Box Temp Warning Box Temperature Warning (>60)
AMB Press Slope
Ambient Pressure Calibration Slope is out of
OOR
range

12.2 Flow Problems

If a flow auditing device indicates any problems with flow, check to ensure the following:
all connections are seated tightly and evenly (no gaps);
the inlet is not clogged or blocked;
the pump is running and is within its PWM range (35% - 80%);
the DFU appear reasonably clean and changed within last 12 months.
After making any adjustments, run a flow calibration and recheck the flow rate. If problems
persist, contact Teledyne-API Technical Support for assistance.
and ventilation for
instrument.
Ensure proper climate and
ventilation for instrument.
Check calibration (make sure
units match calibration
device).
Replace pressure sensor. Call
Tech Support.