Teledyne T400 User Manual page 30

Photometric ozone analyzer with numaview software
Table of Contents

Advertisement

C
ONVERTING
To convert an output configured for current loop operation to the standard 0 to 5 VDC
output operation:
1. Turn off power to the analyzer.
2. If a recording device was connected to the output being modified, disconnect it.
3. Remove the top cover.
4. Remove the screw holding the current loop option to the motherboard.
5. Disconnect the current loop option PCA from the appropriate connector on the
6. Each connector, J19 and J23, requires two shunts. Place one shunt on the two left-
7. Reattach the top case to the analyzer.
8. Calibrate the analog output in the Setup>Analog Outputs>Analog Output Cal menu.
28
C
L
A
URRENT
OOP
NALOG
Remove the screws fastening the top cover to the unit (both sides).
Slide the cover back and lift straight up.
motherboard (see Figure 2-5).
most pins and the second shunt on the two pins next to it (see Figure 2-5).
The analyzer is now ready to have a voltage-sensing, recording device attached
to that output.
Teledyne API – Model T400 O
O
S
UTPUTS TO
TANDARD
Analyzer
3
V
O
OLTAGE
UTPUTS
083730400A DCN8130