11.5 Moisture analyser settings storing

The moisture analyser enables to save 10 different drying settings. Saved settings are
kept in the memory even after unplugging moisture analyzer from the mains.

11.5.1 Saving settings

To save drying settings follow the instructions below:
ms-ma/ms*100%
T = 32.23C
m = 0.000g
t = 0:00:00s
m0= 0.000g
START
SET RAP
SETTINGS
1. Mode
:
2. Calculation
:
3. Drying temp.
:
4. Samples quantity
:
5. Sampling interval
:
6. Drying time
:
7. Drying profiles
8. Settings storing
:
Exit
SETTINGS
1. Mode
:
2. Calculation
:
3. Drying temp.
:
4. Samples quantity
:
5. Sampling interval
:
6. Drying time
:
7. Drying profiles
8. Settings storing
:
Exit
SETTINGS
Save settings?
NO
YES
USER MANUAL
T=120C
ts= 0:10:00s
MEM
STOP
Time mode
m0-m/m0*100%
60C
OFF
10s
0:00:10s
1
<
>
Time mode
m0-m/m0*100%
60C
OFF
10s
0:00:10s
< 2 >
<
>
Press F2 key to choose SET option.
Set necessary drying settings (see
chapter 11.4).
Select Settings storing option using ∧
and ∨ and press ENTER to accept
ENTER
Choose a reference number for
specific setting using < and > keys
and press ENTER to accept. Than
chose Exit and press ENTER.
ENTER
To save settings choose YES and
press ENTER to accept.
ENTER
19
.