Figure 3-28: Zero/Span Calibration Procedure - Teledyne T100 Instruction Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for T100:
Table of Contents

Advertisement

Teledyne ML - T100 UV Fluorescence SO2 Analyzer
3.7.1.4. ZERO/SPAN CALIBRATION
The SPAN key now appears
during the transition from
zero to span.
You may see both buttons.
If either the ZERO or SPAN
buttons fail to appear see the
Troubleshooting section for
tips.
To perform the zero/span calibration procedure, press:
SAMPLE
< TST
TST >
Toggle TST> button until ...
SAMPLE
STABIL= XXXX PPB
< TST TST >
Allow zero gas to enter the sample port
SAMPLE
STABIL= XXXX PPB
< TST TST >
M-P CAL
STABIL= XXXX PPB
<TST TST>
M-P CAL
STABIL= XXXX PPB
<TST TST>
Allow span gas to enter the sample port
at the rear of the analyzer.
SAMPLE
STABIL= XXXX PPB
< TST TST >
M-P CAL
STABIL= XXXX PPB
<TST TST> ZERO
M-P CAL
STABIL= XXXX PPB
<TST TST>
ENTR
M-P CAL
STABIL= XXXX PPB
<TST TST> ENTR
Figure 3-28:
Zero/Span Calibration Procedure
RANGE=500.0 PPB
SO2= XXXX
CAL
SETUP
SO2=XXX.X
CAL
SETUP
at the rear of the analyzer.
SO2=XXX.X
CAL
SETUP
SO2=XXX.X
ZERO
CONC
EXIT
SO2=XXX.X
ENTR
CONC
SO2=XXX.X
CAL
SETUP
SO2=XXX.X
SPAN
CONC
EXIT
SO2=XXX.X
CONC
EXIT
SO2=XXX.X
CONC
EXIT
Set the Display to show
the STABIL test function.
This function calculates
the stability of the SO
measurement.
Wait until STABIL falls
below 0.5 ppb.
This may take several
minutes.
Press ENTR to changes
the OFFSET & SLOPE
values for the SO
measurements.
Press EXIT to leave the
calibration unchanged and
return to the previous
menu.
EXIT
Wait until STABIL falls
below 0.5 ppb.
This may take several
minutes.
Press ENTR to changes
the OFFSET & SLOPE
values for the SO
measurements.
Press EXIT to leave the
calibration unchanged and
return to the previous
menu.
EXIT at this point
returns to the
SAMPLE menu.
Getting Started
2
2
2
75