Teledyne API T802 Paramagnetic O
12.5.6.2. SENSOR INPUTS
12.5.6.3. SAMPLE PRESSURE AND FLOW
12.5.6.4. THERMISTOR INTERFACE
12.5.6.5. ANALOG OUTPUTS
07275B DCN6418
Analyzer Operation Manual
2
functional block selects a particular signal input (e.g. BOX TEMP) and then
converts the selected voltage into a digital word.
The A/D consists of a voltage-to-frequency (V-F) converter, a programmable
logic device (PLD), three multiplexers, several amplifiers and some other
associated devices. The V-F converter produces a frequency proportional to its
input voltage. The PLD counts the output of the V-F during a specified time
period, and sends the result of that count, in the form of a binary number, to the
CPU.
The A/D can be configured for several different input modes and ranges but in
the T802 the A/D is used in uni-polar mode with a +5 V full scale. The converter
includes a 1% over and under-range. This allows signals from –0.05 V to +5.05
V to be fully converted.
For calibration purposes, two reference voltages are supplied to the A/D
converter: Reference Ground and +4.096 VDC. During calibration, the device
measures these two voltages, outputs their digital equivalent to the CPU. The
CPU uses these values to compute the converter's offset and slope and uses these
factors for subsequent conversions.
See Section 5.9.3.2 for instructions on performing this calibration.
The key analog sensor signals are coupled to the A/D through the master
multiplexer from two connectors on the motherboard. 100K terminating resistors
on each of the inputs prevent cross talk from appearing on the sensor signals.
These are analog signals from two sensors that measure the pressure and flow rate
of the gas stream at the outlet of the sample chamber. This information is used in
two ways.
First, the sample pressure is used by the CPU to calculate O
Concentration.
Second, the pressure and flow rate are monitored as a test
function to assist the user in predicting and troubleshooting failures.
This circuit provides excitation, termination and signal selection for several
negative-coefficient, thermistor temperature sensors located inside the analyzer;
there is a thermistor for the O
current temperature of the chamber housing to the CPU as part of the bench
heater control loop. Another thermistor, attached to the motherboard, measures
the analyzer's inside temperature (box temperature). This information is stored
by the CPU and can be viewed by the user for troubleshooting purposes via the
front panel display (See Section 11.1.2).
The analyzer comes equipped with four analog outputs: A1, A2, A3 and A4.
The type of data and electronic performance of these outputs are configurable by
the user. (see Section 5.9.2). All four analog outputs are connected back to the
A/D converter through a Loop-back circuit. This permits the voltage outputs to
be calibrated by the CPU without need for any additional tools or fixtures.
sample chamber housing, which reports the
2
Principles of Operation
2
233