Axis AGS200/T250 User Manual

Ags series moisture analyser

Advertisement

User Manual
Moisture Analyser
AGS series
Measuring method description
Ver. 1 2014-08-28

Advertisement

Table of Contents
loading

Summary of Contents for Axis AGS200/T250

  • Page 1 User Manual Moisture Analyser AGS series Measuring method description Ver. 1 2014-08-28...
  • Page 2: Table Of Contents

    USER MANUAL Contents: 1. General description 2. Completeness 3. Security rules 4. Environment protection 5. Technical data 6. Keys and indicators 7. Preparing moisture analyser to work 8. Interfaces 9. General working rules 10. Description of thermogravimetric analysis 10.1 Infrared radiation source 10.2 Infrared radiation drying description 10.3...
  • Page 3: General Description

    USER MANUAL 1. General description AGS series moisture analyser is destined for fast and precise moisture determination. The moisture analyser is based on two cooperating devices: the balance, used to measure current sample weight, and the dryer, which dries the sample using halogen heaters.
  • Page 4: Security Rules

    USER MANUAL 3. Security rules To avoid electrical shock or damage of the moisture analyzer or connected peripheral devices, it is necessary to follow the security rules below.  To feed the analyser use only mains socket with ground contact. ...
  • Page 5: Environment Protection

    Please return wasted device to the point of purchase or other company specialised in recycling of wasted electronic components. 5. Technical data Technical data Moisture analyzers Model AGS200/T250 Range (Max) 200g Readout unit (d) Work temperature +18 + +33°C Humidity readout precision 0,01% for sample 2g …...
  • Page 6: Keys And Indicators

    USER MANUAL 6. Keys and indicators T = 24,35 C m = 4,321g = 0:00:00s m0= 0.000g START STOP  T  - tare (subtract package weight from weighed mass) - switch- on / switch-off (standby), - confirmation / select the option, ENTER - cancel operation - decimal point,...
  • Page 7: Preparing Moisture Analyser To Work

    USER MANUAL 7. Preparing moisture analyser to work During heating, the halogen heaters 1 warm up to very high temperature. When drying chamber is opened avoid touching the heaters as it may cause severe burns or damage the heaters! Dryer chamber cover 3 heats up to 40°C, but perforated cover may heat up over 60°C.
  • Page 8 USER MANUAL After self-tests and result stabilisation zero indication is displayed. The dryer starts initial heating (signalised with an appropriate communicate). After initial heating the moisture analyser is ready to work. When temperature during initial heating exceeds 105°C or heating time is longer than 1 minute, terminate initial heating with CLR key and check if the temperature sensor 2 works properly and if both halogen heaters light 1 (see chapter 15).
  • Page 9: Interfaces

    USER MANUAL 8. Interfaces The moisture analyser is equipped with RS232C interface to connect a printer or a computer and with PS2 port to connect an external computer keyboard. RS232C Clock RxD (receiver) TxD (transmitter) Data case 9. General working rules During transportation remove the pan, the pan support and the pan shield and place it in a separate package.
  • Page 10: Description Of Thermogravimetric Analysis

    Methods of determining moisture may be grouped in two main categories: absolute and deductive. Absolute methods are based on simple relations, e.g. weight decline during drying. Thermogravimetric analysis used in AXIS moisture analyser is an example of this method. Deductive (indirect) methods measure physical quantity related with moisture, e.g.
  • Page 11: Drawing And Preparation Of A Sample

    USER MANUAL 10.3 Drawing and preparation of a sample As sample of given substance must be representative, drawing and preparing a sample is very important process as it affects repeatability of measurements. The most common method of homogenizing a sample is mixing. The other method is to draw few samples from different but specific points in a substance and calculate an average value.
  • Page 12: Glass Fibre Filter

    USER MANUAL Attention: Due to temperature sensor localisation, max sample height is 10mm. When substance ply is too thick, surface layers will be heated too much and internal – not enough. This may result in burning the sample or surface incrustation, which will make drying process difficult and measuring result false.
  • Page 13: Other Practical Notes

    USER MANUAL 10.9 Other practical notes It is preferable to work with the same mass of the sample at each measurement to measure the size of the sample in a reproducible way. It is best to use the same instruments for the application of the sample. Put a sample on the pan as quickly as possible to avoid losing moisture.
  • Page 14: Moisture Analyser Functioning Description

    USER MANUAL 11. Moisture analyser functioning description 11.1 Switching on After switching-on the moisture analyser proceeds with self-tests. INITIALIZATION model program version EEPROM After completing self-tests, the analyser is tared and the dryer begins initial heating necessary to create thermal conditions suitable for measurements m0-m/m0*100% Ts=120C ts= 0:10:00s...
  • Page 15: Drier Operation Modes

    USER MANUAL 11.2 Drier operation modes During the balance – drier operation sampling of the mass on the pan takes place. Sampling time is set by the user, according to drying process speed. As a result of sampling the current humidity value is calculated and displayed. Measurement is finished depending on selected Drying mode: 1.
  • Page 16: Calculation Methods

    USER MANUAL 11.3 Calculation methods Humidity may be calculated upon the basis of various mathematic formulas, defined in balance – drier as Calculation method: 1. Relative humidity, defined in relation to initial mass w [%] = m -m/m *100% , –...
  • Page 17: Drier Operation Parameters Setting

    USER MANUAL 11.4 Drier operation parameters setting m0-m/m0*100% Ts=120C ts= 0:10:00s Select Settings option by pressing F2 T = 32.23C key. m = 0.000g t = 0:00:00s 0.00 and ∨ m0= 0.000g successive menu positions selected, e.g. Mode. START STOP SETTINGS To go to options selection press 1.
  • Page 18: Drying Profiles

    USER MANUAL 11.4.1 Drying profiles DRYING PROFILES Drying profile will used 1. Drying temp. 120C optimization of drying process by 2. Mode Short mode 3. Calculation m0-m/m0*100% accommodation process 4. Samples quantity 2 samples physical properties sample 5. Sampling interval material 6.
  • Page 19: Moisture Analyser Settings Storing

    USER MANUAL 11.5 Moisture analyser settings storing The moisture analyser enables to save 10 different drying settings. Saved settings are kept in the memory even after unplugging moisture analyzer from the mains. 11.5.1 Saving settings To save drying settings follow the instructions below: ms-ma/ms*100% T=120C ts= 0:10:00s...
  • Page 20: Loading Saved Settings

    USER MANUAL SETTINGS analyser displays short communicate Saving..After parameters are saved the analyser is Saving . . . ready to work with new drying parameters. 11.5.2 Loading saved settings ms-ma/ms*100% Ts=120C ts= 0:10:00s To load saved settings, stored in the memory, choose MEM option using T = 32.23C F4 key.
  • Page 21: Initial Moisture Analysis

    USER MANUAL 11.6 Initial moisture analysis To determine optimal drying parameters for unknown sample, it is recommended to perform initial measurement with activated drying chart displaying. To do this, set the following drying parameters (see Drying parameters setting): - Operation Mode: Time mode - Calculation method: m0-m/m0*100% - Drying temperature: organic substances: 80 - 120...
  • Page 22 Observing drying process chart it is possible to evaluate its course and define time required for complete drying. The chart shows 160 time samples on the X axis (for longer times chart is scaled to 360 samples, 720, etc.) and humidity value according to selected formula on the Y axis (chart is automatically scaled to 10%, 30%, 50%, etc.).
  • Page 23: Proper Moisture Analysis

    USER MANUAL 11.7 Proper moisture analysis Before measurement carefully prepare the sample (as described in chapter Description of Thermogravimetric Analysis) and set correct drying parameters (see chapter Working Parameters Setting) m0-m/m0*100% Ts=120C ts= 0:10:00s Place an empty single-use pan and tare the balance with ...
  • Page 24: Internal Thermometer Indications Correction

    USER MANUAL 11.8 Internal thermometer indications correction necessary point indication m0-m/m0*100% Ts=120C ts= 0:10:00s correction internal moisture analyzer thermometer based T = 32.23C measurement made external m= 0.000g 0.00 control thermometer available t =0:00:00s 0.00 (inscribe temperature values T1 and m0= 0.000g their equivalent...
  • Page 25 USER MANUAL The way of entering control thermometer probe to moisture analyzer drying chamber 1 – control thermometer probe 2 – 15 mm distance sleeve that ensures proper level above pan 3 – moisture analyzer internal sensor Before executing temperature correction (inscribing T1 and T2 temperature) drying cycle must be made with inscribed T1 temperature and drying time 15 minutes.
  • Page 26 USER MANUAL Moisture analyzer internal thermometer correction is made with internal thermometer and control thermometer on the same level (11mm) above the sample. Attention: The temperature indicated by thermometer situated on some level above the sample can differ from real temperature of the sample. In this case if there is a need for temperature indication correction simply lower the level of control thermometer by removing distance sleeve.
  • Page 27: Connecting To A Computer Or A Printer

    USER MANUAL 11.9 Connecting to a computer or a printer When drying process is finished measurement result can be printer or a computer via RS232C interface. Measuring data can be also completed with text information. To enter text descriptions it is necessary to connect a computer keyboard to PS2 port at the back of the moisture analyser.
  • Page 28 USER MANUAL A set of characters available using the keyboard while you use Product name, Executive or Notes: 1 . , ‟ ? ! ” - ( ) @ / : _ ; + & % * = < > $ [ ] { }\ ~ ^ „ # | 2 A B C a b c 3 D E F d e f 4 G H I g h i...
  • Page 29 15. Quantity 16. Average 17. Remarks 1. Units 18. Performance 2. Auto-zeroing 19. Signature 3. Interface 4. LCD settings <PL><ENG><DE><ESP><ITA><FRA> 5. Language 6. Time&Date 7. Keyboard 8. Defaults 9. Calibration 10. Info MODEL 11. Exit SOFT AXIS Sp. z o.o.
  • Page 30: Testing And Calibration Of The Balance

    USER MANUAL 13. Testing and calibration of the balance To check the weighing function of balance – drier, switch it to the simple weighing (  key) and check it by putting precisely weighed object, e.g. calibration weight F2 (OIML), equal to device measurement range.
  • Page 31: Moisture Analyser As A Balance

    USER MANUAL 14. Moisture analyser as a balance The moisture analyser may be also used as a normal balance. To switch between analyser / weighing mode press  key. In weighing mode MENU key opens a set of special functions. Standard functions are described below.
  • Page 32: Auto-Zeroing

    USER MANUAL 14.2 Auto-zeroing Auto-zeroing function causes that the close to zero indication will be corrected automatically and when the pan is unbiased zero indication will be hold independently even when environment conditions change (temperature, air density etc). USER MENU 1.
  • Page 33: Calibration

    USER MANUAL 14.3 Calibration Calibration with external weight standard should be performed in case indications exceed permissible error (for example more than 5 graduation overflow). To scale calibration use weight standard presented in technical data table (or more precise). Depending on the value of gravity acceleration the producer sets the scale to specific location of use.
  • Page 34 USER MANUAL CONFIGURATION 1. Units 2. Auto-zeroing 3. Interface 4. LCD settings 5. Language 6. Time&date 7. Keyboard 8. Defaults 9. Calibration 10. Info Exit CALIBRATION 1. Calibration: 2. Load : 100g 3. Exit CALIBRATION Please wait ... CALIBRATION Taring CALIBRATION Put load Put the standard of mass...
  • Page 35: Maintenance And Repairs Of Small Defects

    USER MANUAL 15. Maintenance and repairs of small defects 1. A moisture analyser should be kept clean. 2. Take care that no dirt gets between the casing and the pan. If found any, remove the pan (lift it up), remove dirt and then replace the pan. 3.
  • Page 36: Troubleshooting

    USER MANUAL Troubleshooting: Display indication Possible cause Remedy Initial heating Ts temperature exceeds The temperature sensor is Contact an authorised service 105°C, the sensor damaged. point. does not react when touched with a finger Initial heating Ts temperature does not reach 105°C, The heater is damaged.
  • Page 37: Declaration Of Conformity

    USER MANUAL Declaration of Conformity AXIS Spółka z o.o. 80-125 Gdańsk, ul.Kartuska 375B, Poland confirm with all responsibility that moisture analysers: AGS60, AGS120 i AGS210 AGS60/T250, AGS120/T250, AGS210/T250 marked with CE mark comply with the following: 1. EN 61010-1 standard Safety requirements for electrical equipment for measurement, control and laboratory use.
  • Page 38: Drying Parameters For Different Substances (Examples)

    USER MANUAL Drying parameters for different substances (examples) Initial weight Temperature Substance Preparation Analysing time (min) (°C) Acrylate seal mix a sample Granulated acryl 10-15 Acryl ester mix a sample Active coal Active coal Cream 10.9 Cream 10.8 Cotton seeds grind a sample for 1 min.
  • Page 39 USER MANUAL Initial weight Temperature Substance Preparation Analysing time (min) (°C) Latex Latex LE1 10.8 Latex LE2 Latex O44 Lentil grind a sample for 30 sec. Loess soil 10-15 Loess soil cut into small pieces 14.5 Skimmed milk mix a sample Skimmed milk powder Cottage cheese mix a sample...
  • Page 40 USER MANUAL Notes...