Appendix C
Warranty/Repair Questionnaire
T80X, M80XE
(06532C DCN 5798)
CUSTOMER: _______________________________
CONTACT NAME: ___________________________
SITE ADDRESS: ___________________________________________________________________________
MODEL TYPE: ______________ SERIAL NO.: ________________ FIRMWARE REVISION: ____________
Are there any failure messages? _______________________________________________________________
________________________________________________________________________________________________________________________________
________________________________________________________________________
PLEASE COMPLETE THE FOLLOWING TABLE:
PARAMETER
O2 RANGE
O2 CELL TEMP
O2 SLOPE
O2 OFFSET
CO2 RANGE
CO2 CELL TEMP
CO2 SLOPE
CO2 OFFSET
SAMPLE FLOW
REF_4096_MV
1
Not all models are equipped with both an O
Cap the SAMPLE inlet and record the flow rate and pressure readings:
What is PRESS____________________in-Hg-A
What is the SAMPLE FLOW__________ cc/min
What are the failure symptoms? __________________________________________________________
____________________________________________________________________________________
____________________________________________________________________________________
____________________________________________________________________________________
____________________________________________________________________________________
____________________________________________________________________________________
What test have you done trying to solve the problem? ________________________________________
____________________________________________________________________________________
____________________________________________________________________________________
____________________________________________________________________________________
07275B DCN6418
1
1
1
1
1
1
1
1
STABIL
PRESS
BOX TEMP
following values are under the signal i/o submenu
REF_GND
2
TELEDYNE API CUSTOMER SERVICE
EMAIL:
PHONE: (858) 657-9800
PHONE: _____________________________________
FAX NO. _____________________________________
RECORDED VALUE
and a CO
sensor.
2
api-customerservice@teledyne.com
TOLL FREE: (800) 324-5190
(Continue on back if necessary)
ACCEPTABLE VALUE
%
0-100%
C
50 ± 5
º
1.0 ± 0.3
-
10 to 10%
%
0 to 20%
C
50 ± 5
º
1.0 ± 0.3
-
10 to 10%
 0.2% with zero air
%
in-Hg-A
ambient ± 1
3
120 ± 20
cm
/min
C
ambient ± 5ºC
º
mV
4096mV ±2 mV and Must be Stable
mV
0± 0.5 and Must be Stable
FAX: (858) 657-9816
C-1