CO Analyzer with Auto-Reference
(Addendum to T300/T300M Operation Manual, P/N 06864)
Copyright 2010-2018
Teledyne API
MANUAL ADDENDUM
MODEL T300U
© TELEDYNE API (TAPI)
9970 CARROLL CANYON ROAD
SAN DIEGO, CA 92131-1106
Toll-free Phone:
Phone:
Fax:
Email:
Website:
800-324-5190
+1 858-657-9800
+1 858-657-9816
api-sales@teledyne.com
http://www.teledyne-api.com/
06867D DCN7970
21 August 2018

Advertisement

Table of Contents
loading
Need help?

Need help?

Do you have a question about the T300 and is the answer not in the manual?

Questions and answers

Subscribe to Our Youtube Channel

Summary of Contents for TELEDYNE API T300

  • Page 1 MANUAL ADDENDUM MODEL T300U CO Analyzer with Auto-Reference (Addendum to T300/T300M Operation Manual, P/N 06864) © TELEDYNE API (TAPI) 9970 CARROLL CANYON ROAD SAN DIEGO, CA 92131-1106 Toll-free Phone: 800-324-5190 Phone: +1 858-657-9800 Fax: +1 858-657-9816 Email: api-sales@teledyne.com Website: http://www.teledyne-api.com/...
  • Page 3 See Warranty page in the T300/T300M Manual - P/N 06864. ABOUT THIS MANUAL This manual is intended for use in conjunction with the Model T300/T300M Operation Manual, part number 06864. Where operation of the Model T300U diverges from that of the Model T300/T300M, this addendum takes precedence.
  • Page 4 Teledyne API - T300U Addendum to T300/T300M Manual This page intentionally left blank. 06867D DCN7970...
  • Page 5 NEVER use any gas analyzer to sample combustible gas(es)! Technical Assistance regarding the use and maintenance of the Note T300/T300M or any other Teledyne API product can be obtained by contacting Teledyne API’s Technical Support Department: Phone: 800-324-5190 Email: api-techsupport@teledyne.com...
  • Page 6 Teledyne API - T300U Addendum to T300/T300M Manual CONSIGNES DE SÉCURITÉ Des consignes de sécurité importantes sont fournies tout au long du présent manuel dans le but d’éviter des blessures corporelles ou d’endommager les instruments. Veuillez lire attentivement ces consignes. Chaque consigne de sécurité...
  • Page 7: Table Of Contents

    3.5.2.1. Calibration Gases ............................22 3.5.3. Basic CO Calibration Procedure ..........................23 4. T300U OPERATING INSTRUCTIONS ..........................25 4.1. Summary of Setup and operation Differences between T300U and T300/T300M Analyzers......... 25 4.2. Operating Modes ................................26 4.2.1. Auto-Reference Mode (A-REF) ..........................27 4.2.2.
  • Page 8 Teledyne API - T300U Addendum to T300/T300M Manual Table of Contents 7.1. Perfoming an Auto-Reference Measurement after Maintenance ..................55 8. PRINCIPLES OF OPERATION............................. 59 8.1. Measurement Method ..............................59 8.1.1. Auto-Reference (A-REF) Ratio ..........................59 8.2. Pneumatic Operation ..............................60 8.2.1.
  • Page 9 Table of Contents Teledyne API - T300U Addendum to T300/T300M Manual Table 4-3: T300U VARS Menu ........................39 Table 4-4: T300U Diagnostic (DIAG) Submenus ..................40 Table 4-5: DIAG - Analog I/O Functions .......................41 Table 4-6: Test Channels Functions Available on the T300U’s Analog Output ..........45 Table 4-7: T300U Hessen GAS ID List ......................47...
  • Page 10 Teledyne API - T300U Addendum to T300/T300M Manual Table of Contents This page intentionally left blank. viii 06867D DCN7970...
  • Page 11: Introduction

    INTRODUCTION This addendum for the Teledyne API Model T300U is a supplement to the T300/T300M manual (P/N 06864) and provides an overview of the instrument operation and specific details regarding those areas where the T300U is different in design or operating method from the T300.
  • Page 12: Calibration Valve Options For The T300U

    1.3. CALIBRATION VALVE OPTIONS FOR THE T300U The hardware options for the T300x family of analyzers are available in the T300U. For a list and descriptions of these options please see Section 1 of the T300/T300M Operators Manual (P/N 06864).
  • Page 13: Specifications, Approvals, And Warranty

    SPECIFICATIONS, APPROVALS, AND WARRANTY 2.1. SPECIFICATIONS Table 2-1: T300U Basic Unit Specifications PARAMETER SPECIFICATION Min: 0-100 ppb Full scale Ranges Max: 0-100 ppm Full scale (selectable dual ranges and auto ranging supported) Measurement Units ppb, ppm, µg/m , mg/m (selectable) ≤...
  • Page 14: Epa Equivalency Designation

    Designated Reference and Equivalent Methods is published in the U.S. Federal Register – http://www3.epa.gov/ttn/amtic/criteria.html. 2.3. APPROVALS AND CERTIFICATIONS The Teledyne API Model T300/T300M analyzer was tested and certified for Safety and Electromagnetic Compatibility (EMC). This section presents the compliance statements for those requirements and directives. 2.3.1. SAFETY IEC 61010-1:2001 (3rd Edition), Safety requirements for electrical equipment for measurement, control, and laboratory use.
  • Page 15: Getting Started

    T300/T300M manual (P/N 06864). 3.2. INSTRUMENT LAYOUT The front panel of the T300U is identical to that of the T300/T300M (see Figure 3-1 of the T300/T300M Operators Manual, P/N 06864). The Rear Panel is also very similar to that of the T300/T300M (see Figure 3-4 of the T300/T300M Operators Manual, P/N 06864), the only difference being that the instrument’s particulate filter is mounted...
  • Page 16: Figure 3-1: T300U Internal Layout

    Getting Started Teledyne API - T300U Addendum to T300/T300M Manual Figure 3-1: T300U Internal Layout 06867D DCN7970...
  • Page 17: Connections And Setup

    The electrical connection instructions in the T300/T300M Operators Manual apply except as differentiated herein. 3.3.1.1. Connecting Analog Outputs The analog outputs for the T300U are different from those of the T300/T300M. Unlike the T300/T300M which can include options for measuring both O and CO, the T300U is a single gas analyzer (CO) and therefore does not require the fully configurable version of the analog outputs.
  • Page 18: Figure 3-2: T300U Analog Output Connector

    CALIBRATED OUTPUT See Table A-6 of T300/T300M Appendix A for definitions of these DAS data types Optional current loop outputs are available for analog output channels A1& A2. To access these signals attach a strip chart recorder and/or data-logger to the appropriate analog output connections on the rear panel of the analyzer.
  • Page 19: Connecting Status Outputs

    Teledyne API - T300U Addendum to T300/T300M Manual Getting Started 3.3.2. CONNECTING STATUS OUTPUTS The information found in Section 3.3.1.5. of the T300/T300M Operators Manual is correct with the following exception: Table 3-3: Status Output Pin Assignments Status Output #...
  • Page 20: Pneumatics

    Getting Started Teledyne API - T300U Addendum to T300/T300M Manual 3.3.3. PNEUMATICS The pneumatic flow of the T300U is different from the T300/T300M in two ways: •  The addition of a high-flow multi-tube Nafion dryer. • An additional gas path that passes the sample gas though a CO scrubber used during the auto-reference cycle.
  • Page 21: Figure 3-4: T300U Internal Pneumatic Flow - Ambient Zero/Span Valves (Opt 50A)

    Teledyne API - T300U Addendum to T300/T300M Manual Getting Started INSTRUMENT CHASSIS  Multi-Tube Nafion Dryer Sample SAMPLE GAS INLET Sample / Cal Valve Span SPA N INLET Zero / Span Valve Ze ro ZERO INLET SAMPLE FILTER OUT Scrubber...
  • Page 22: Startup, Functional Checks, And Initial Calibration

    3.4.2. WARNING MESSAGES The information found in Section 3.4.2 of the T300/T300M Operators Manual (P/N 06864) is applicable to the T300U with the following exceptions (marked by bold print...
  • Page 23: Functional Checks

    Teledyne API - T300U Addendum to T300/T300M Manual Getting Started 3.4.3. FUNCTIONAL CHECKS The information found in Section 3.4.3 of the T300/T300M Operators Manual (P/N 06864) regarding performing an initial functional check of the analyzer is applicable to the T300U with the following exception(s).
  • Page 24: Initial Calibration

    To perform the following calibration you must have sources for zero air and span gas available for input into the sample port on the back of the analyzer. See Section 3.3.2.1 of the T300/T300M Operators Manual (P/N 06864) for instructions for connecting these gas sources.
  • Page 25: Basic Co Calibration Procedure

    Final Test and Validation Data Sheet. Because the T300U’s analog output setup differs from that of the T300/T300M and because the T300U could not be modified to include the optional CO...
  • Page 26 Getting Started Teledyne API - T300U Addendum to T300/T300M Manual This page intentionally left blank. 06867D DCN7970...
  • Page 27: T300U Operating Instructions

    BETWEEN T300U AND T300/T300M ANALYZERS. For the most part the setup and operation instruction for the T300U are the same as those described in Sections 4, 5, and 6 of the T300/T300M manual (P/N 06864) with the following exceptions: •...
  • Page 28: Operating Modes

    Teledyne API - T300U Addendum to T300/T300M Manual 4.2. OPERATING MODES The information found in Section 4 of the T300/T300M Operators Manual (P/N 06864) is applicable to the T300U with the following exception(s). The following table supersedes Table 4-1 of the T300/T300M Operators Manual (P/N 06864).
  • Page 29: Auto-Reference Mode (A-Ref)

    T300U Operating Instructions 4.2.1. AUTO-REFERENCE MODE (A-REF) One of the most significant differences between the T300/T300M and the T300U analyzers is the auto-reference measurement feature. In this mode, the analyzer makes special measurements and calculations that are applied to the CO concentration...
  • Page 30: Test Functions

    Teledyne API - T300U Addendum to T300/T300M Manual 4.2.2. TEST FUNCTIONS The information found in Section 4.1.1 of the T300/T300M Operators Manual (P/N 06864) is applicable to the T300U with the exception that the following table supersedes Table 4-2 of the T300/T300M Operators Manual (P/N 06864).
  • Page 31: Setup  Rnge: Analog Output Reporting Range Configuration

    Teledyne API - T300U Addendum to T300/T300M Manual T300U Operating Instructions 4.3. SETUP  RNGE: ANALOG OUTPUT REPORTING RANGE CONFIGURATION 4.3.1. PHYSICAL RANGE VERSUS ANALOG OUTPUT REPORTING RANGES Functionally, the T300U analyzer has one hardware “physical range” that is capable of determining CO concentrations between 0 ppb and 100,000 ppb.
  • Page 32 T300U Operating Instructions Teledyne API - T300U Addendum to T300/T300M Manual The A1 and A2 channels output a signal that is proportional to the CO concentration of the sample gas. They can be configured: • With independent reporting ranges reporting a “single” output signal (SNGL Mode, see Section 4.3.3).
  • Page 33: Rnge  Mode  Sngl: Configuring The T300U Analyzer For Single Range Mode

    Teledyne API - T300U Addendum to T300/T300M Manual T300U Operating Instructions 4.3.3. RNGE  MODE  SNGL: CONFIGURING THE T300U ANALYZER FOR SINGLE RANGE MODE The single range mode sets a single maximum range for the both the A1 and A2 analog outputs.
  • Page 34: Rnge  Mode  Dual: Configuring The T300U Analyzer For Dual Range Mode

    T300U Operating Instructions Teledyne API - T300U Addendum to T300/T300M Manual 4.3.4. RNGE  MODE  DUAL: CONFIGURING THE T300U ANALYZER FOR DUAL RANGE MODE DUAL range mode allows the A1 and A2 outputs to be configured with separate reporting range spans as well as separate electronic signal levels. The analyzer software calls these two ranges LOW and HI.
  • Page 35: Rnge  Mode  Auto: Configuring The T300U Analyzer For Auto Range Mode

    Teledyne API - T300U Addendum to T300/T300M Manual T300U Operating Instructions 4.3.5. RNGE  MODE  AUTO: CONFIGURING THE T300U ANALYZER FOR AUTO RANGE MODE AUTO range mode gives the analyzer to ability to output data via a LOW range (displayed on the front panel as RANGE1) and HIGH range (displayed on the front panel as RANGE2) on a single analog output.
  • Page 36 T300U Operating Instructions Teledyne API - T300U Addendum to T300/T300M Manual To set the ranges press: SAMPLE RANGE=50.000 PPM CO= XXXX <TST TST> CAL SETUP SETUP PRIMARY SETUP MENU CFG DAS RNGE PASS CLK MORE EXIT SETUP RANGE MODE MENU...
  • Page 37: Setup  Rnge  Unit: Setting The Reporting Range Unit Type

    Teledyne API - T300U Addendum to T300/T300M Manual T300U Operating Instructions 4.3.6. SETUP  RNGE  UNIT: SETTING THE REPORTING RANGE UNIT TYPE The T300U can display concentrations in ppb, ppm, ug/m , mg/m units. Changing units affects all of the COM port values, and all of the display values for all reporting ranges.
  • Page 38: Setup  More  Aref: Configuring And Performing Auto-Reference Measurements

    ACAL sequence already programmed and activated with the DELTA TIME parameter set for 4 hours (for a more detailed discussion of the ACAL system see Section 9.4 of the T300/T300M Operators Manual).
  • Page 39 Teledyne API - T300U Addendum to T300/T300M Manual T300U Operating Instructions SAMPLE RANGE=50.000 PPM CO= XXXX <TST TST> CAL SETUP SETUP PRIMARY SETUP MENU ACAL DAS RNGE PASS CLK MORE EXIT NOTE SETUP SEQ1) AUTO-REF,0:04:00 DO NOT change the MODE.
  • Page 40: Forcing An Auto-Reference Measurement

    T300U Operating Instructions Teledyne API - T300U Addendum to T300/T300M Manual 4.4.2. FORCING AN AUTO-REFERENCE MEASUREMENT Sometimes it is advisable to perform an auto-reference measurement at other times such as just before calibrating the analyzer. To manually start an auto-reference measurement cycle, press: SAMPLE RANGE=50.000 PPM...
  • Page 41: Setup  More  Vars: Variables Setup And Definition

    T300U Operating Instructions 4.5. SETUP  MORE  VARS: VARIABLES SETUP AND DEFINITION The information found in Section 5.8 of the T300/T300M Operators Manual (P/N 06864) is applicable to the T300U with the following exception(s): • There is no STABIL_GAS variable available under the VARS menu of the T300U.
  • Page 42: Setup  More  Diag: Diagnostics Functions

    T300U Operating Instructions Teledyne API - T300U Addendum to T300/T300M Manual 4.6. SETUP  MORE  DIAG: DIAGNOSTICS FUNCTIONS The information found in Section 5.9 of the T300/T300M Operators Manual (P/N 06864) is applicable to the T300U with the following exception(s): •...
  • Page 43: Analog Output Configuration

    4.7.1. ANALOG I/O CONFIGURATION Because the manner in which the analog outputs are implemented differs between the T300/T300M and the T300U, there are significant differences in the manner in which the ANALOG I/O are manipulated. This section replaces Section 5.9.3 of the T300/T300M Operators Manual (P/N 06864).
  • Page 44 T300U Operating Instructions Teledyne API - T300U Addendum to T300/T300M Manual To configure the analyzer’s three analog outputs, set the electronic signal type of each channel and calibrate the outputs by: 1. Selecting an output type (voltage or current, if an optional current output driver has been installed) and the signal level that matches the input requirements of the recording device attached to the channel.
  • Page 45 Teledyne API - T300U Addendum to T300/T300M Manual T300U Operating Instructions SAMPLE RANGE=50.000 PPM CO= XXXX <TST TST> CAL SETUP SETUP PRIMARY SETUP MENU CFG DAS RNGE PASS CLK MORE EXIT SETUP SECONDARY SETUP MENU COMM VARS DIAG AREF EXIT...
  • Page 46: Analog Output Calibration

    T300U Operating Instructions Teledyne API - T300U Addendum to T300/T300M Manual 4.7.2. ANALOG OUTPUT CALIBRATION This section corresponds to Section 5.9.3.2 of the T300/T300M Operators Manual (P/N 06864). ANALOG OUPUT calibration needs to be carried out on first startup of the analyzer (performed in the factory as part of the configuration process) or whenever re-calibration is required.
  • Page 47: Selecting A Test Channel Function For Output A4

    Teledyne API - T300U Addendum to T300/T300M Manual T300U Operating Instructions 4.7.3. SELECTING A TEST CHANNEL FUNCTION FOR OUTPUT A4 This section replaces Section 5.9.8.1 of the T300/T300M Operators Manual (P/N 06864) The Test Functions available to be reported on analog output A4 are: Table 4-6: Test Channels Functions Available on the T300U’s Analog Output...
  • Page 48 T300U Operating Instructions Teledyne API - T300U Addendum to T300/T300M Manual To activate the TEST Channel and select a function, press: SAMPLE RANGE=50.000 PPM CO= XXXX <TST TST> CAL SETUP SETUP PRIMARY SETUP MENU CFG DAS RNGE PASS CLK MORE...
  • Page 49: Hessen Protocol

    Teledyne API - T300U Addendum to T300/T300M Manual T300U Operating Instructions 4.7.4. HESSEN PROTOCOL The information found in Section 6.7.2 of the T300/T300M Operators Manual (P/N 06864) is applicable to the T300U with the following exception(s): • There is only one default gas type programmed into the T300U, as follows:...
  • Page 50 T300U Operating Instructions Teledyne API - T300U Addendum to T300/T300M Manual This page intentionally left blank. 06867D DCN7970...
  • Page 51: Calibration Procedures

    CALIBRATION PROCEDURES Calibration of the T300U should be performed according to the procedures described in Sections 9 & 10 of the T300/T300M Manual - P/N 06864 with the following notes and exceptions: • Delivering span and zero gases for the higher resolution the T300U can be difficult.
  • Page 52: Manual Calibration Checks

    Calibration Procedures Teledyne API - T300U Addendum to T300/T300M Manual 5.2.1. MANUAL CALIBRATION CHECKS The information found in Section 9.2.2 of the T300/T300M Operators Manual (P/N 06864) is applicable to the T300U with the following exception(s). • Set the display to show the STABILITY test function.
  • Page 53: Co Calibration Quality

    Teledyne API - T300U Addendum to T300/T300M Manual Calibration Procedures 5.2.6. CO CALIBRATION QUALITY The information found in Section 9.5 of the T300/T300M Operators Manual (P/N 06864) is applicable to the T300U. 5.2.7. CALIBRATION OF OPTIONAL SENSORS There are no optional sensors in the T300U.
  • Page 54 Calibration Procedures Teledyne API - T300U Addendum to T300/T300M Manual This page intentionally left blank. 06867D DCN7970...
  • Page 55: Epa Protocol Calibration

    EPA PROTOCOL CALIBRATION The information found in Section 10 of the T300/T300M Operators Manual (P/N 06864) is applicable to the T300U. 06867D DCN7970...
  • Page 56 EPA Protocol Calibration Teledyne API - T300U Addendum to T300/T300M Manual This page intentionally left blank. 06867D DCN7970...
  • Page 57: Instrument Maintenance

    INSTRUMENT MAINTENANCE The information found in Section 11 of the T300/T300M Operators Manual (P/N 06864) is applicable to the T300U with the following exception(s): • The oven temperature should be tracked along with the other parameters listed on Table 11-2 of the T300/T300M Operators Manual (see Table 7-1 below).
  • Page 58 Instrument Maintenance Teledyne API - T300U Addendum to T300/T300M Manual This page intentionally left blank. 06867D DCN7970...
  • Page 59 Table 7-1: T300U Test Function Record DATE RECORDED OPERATING FUNCTION MODE* STABILITY ZERO CAL CO MEAS ZERO CAL ZERO CAL MR RATIO SPAN CAL PRES SAMPLE SAMPLE AFTER DRIVE WARM-UP SLOPE SPAN CAL OFFSET ZERO CAL OVEN SAMPLE TEMP 06867D DCN7970...
  • Page 60 Instrument Maintenance TELEDYNE API - T300U ADDENDUM TO T300/T300M MANUAL This page intentionally left blank. 06867D DCN7970...
  • Page 61: Principles Of Operation

    06864) is applicable to the T300U with the following addition. It is recommended that you read that section before continuing. 8.1.1. AUTO-REFERENCE (A-REF) RATIO The higher resolution of the T300U makes it more susceptible then the T300/T300M to the effects of a variety of environmental conditions such as: •...
  • Page 62: Pneumatic Operation

    With the exceptions to the difference discussed on other portions of this addendum  regarding the A-REF CO scrubber and the Nafion dryer, the pneumatic operation is the same as that described in Section 13.3 of the T300/T300M Operators Manual (P/N 06864). 06867D DCN7970...
  • Page 63: The A-Ref Carbon Monoxide Scrubber

    Teledyne API - T300U Addendum to T300/T300M Manual Principles of Operation 8.2.1. THE A-REF CARBON MONOXIDE SCRUBBER A special CO scrubber is used to remove all of the CO from the gas stream when the instrument is performing an auto-reference measurement. The scrubber contains a...
  • Page 64: Figure 8-3: Semi-Permeable Membrane Drying Process

    Principles of Operation Teledyne API - T300U Addendum to T300/T300M Manual Figure 8-3: Semi-Permeable Membrane Drying Process 06867D DCN7970...
  • Page 65 Teledyne API - T300U Addendum to T300/T300M Manual Principles of Operation  The dryer consists of a bundle of 0.030" O.D. Nafion tubes housed within a single large tube shell. Sample gas flows within the Nafion tubes while water vapor absorbs into the tubular membrane walls and is removed.
  • Page 66: Electronic Operation

    Teledyne API - T300U Addendum to T300/T300M Manual 8.3. ELECTRONIC OPERATION 8.3.1. OVERVIEW The information found in Section 13.4 of the T300/T300M Operators Manual (P/N 06864) is applicable to the T300U with the following exception(s): • This figure replaces Figure 13-9 of the T300/T300M Operators Manual.
  • Page 67: The Relay Pca

    INTERNAL PUMP M/R Sensor Oven Heater Figure 8-4: T300U Electronic Overview Block Diagram 8.3.2. THE RELAY PCA The information found in Section 13.4.4 of the T300/T300M Operators Manual (P/N 06864) is applicable to the T300U with the following exception(s): 06867D DCN7970...
  • Page 68: Temperature Control Of The Convection Oven

    Principles of Operation Teledyne API - T300U Addendum to T300/T300M Manual 8.3.2.1. Temperature Control of the Convection Oven The information found in Section 13.4.4.1 of the T300/T300M Operators Manual (P/N 06864) is applicable to the T300U with the following exception(s): •...
  • Page 69: Oven Heater Ac Power Configuration

    Teledyne API - T300U Addendum to T300/T300M Manual Principles of Operation 8.3.2.2. Oven Heater AC Power Configuration The two main heaters for the convection oven area of the T300U are AC heaters. A jumper, located in line on the power connection between the relay PCA and the heaters, set the heaters in one of two configurations: •...
  • Page 70: Status Leds

    Principles of Operation Teledyne API - T300U Addendum to T300/T300M Manual 8.3.2.3. Status LEDs This information replaces that found in Sections 13.4.4.5 of the T300/T300M Operators Manual. • Eight LEDs, described in Table 8-2 are located on the analyzer’s relay board to show the current status on the various control functions performed by the relay board (see Figure 13-15 of the T300/T300M Operators Manual).
  • Page 71: Power Distribution

    Teledyne API - T300U Addendum to T300/T300M Manual Principles of Operation 8.3.4. POWER DISTRIBUTION The information found in Section 13.4.7 of the T300/T300M Operators Manual (P/N 06864) is applicable to the T300U with the following exception(s): • The following figure replaces Figure 13-16 of the T300/T300M Operators Manual (P/N 06864).
  • Page 72 Principles of Operation Teledyne API - T300U Addendum to T300/T300M Manual This page intentionally left blank. 06867D DCN7970...
  • Page 73: Troubleshooting & Service

    DO NOT SHORTEN OR TOUCH ELECTRIC CONNECTIONS WITH METALLIC TOOLS WHILE OPERATING INSIDE THE ANALYZER. USE COMMON SENSE WHEN OPERATING INSIDE A RUNNING ANALYZER. The information found in Section 12 of the T300/T300M Operators Manual (P/N 06864) is applicable to the T300U with the exceptions below. CAUTION ALWAYS ALLOW THE T300U TO OPERATE FOR AT LEAST 24 HOURS AFTER ALL REPAIRS HAVE BEEN FINISHED AND THE INSTRUMENT’S CASE IS CLOSED.
  • Page 74: Fault Diagnosis With Warning Messages

    Troubleshooting & Service Teledyne API - T300U Addendum to T300/T300M Manual 9.2. FAULT DIAGNOSIS WITH WARNING MESSAGES The following Table replaces Table 13-1 of the T300/T300M Operators Manual. Table 9-1: Warning Messages - Indicated Failures WARNING FAULT CONDITION POSSIBLE CAUSES...
  • Page 75 Teledyne API - T300U Addendum to T300/T300M Manual Troubleshooting & Service WARNING FAULT CONDITION POSSIBLE CAUSES MESSAGE If sample pressure is < 10 in-hg: o Blocked particulate filter Sample Pressure is o Blocked sample inlet/gas line <10 in-Hg or > 35 in-Hg...
  • Page 76: Fault Diagnosis With Test Functions

    If the Range selected is too small, the recording device will over range. If the Range is too big, the device will show minimal or no apparent change in readings. Indicates noise level of instrument or CO concentration of sample gas (See Section 12.4.2 of the T300/T300M STABIL Operators Manual for causes).
  • Page 77 Teledyne API - T300U Addendum to T300/T300M Manual Troubleshooting & Service TEST INDICATED FAILURE(S) FUNCTIONS As Displayed) If this drive voltage is out of range it may indicate one of several problems: - A poor mechanical connection between the various components in inside the detector housing PHT DRIVE - An electronic failure of the IR Photo-Detector’s built-in cooling circuitry, or;...
  • Page 78: Relay Board Status Leds

    Troubleshooting & Service Teledyne API - T300U Addendum to T300/T300M Manual 9.2.2. RELAY BOARD STATUS LEDs The information found in Section 12.1.4.3 of the T300/T300M Operators Manual (P/N 06864) is applicable to the T300U with the following exception: • The following table replaces Table 12-5 of the T300/T300M Operators Manual.
  • Page 79: Gas Flow Problems

    Teledyne API - T300U Addendum to T300/T300M Manual Troubleshooting & Service 9.3. GAS FLOW PROBLEMS Along with the information found in Section 12.2 of the T300/T300M Operators Manual  (P/N 06864), also be aware that problems with the flow of gas through the Nafion dryer could also be a cause of pressure or flow problems through the T300U.
  • Page 80: Subsystem Checkout

    Teledyne API - T300U Addendum to T300/T300M Manual 9.5. SUBSYSTEM CHECKOUT 9.5.1. RELAY BOARD The information found in Section 12.5.6 of the T300/T300M Operators Manual (P/N 06864) is applicable to the T300U with the following exception: • The following table replaces Table 12-8 of the T300/T300M Operators Manual.
  • Page 81: Technical Assistance

    Teledyne API - T300U Addendum to T300/T300M Manual Troubleshooting & Service 9.6. TECHNICAL ASSISTANCE If this addendum and its trouble-shooting / repair sections do not solve your problems, technical assistance may be obtained from: Teledyne API Technical Support 9970 Carroll Canyon Rd...
  • Page 82 Troubleshooting & Service Teledyne API - T300U Addendum to T300/T300M Manual This page intentionally left blank. 06867D DCN7970...
  • Page 83 T300U and the M300EU models. For the additional software documentation, please refer to Appendix A in the related operation manual:  T300U - T300/T300M operation manual, PN 06864  M300EU - M300E/EM operation manual, PN 04288...
  • Page 84 APPENDIX A: T300U, M300EU Software Menu Trees, Revision L.8 Teledyne API - T300U, M300EU 05924B DCN5871 This page intentionally left blank. 06867D DCN7970...
  • Page 85 Teledyne API - T300U, M300EU 05924B DCN5871 APPENDIX A: T300U, M300EU Software Menu Trees, Revision L.8 APPENDIX A: T300U, M300EU Software Menu Trees, Revision L.8 SAMPLE TEST CALZ CALS SETUP Press to <TST TST> HIGH HIGH HIGH Press to cycle...
  • Page 86 APPENDIX A: T300U, M300EU Software Menu Trees, Revision L.8 Teledyne API - T300U, M300EU 05924B DCN5871 Figure A-2: Primary Setup Menu (Except DAS) 06867D DCN7970...
  • Page 87 Teledyne API - T300U, M300EU 05924B DCN5871 APPENDIX A: T300U, M300EU Software Menu Trees, Revision L.8 Primary Setup Menu  DAS Submenu Figure A-3: 06867D DCN7970...
  • Page 88 APPENDIX A: T300U, M300EU Software Menu Trees, Revision L.8 Teledyne API - T300U, M300EU 05924B DCN5871 SAMPLE SETUP MORE RNGE PASS ACAL COMM DIAG VARS ENTER PASSWORD: 818 INET HESN COM1 COM2 <SET SET> EDIT PREV NEXT JUMP EDIT PRNT...
  • Page 89 Teledyne API - T300U, M300EU 05924B DCN5871 APPENDIX A: T300U, M300EU Software Menu Trees, Revision L.8 Secondary Setup Menu  Hessen Protocol Submenu Figure A-5: 06867D DCN7970...
  • Page 90 APPENDIX A: T300U, M300EU Software Menu Trees, Revision L.8 Teledyne API - T300U, M300EU 05924B DCN5871 Figure A-6: DIAG Menu 06867D DCN7970...
  • Page 91 Appendix B - Interconnect Diagram 06867D DCN7970...