11. Moisture analyser functioning description

11.1 Switching on

After switching-on the moisture analyser proceeds with self-tests.
INITIALIZATION
model
CPU
EEPROM
...
...
After completing self-tests, the analyser is tared and the dryer begins initial heating
necessary to create thermal conditions suitable for measurements
m0-m/m0*100%
T= 32.23C
m=0.000g
INITIAL HEATING
Initial heating should warm the drying chamber up to 105°C within 1 minute.
When temperature during initial heating exceeds 105°C or heating time is longer than 1
minute, terminate initial heating with CLR key and check if the analyser is not damaged
(see chapter 15).
After initial heating is completed (or terminated), the device displays the following
information:
m0-m/m0*100%
T = 32.23C
m= 0.000g
t =0:00:00s
m0= 0.000g
START
SET
Legend:
m0-m/m0*100%
-
formula used to calculate the moisture
Ts
defined drying temperature
ts
defined drying time
T –
current temperature in the drying chamber
m –
current weight,
t –
current drying time
m0 –
initial weight
-
indicator of drying profile