In general, flow problems can be divided into three categories:
Flow is too high
Flow is greater than zero, but is too low, and/or unstable
Flow is zero (no flow)
When troubleshooting flow problems, it is crucial to confirm the actual flow rate
without relying on the analyzer's flow display. The use of an independent,
external flow meter to perform a flow check as described in Section 10.3.4 is
essential.
The flow diagrams found in a variety of locations within this manual depicting
the T802 in its standard configuration and with options installed, can help in
trouble-shooting flow problems. For your convenience the diagrams are collected
here.