9.4.2. FLOW CALIBRATION

Toggle to match the actual
flow as measured by the
external flow meter
(The ACAL submenu in the Primary Setup Menu is a special configuration; consult factory).
178
Teledyne API T802 Paramagnetic O
The flow calibration allows the user to adjust the values of the sample flow rates
as they are displayed on the front panel and reported through COM ports to match
the actual flow rate measured at the sample inlet. This does not change the
hardware measurement of the flow sensors, only the software-calculated values.
To carry out this adjustment, connect an external, sufficiently accurate flow meter
to the sample inlet. Once the flow meter is attached and is measuring actual gas
flow, press:
SAMPLE
RANGE=100.00 %
<TST
TST>
CAL
SETUP X.X
PRIMARY SETUP MENU
CFG
DAS ACAL RANG PASS CLK
SETUP X.X
SECONDARY SETUP MENU
COMM VARS
DIAG
SETUP X.X
ENTER PASSWORD:818
8
1
8
DIAG
PREV NEXT
Continue pressing NEXT until ...
DIAG
FLOW CALIBRATION
PREV NEXT
DIAG FCAL
ACTUAL FLOW: 120 CC/M
0
1
2
07275A (DCN6005)
O2=XXX.XX
MORE
ENTR
SIGNAL I/O
ENTR
ENTR
0
ENTR
Analyzer Operation Manual
2
SETUP
EXIT
EXIT
EXIT
EXIT
EXIT
EXIT
EXIT discards the new
setting
ENTR accepts the
new setting