Ndir Measurement Of Co - Teledyne T802 Operation Manual

Paramagnetic oxygen analyzer
Table of Contents

Advertisement

Teledyne API T802 Paramagnetic O

12.2. NDIR MEASUREMENT OF CO

07275B DCN6418
Analyzer Operation Manual
2
This current is passed to a feedback loop, which generates a second current to a
wire winding (in effect, a small DC electric motor) mounted on the suspended
mirror. The more O
2
moves and the more current is needed to move the dumbbell back to it's zero
position. Finally, sensor measures the amount of current generated by the
feedback control loop which is directly proportional to the concentration of
oxygen within the sample gas mixture.
The optional CO
sensor is a silicon based Non-Dispersive Infrared (NDIR)
2
sensor. It uses a single-beam, dual wavelength measurement method.
An infrared source at one end of the measurement chamber emits IR radiation
into the sensor's measurement chamber where light at the 4.3 μm wavelength is
partially absorbed by any CO
Interferometer (FPI) is electronically tuned so that only light at the absorption
wavelength of CO
is allowed to pass and be detected by the sensor's IR detector.
2
A reference measurement is made by electronically shifting the filter band pass
wavelength so that no IR at the CO
Figure 12-3:
CO
2
The sensor computes the ratio between the reference signal and the measurement
signal to determine the degree of light absorbed by CO
chamber. This dual wavelength method of measuring CO
to compensate for ancillary effects like sensor aging and contamination.
present, the more the dumbbell and its attached mirror
2
present. A special light filter called a Fabry-Perot
2
absorption wavelength is let through.
2
Sensor Theory of Operation
Principles of Operation
present in the sensor
2
allows the instrument
2
221