Section II Operating Instructions
ATTENTION
Figure 3-16:
3.3.2.5. SAMPLE GAS SOURCE
48
Teledyne API T802 Paramagnetic O
COULD DAMAGE INSTRUMENT AND VOID WARRANTY
Remove dust plugs from rear panel exhaust and supply line fittings
before powering on/operating instrument. These plugs should be
kept for reuse in the event of future storage or shipping to prevent
debris from entering the pneumatics.
The exhaust from the analyzer's internal or customer supplied external
pump MUST be vented outside the immediate area or shelter surrounding
the instrument.
Pneumatic Connections–Basic Configuration–Using Bottled Span Gas
Attach a sample inlet line to the SAMPLE inlet port. The SAMPLE input line
should not be more than two (2) meters long.
Maximum pressure of any gas at the SAMPLE inlet should not exceed 1.5 in-
Hg above ambient pressure and ideally should equal ambient atmospheric
pressure.
In applications where the sample gas is received from a pressurized
manifold, a vent must be placed on the sample gas before it enters the
analyzer. Please refer to Figure 3-16.
CAUTION
G
S
H
ENERAL
AFETY
AZARD
Analyzer Operation Manual
2
07275B DCN6418