Flow Problems; Calibration Problems; Technical Assistance - Teledyne T640 User Manual

Pm mass monitor
Hide thumbs Also See for T640:
Table of Contents

Advertisement

Message
Box Temp Warning
AMB Press Slope OOR Ambient Pressure Calibration Slope is
5.3.2.

FLOW PROBLEMS

5.3.3.

CALIBRATION PROBLEMS

5.3.4.

TECHNICAL ASSISTANCE

08354A DCN7303
Description
Box Temperature Warning (>60)
out of range
If a flow auditing device indicates any problems with flow, check to ensure the
following:
all connections are seated tightly and evenly (no gaps)
the inlet is not clogged or blocked
the pump is running and is within its PWM range (35% - 80%).
After making any adjustments, run a flow calibration and recheck the flow rate. If
problems persist, please contact Technical Support.
Call our Technical Support Department for technical assistance.
If this manual and its troubleshooting / service sections do not solve your
problems, technical assistance may be obtained from:
Teledyne API Technical Support,
9970 Carroll Canyon Road, Ste A
San Diego, California 92131-1106 USA
Toll-free Phone:
Phone:
Fax:
Email:
Website:
Before contacting Teledyne API Technical Support, fill out the problem report
form in the Appendix, which is also available online for electronic submission at
http://www.teledyne-api.com
Teledyne API T640 PM Mass Monitor
Ensure proper climate and ventilation
for instrument.
Check calibration (make sure units
match calibration device).
Replace pressure sensor.
Call Tech Support.
800-324-5190
+1 858-657-9800
+1 858-657-9816
sda_techsupport@teledyne.com
http://www.teledyne-api.com/
under Technical Support Forms.