
Analyzer
3
G
O
ENERATOR
PTION
3
083730400A DCN8130
'