6.5. MANUAL CALIBRATION WITH IZS OPTION

Zero air and span gas must be supplied to the analyzer through the sample gas inlet as
depicted in Figure 6-1 of Section 6.2.
The calibration procedure must be initiated using the CAL button, not the CALZ or CALS
buttons, using the procedure defined in Section 6.2.
Using the CAL button does not activate the zero/span or sample/cal valves of the IZS
option, thus allowing the introduction of zero air and sample gas through the sample port
from more accurate, external sources such as a calibrated bottle of SO
Dilution Calibrator.
6.6. MANUAL CALIBRATION CHECKS WITH IZS OR ZERO/SPAN
VALVES
On units with an IZS option installed, zero air and span gas are supplied to the analyzer
through the zero gas inlet and from ambient air.
On units with a zero/span valve option installed, zero air and span gas are supplied to the
analyzer through the zero gas and span gas inlets from two different sources.
The zero and calibration operations are initiated directly and independently with dedicated
buttons CALZ and CALS.
Under the best conditions, the accuracy off the SO
permeation tube is about ±5%. This can be subject to significant amounts of drift as the
tube ages and the amount of SO
sufficient for informal calibration checks, it is not adequate for formal calibrations and is
not approved for use by the US EPA as a calibration source.
Therefore, for formal calibrations of an instrument with an IZS option installed the
following provisions must be followed.
SAMPLE
< TST TST >
Use for formal
calibration
operations.
Figure 6-8: Manual Calibration with IZS Option
Zero and span checks using the zero/span valve or IZS option are similar to that
described in Section 6.3, with the following exceptions:
contained in the tube is depleted. Whereas this may be
2
RANGE = 500.000 PPB
CAL
CALZ
CALS
Calibration Procedures
effusion rate of the IZS option's
2
or a Model T700
2
SO2 =XXX.X
SETUP
Use only for
informal calibration
checks.