TELEDYNE API T100 User Manual

Uv fluorescence so2 analyzer
Table of Contents

Advertisement

Quick Links

UV Fluorescence SO
Copyright 2019
Teledyne API
User Manual
Model T100
with NumaView™ software
© TELEDYNE API (TAPI)
9970 CARROLL CANYON ROAD
SAN DIEGO, CALIFORNIA 92131-1106
USA
Toll-free Phone:
+1 800-324-5190
Phone:
+1 858-657-9800
Fax:
+1 858-657-9816
Email:
api-sales@teledyne.com
Website:
http://www.teledyne-api.com/
Analyzer
2
083730100 DCN8060
25 March 2019

Advertisement

Table of Contents
loading
Need help?

Need help?

Do you have a question about the T100 and is the answer not in the manual?

Questions and answers

Subscribe to Our Youtube Channel

Summary of Contents for TELEDYNE API T100

  • Page 1 User Manual Model T100 UV Fluorescence SO Analyzer with NumaView™ software © TELEDYNE API (TAPI) 9970 CARROLL CANYON ROAD SAN DIEGO, CALIFORNIA 92131-1106 Toll-free Phone: +1 800-324-5190 Phone: +1 858-657-9800 Fax: +1 858-657-9816 Email: api-sales@teledyne.com Website: http://www.teledyne-api.com/ Copyright 2019 083730100 DCN8060...
  • Page 3 NOTICE OF COPYRIGHT © 2019 Teledyne API (TAPI). All rights reserved. TRADEMARKS All trademarks, registered trademarks, brand names or product names appearing in this document are the property of their respective owners and are used herein for identification purposes only.
  • Page 4 NEVER use any gas analyzer to sample combustible gas(es)! For Technical Assistance regarding the use and maintenance of this instrument or any other Teledyne API product, contact Teledyne API’s Technical Support Department: Telephone: 800-324-5190 Email: api-techsupport@teledyne.com...
  • Page 5 été prévu, l’instrument pourrait se comporter de façon imprévisible et entraîner des conséquences dangereuses. NE JAMAIS utiliser un analyseur de gaz pour échantillonner des gaz combustibles! 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 6: Warranty

    PRODUCT RETURN All units or components returned to Teledyne API should be properly packed for handling and returned freight prepaid to the nearest designated Service Center. After the repair, the equipment will be returned, freight prepaid.
  • Page 7 Provides information about that which could either affect accuracy of instrument readings or cause loss of data. Provides information pertinent to the proper care, Note operation or maintenance of the instrument or its parts. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 8: Table Of Contents

    Alerts: Warnings and Other Messages ..............52 Functional Checks ....................53 Calibration......................53 Menu Overview ..........................54 Home Page .......................... 55 Dashboard ..........................57 Alerts ............................ 58 Calibration ..........................59 Utilities ..........................60 Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 9 Zero Calibration Check and Actual Calibration ............. 94 Span Calibration Check and Actual Calibration ............ 94 Calibration and Check Procedures with Valve Options Installed ......... 95 Use of Zero/Span Valve with Remote Contact Closure ........96 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 10 AC Main Power ....................133 DC Power Supply ....................134 C Bus ......................... 135 LCD/Display Module .................... 135 Relay PCA ......................135 Motherboard ......................136 Pneumatic Pressure / Flow Sensor Assembly ............ 137 viii Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 11 Light Pollution ...................... 173 Oxygen (O ) Sensor Principles of Operation ................173 Paramagnetic Measurement of O ..................173 Sensor Operation within the T100 Analyzer ..............174 Carbon Dioxide (CO ) Sensor Principles of Operation ..............175 NDIR Measurement of CO ....................175 Operation within the T100 Analyzer ................
  • Page 12: List Of Figures

    Figure 2-21. Gas Line Connection w/Zero Scrubber and Internal Span Source (IZS) Option (OPT 50G) . 44 Figure 2-22. Pneumatic Connections for Precision Calibration when IZS Generator Present ....44 Figure 2-23. Pneumatics, Basic Configuration.................... 46 Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 13 Figure 5-8. Critical Flow Orifice Assembly ....................112 Figure 5-9. Flow Calibration Menu ......................116 Figure 5-10. Simple Leak Check Fixture ....................117 Figure 5-11. Hydrocarbon Scrubber Leak Check Setup ................117 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 14 Figure 6-10. Gas Flow and Location of Critical Flow Orifice ..............177 Figure 6-11. Flow Control Assembly & Critical Flow Orifice ..............178 Figure 6-12. T100 Hydrocarbon Scrubber (Kicker) ................... 179 Figure 6-13. Electronic Block Diagram ..................... 181 Figure 6-14. CPU Board ..........................183 Figure 6-15.
  • Page 15 Table 2-18. LAN/Ethernet Configuration Properties ................... 78 Table 3-1. Ethernet Status Indicators ......................81 Table 3-2. Teledyne API's Hessen Protocol Response Modes ..............85 Table 3-3. Hessen Status Flags and Default Bit Assignments ..............86 Table 3-4. Hessen Gas List Definitions ....................... 87 Table 4-1.
  • Page 16: Introduction, Specifications, Approvals, & Compliance

    SO present in the sample gas. The T100’s exceptional stability is achieved with the use of an optical shutter to compensate for sensor drift and a reference detector to correct for changes in UV lamp intensity.
  • Page 17: Specifications

    Intended for Indoor Use Only at Altitudes ≤ 2000m As defined by the US EPA. Defined as twice the zero noise level by the US EPA. All specifications are based on constant conditions. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 18: Epa Designation

    EPA Federal Equivalent Method Number EQSA-0495-100 for sulfur dioxide measurement. The official List of Designated Reference and Equivalent Methods is published in the U.S. Federal Register: http://www3.epa.gov/ttn/amtic/criteria.html, and specifies the instrument’s settings and configurations. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 19: Safety

    EN 15267 – Air Quality – Ambient Air Automated Measuring Systems EN 14212 – Ambient Air Measurement for SO2 For additional certifications, please contact Technical Support: Toll-free Phone: +1 800-324-5190 Phone: +1 858-657-9800 Fax: +1 858-657-9816 Email: api-techsupport@teledyne.com 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 20: Getting Started

    Do not operate this instrument without first removing dust plugs from SAMPLE and EXHAUST ports on the rear panel. Note Teledyne API recommends that you store shipping containers and materials for future use if/when the instrument should be returned to the factory for repair and/or calibration service. See...
  • Page 21: Ventilation Clearance

    MINIMUM REQUIRED CLEARANCE 10 cm / 4 in Back of the instrument 2.5 cm / 1 in Sides of the instrument 2.5 cm / 1 in Above and below the instrument 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 22: Instrument Layout

    The front panel (Figure 2-1) includes two USB ports for peripheral device connections, which can be used with mouse and keyboard as alternatives to the touchscreen interface, or with flash drive for uploads/downloads (devices not included). Figure 2-1. Front Panel Layout Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 23: Rear Panel

    REAR PANEL Figure 2-2 shows the layout of the rear panel. Figure 2-2. Rear Panel Layout, Base Unit (options include additional pneumatic ports) 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 24: Table 2-2. Rear Panel Description

    ANALOG IN Option for external voltage signals from other instrumentation and for logging these signals. Connector for direct connection to laptop computer, using USB cable. Label Includes voltage and frequency specifications. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 25: Internal Chassis

    INTERNAL CHASSIS Figure 2-3 shows internal chassis configurations without options. Figure 2-3. Internal Chassis Layout, Basic (no valve or second gas option) 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 26: Connections And Startup

    To maintain compliance with EMC standards, cable length must be go greater than 3 meters for all I/O connections. Teledyne API recommends that you store shipping containers and materials for future use if/when the instrument should be returned to the factory for repair and/or calibration service.
  • Page 27: Connecting Analog Inputs (Option)

    Analog input # 6 Channel 6 Analog input # 7 Channel 7 Analog input # 8 Channel 8 Analog input Ground See Section 2.5.1 to set up the Data Logger. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 28: Connecting Analog Outputs

    I Out + through the Ground I Out - Setup>Analog I Out + Outputs menu. Ground I Out - I Out + Ground I Out - Not Available Ground Not Available Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 29: Current Loop Analog Outputs (Option 41) Setup

    Fundamentals of ESD, PN 04786, which can be downloaded from our website at http://www.teledyne-api.com under Help Center > Product Manuals in the Special Manuals section.. Figure 2-6. Current Loop Option Installed on the Motherboard 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 30: Connecting The Status Outputs (Digital Outputs)

    1.2V from its collector to emitter. STATUS Map the digital outputs 1 thru 8 through the Setup>Digital Outputs menu. +5V to external device Figure 2-7. Status Output Connector for Digital Outputs Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 31: Connecting The Control Inputs (Digital Inputs)

    (more convenient), or by a separate external 5 VDC power supply (ensures that these inputs are truly isolated). Refer to Figure 2-8. CONTROL IN CONTROL IN 5 VDC Power Supply External Power Connections Local Power Connections Figure 2-8. Energizing the Control Inputs 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 32: Concentration Alarm Relay (Option 61)

    Ethernet interface connector to an Ethernet port. Although the analyzer is shipped with DHCP enabled by default, it should be manually configured with a static IP address. Configuration: Section 2.5.10.5 Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 33: Figure 2-10. Rear Panel Connector Pin-Outs For Rs-232 Mode

    For RS-232 communications with data terminal equipment (DTE) or with data communication equipment (DCE) connect either a DB9-female-to-DB9-female cable (Teledyne API part number WR000077) or a DB9-female-to-DB25-male cable (Option 60A), as applicable, from the analyzer’s RS-232 port (see Figure 2-10 for connector pin assignments) to the device.
  • Page 34: Figure 2-11. Default Pin Assignments For Cpu Com Port Connector (Rs-232)

    Figure 2-11. Default Pin Assignments for CPU COM Port Connector (RS-232) Teledyne API offers two mating cables, one of which should be applicable for your use. • P/N WR000077, a DB-9 female to DB-9 female cable, 6 feet long. Allows connection of the serial ports of most personal computers.
  • Page 35 2. On the Multidrop/LVDS PCA’s JP2 connector, remove the shunt that jumpers Pins 21 ↔ 22 as indicated in. (Do this for all but the last instrument in the chain where the shunt should remain at Pins 21 ↔ 22). 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 36: Figure 2-12. Jumper And Cables For Multidrop Mode

    RS232 port; from the second analyzer’s COM2 port to the third analyzer’s RS232 port, etc., connecting in this fashion up to eight analyzers, subject to the distance limitations of the RS-232 standard. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 37: Figure 2-13. Rs-232 Multidrop Pca Option Host/Analyzer Interconnect Diagram

    The (communication) Host instrument can address only one Note instrument at a time, each by its unique ID (see step 7 above). Teledyne API recommends setting up the first link, between the Note Host and the first analyzer, and testing it before setting up the rest of the chain.
  • Page 38: Pneumatic Connections

    Remove the permeation tube when taking the analyzer out of operation and store in sealed container (use the original shipping packaging). (See Section 5.6.2 for instructions on how to remove the permeation tube when the unit is not in operation). Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 39: Critical Tubing, Pressure, Venting And Exhaust Requirements

    Venting (to prevent back diffusion and pressure effects): Run tubing outside the enclosure or at least away from immediate area surrounding • the instrument. Exhaust Outlet: • Run tubing outside the enclosure. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 40: Basic Connections From Calibrator

    SAMPLE GAS Calibrated (Remove during calibration) span gas concentration MODEL T701 SAMPLE Zero Gas Chassis Generator VENT EXHAUST Figure 2-15. Gas Line Connections from Bottled Span Gas – Basic Configuration Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 41: Connections W/Ambient Zero/Ambient Span (Z/S) Valves (Opt 50A)

    (ZERO AIR and SPAN1) • Figure 2-16. Rear Panel Layout with Z/S Valve Options (OPT 50A) 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 42: Figure 2-17. Gas Line Connections With Z/S Valves Option (Opt 50A)

    ALIBRATION OURCES SPAN GAS Attach a gas line from the source of calibration gas (e.g. a Teledyne API's T700 Dynamic Dilution Calibrator) to the SPAN1 inlet. ZERO AIR Zero air is supplied by the zero air generator such as a Teledyne API's M701.
  • Page 43: Connections W/Ambient Zero/Pressurized Span Valves (Opt 50E)

    (ZERO AIR, SPAN and VENT) • Figure 2-18. Rear Panel Layout with Ambient Zero/Pressurized Span Valves (OPT 50E) 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 44: Figure 2-19. Gas Line Connection W/Ambient Zero/Pressurized Span Valves Option (Opt 50E)

    SPAN GAS Attach a gas line from the pressurized source of calibration gas (e.g. a bottle of NISTSRM gas) to the SPAN1 inlet. Use PTFE tubing, minimum O.D ¼”. ZERO AIR (the dual-stage zero Air Scrubber makes zero air) ENTING Vent the SPAN2/VENT outlet. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 45: Zero Scrubber And Internal Span Source (Izs) (Opt 50G)

    • the zero/span subsystem Figure 2-20. Rear Panel Layout with Internal Span Source (IZS) OPT 50G 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 46: Figure 2-21. Gas Line Connection W/Zero Scrubber And Internal Span Source (Izs) Option (Opt 50G)

    ZERO AIR Generator Vent here if output of calibrator External Zero is not already vented. FROM DRYER Air Scrubbe PUMP Figure 2-22. Pneumatic Connections for Precision Calibration when IZS Generator Present Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 47 Although the factors that influence the resulting span gas concentration can be kept at a constant, they also can be manipulated, thereby impacting the concentration. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 48: Pneumatic Flow Diagrams

    ONFIGURATION Chassis HYDROCARBON Particulate SCRUBBER SAMPLE (Kicker) gas inlet Filter EXHAUST gas outlet PUMP LAMP REACTION CELL FLOW SENSOR SAMPLE PRESSURE FLOW PRESSURE SENSOR SENSOR PCA Figure 2-23. Pneumatics, Basic Configuration Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 49: Pneumatic Flow For Zero/Span Valves Option

    ZERO CAL Zero/Span Open to ZERO AIR inlet NO  COM Sample/Cal Open to ZERO/SPAN valve NC  COM SPAN CAL Zero/Span Open to SPAN GAS inlet NC  COM 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 50: Pneumatic Flow For Ambient Zero/Pressurized Span Option

    Span Shutoff Closed Zero Air Shutoff NC  COM Sample/Cal Open to ZERO/SPAN valve NC  COM Zero/Span Open to SPAN inlet SPAN CAL Closed Span Shutoff OPEN Zero Air Shutoff Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 51: Pneumatic Flow For Internal Zero/Span (Izs) Gas Generator Option

    NC  COM SPAN CAL Zero/Span Open to SPAN GAS inlet NC  COM The permeation tube is not included in the IZS Option and must Note be ordered separately. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 52: Pneumatic Flow With O2 Sensor Option

    When the instrument is first started (Section 2.3.4.1), check its functionality (Section 2.3.4.3) and run an initial calibration (Section 2.3.4.4). Section 2.4 introduces the menu system, and Section 2.5 provides setup/customization instructions. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 53: Startup

    (Figure 2-29). Figure 2-28. Status Screens at Startup Upon any startup, this instrument should warm up for approximately one hour before reliable measurements can be taken. Figure 2-29. Home Page Example 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 54: Alerts: Warnings And Other Messages

    Figure 2-30. Viewing Active Alerts Page If alerts of warning conditions persist after the warm up period or after being cleared, investigate their cause using the troubleshooting guidelines in Section 5.7. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 55: Functional Checks

    The method for performing a calibration differs slightly depending on whether or not any of the available internal zero air or valve options are installed. Follow the appropriate calibration instructions presented in Section 4. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 56: Menu Overview

    Time settings*; and check for firmware updates when Section 2.5.9 connected to a network that is connected to the Internet. Time Zone change requires special procedures (Section 5.5). Comm View and configure network and serial communications. Section 2.5.10 Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 57: Home Page

    Section 2.5.6 provides configuration instructions. Figure 2-32. User Interface Orientation 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 58: Figure 2-33. Concentration And Stability Graph (Top) And Meter Graph (Bottom)

    Figure 2-33. Concentration and Stability Graph (top) and Meter Graph (bottom) Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 59: Dashboard

    2.5.3 provides configuration instructions). If there is a graphing icon in the upper right corner of a parameter, pressing that parameter displays a live plot of its readings as in Figure 2-34. Figure 2-34. Dashboard Page 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 60: Alerts

    To clear Alerts from the Active Alerts page, either check individual boxes to choose specific Alerts, or check the Select All box to choose all Alerts, then press the Clear Selected button. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 61: Calibration

    Figure 2-37. Utilities>Alerts Log of Active and Past Alerts and Events CALIBRATION The Calibration menu is used for multipoint calibrations and for external calibration with valve options installed. Calibration procedures are presented in Section 4. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 62: Utilities

    The Setup menu is used to configure the instrument’s software features, gather information on the instrument’s performance, and configure and access data from the Datalogger, the instrument’s internal data acquisition system (DAS). Section 2.5 provides details for the menus under Setup. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 63: Setup Menu: Features/Functions Configuration

    Sections 2.5.1.1 and 2.5.1.2 providing additional details. To transfer captured instrument data to a flash drive see Section 2.5.1.3. Figure 2-38. Datalog Configuration, New Log Page Figure 2-39. Datalog Configuration, Existing Log 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 64: Figure 2-40. Creating A New Data Log

    Figure 2-40. Creating a New Data Log The parameters available in the list of Log Tags include the names of Events configured in the Events page (Section 2.5.2). Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 65: Configuring Trigger Types: Periodic

    Periodic Trigger requires an interval that is set to number of minutes and a start time that is set to date and clock time. Figure 2-41. Datalog Periodic Trigger Configuration 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 66: Configuring Trigger Types: Conditional

    2. Select all or define a period from which to download the collected data. 3. Press the Download button, and when complete, as indicated in the Status field, press the Done button (changed from “Cancel”) and remove the flash drive. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 67: Setup>Events

    Press ADD to create a new Event (refer to Figure 2-45 for details), or select an existing Event to either Edit or Delete it (Figure 2-47). Figure 2-45. Event Configuration 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 68: Editing Or Deleting Events

    Select an Event from the list (Figure 2-44) and press the Edit button to view or edit the details (Figure 2-46), or press the Delete button to delete the Event. Figure 2-47. Edit or Delete an Event Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 69: Using Events As Triggers For Data Logging

    Name field of the Events Configuration page will appear in the list of Log Tags of the Datalog Configuration page. The Data Logger is presented in Section 2.5.1. SETUP>DASHBOARD Figure 2-48. Dashboard Display and Configuration 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 70: Setup>Autocal (With Valve Option)

    MODBUS or Hessen protocols (see Setup>Comm). Max Concentration Set the highest concentration expected, as this is used by the CPU to adjust Range Preamp physical ranges. (Section 4.1.5 provides more information). Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 71: Setup>Homescreen

    Setup>Homescreen menu or from Home page using the configuration icon (Figure 2-49). Figure 2-49. Homescreen Configuration An orientation to the Homescreen was presented in Section 2.4.1, including Figure 2-32 and Figure 2-33. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 72: Setup>Digital Outputs

    Go to the Utilities>Diagnostics>Digital Outputs menu to change the state (ON/OFF) of individual digital outputs. Figure 2-50. Digital Outputs Setup Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 73: Setup>Analog Outputs

    Recorder Offset: add a zero offset for recording slightly negative readings from noise around the zero point. • Allow Overrange: check to allow a ± 5% over-range; uncheck to disable over-range if the recording device is sensitive to excess voltage or current. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 74: Figure 2-52. Analog Outputs Group Calibration Screen

    While these are the physical limits of the current loop modules, typical applications use 2-20 mA or 4-20 mA for the lower and upper limits. For manual calibration adjustments, see Section 2.5.8.1 for voltage and Section 2.5.8.2 for current. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 75: Manual Calibration Of Voltage Range Analog Outputs

    ±0.0005V 90 mV ±0.001V 0.02 mV 1 VDC ±0.001V 900 mV ±0.001V 0.24 mV 5 VDC ±0.002V 4500 mV ±0.003V 1.22 mV 10 VDC ±0.004V 4500 mV ±0.006V 2.44 mV 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 76: Manual Adjustment Of Current Range Analog Outputs

    Figure 2-5 for pin assignments and diagram of the analog output connector). This allows the use of a voltmeter connected across the resistor to measure converter output as VDC or mVDC. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 77: Setup>Instrument

    When an instrument is connected to a network that is connected to the Internet, follow the instructions on this Remote Update page to check for and activate software/firmware updates. (Also refer to Section 5.3). 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 78: Setup>Comm (Communications)

    Enable/disable the requirement for a password for this serial port to respond. The Security only command that is active is the request-for-help command (? CR). Stop bits Select either 0 or 1 stop bit (typically set in conjunction with Parity and Data bits). Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 79: Tcp Port1

    (DHCP). Most users will want to configure the instrument with a static IP address: click the Static radio button to manually assign a static IP address (consult your network administrator, and see Table 2-18 for information). Figure 2-57. Communications Configuration, Network Settings 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 80: Transferring Configuration To Other Instruments

    2. Insert a flash drive into either of the two front panel USB ports. 3. When the Status field indicates that the USB drive has been detected, press the “Download Configuration from Instrument” Start button. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 81 7. When the Status field indicates that the USB drive has been detected, press the “Upload Configuration to Instrument” Start button. 8. When the Status field indicates that the upload is complete, remove the flash drive. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 82: Communications And Remote Operation

    RS232 on instrument rear panel) and/or COM2 (labeled COM2 on instrument rear panel) for communication modes, baud rate and serial communications. If using a USB option communication connection, setup requires that the instrument’s baud rate and personal computer baud rate match. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 83: Serial Communication: Rs-232

    A code-activated switch (CAS), can also be used on either port to connect typically between 2 and 16 send/receive instruments (host computer(s) printers, data loggers, analyzers, monitors, calibrators, etc.) into one communications hub. Contact Teledyne API Sales (front cover, this manual) for more information on CAS systems.
  • Page 84: Communications Protocols

    MODBUS registers are provided in Appendix A. MODBUS These instructions assume that the user is familiar with MODBUS communications, and provide minimal information to get started. Please refer to the Teledyne API MODBUS manual, PN 06276, and to www.modbus.org for MODBUS communication protocols. Minimum Requirements: •...
  • Page 85: Figure 3-2. Modbus Via Serial Communication (Example)

    Setup>Vars>Instrument ID menu. Next, for the settings to take effect, power off the analyzer, wait 5 seconds, and power it on again. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 86: Hessen

    Create a unique identification number for each instrument in Important the multidrop chain via the Setup>Vars>Instrument ID menu. The Hessen protocol is not strictly defined; therefore, while Teledyne API’s application is completely compatible with the protocol itself, it may be different from implementations by other companies.
  • Page 87: Hessen Settings Configuration

    ESPONSE Set the response mode under Hessen Response Mode, referring to Table 3-2 for descriptions. Table 3-2. Teledyne API's Hessen Protocol Response Modes MODE ID MODE DESCRIPTION This is the default setting. Reponses from the instrument are encoded as the traditional command format.
  • Page 88: Table 3-3. Hessen Status Flags And Default Bit Assignments

    Be careful not to assign conflicting flags to the same bit as each status bit will be triggered if any of the assigned flags is active. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 89: Hessen Gas List Configuration

    1 or low range is active only when range 2 or high range is active Not Applicable specific gas identification Reported choice of whether to report when polled by the Hessen network 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 90: Calibration

    Optional equipment: A recording device such as a strip-chart recorder and/or data logger should be used to record data from the T100’s serial or analog outputs. If analog readings are used, the response of the recording system should be checked against a NIST traceable voltage source or meter.
  • Page 91: Zero Air

    If calibrator and zero air source are not available, then use a bottle of SO air. Teledyne API offers an IZS option operating with permeation devices. The accuracy of these devices is about ±5%. Whereas this may be sufficient for quick, daily calibration checks, we strongly recommend using certified SO span gases for accurate calibration.
  • Page 92: Span Gas For Multipoint Calibration

    (dilution ratio of 222, in the mid-range of the system’s capabilities). PHYSICAL RANGE MEASUREMENTS The entire measurement range of the T100 is 0 – 20,000 ppb, but many applications use only a small part of the analyzer’s full measurement range, which then becomes a data resolution challenge where the reported measurement may be barely perceptible.
  • Page 93: Interferents

    NO (contact Technical Support for more information). PERMEATION TUBE OPTIONS Teledyne API offers an optional internal span gas generator that utilizes a permeation tube as a span gas source (see Section 2.3.2.5). The accuracy of these devices is only about ±5%.
  • Page 94: Calibration Procedures

    1.0 PPB (either in the gas graph or in the Dashboard). Otherwise, follow the steps presented in Sections 4.2.1.1 and 4.2.1.2. after connecting the sources of zero air and span gas in either of the following two ways: Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 95: Figure 4-1. Setup For Manual Calibration Without Z/S Valve Or Izs Option

    SAMPLE Gas (Remove 3-way during Valve calibration) Needle valve to control flow Calibrated Chassis SAMPLE (At high EXHAUST concentration) Figure 4-1. Setup for Manual Calibration without Z/S Valve or IZS Option 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 96: Zero Calibration Check And Actual Calibration

    3. Press the Stop button and return to Home screen. 4. In the Dashboard, check and record the Slope(s) and the Offset(s). (See Table Section 4.4, Calibration Quality Analysis, expected/acceptable values). Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 97: Calibration And Check Procedures With Valve Options Installed

    Figure 4-3. Setup for Manual Calibration Check with Z/S Valve or IZS Option Navigate to the Calibration>Zero menu for Zero cal and to the Calibration>Span menu for Span cal (see Figure 4-4) and follow the instructions in Sections 4.2.1.1 and 4.2.1.2 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 98: Use Of Zero/Span Valve With Remote Contact Closure

    It is recommended that contact closures remain closed for at least 10 minutes to establish a reliable reading. • The instrument will stay in the selected mode for as long as the contacts remain closed. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 99: Automatic Zero/Span Cal/Check (Auto Cal)

    Table 4-1 and Table 4-2 show how to set up the operating states of each calibration or check, and Table 4-3 shows how to program the execution of each. Figure 4-5. Auto Cal Page 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 100: Table 4-1. Auto Cal States

    Check Zero Low High Calibrate For each sequence, there are four parameters that control operational details: Date, Time (both in the Start field), Interval, and Duration, as presented in Table 4-3. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 101: Calibration Quality Analysis

    Otherwise, refer to the troubleshooting Section 5.7.6. Table 4-4. Calibration Data Quality Evaluation FUNCTION MINIMUM VALUE OPTIMUM VALUE MAXIMUM VALUE SLOPE -0.700 1.000 1.300 Norm Offset 50.0 mV 250.0 mV 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 102: Epa Protocol Calibration

    Government Publishing Office at http://www.gpo.gov/fdsys/) and with Quality Assurance Guidance documents (available on the EPA website: http://www3.epa.gov/ttn/amtic/qalist.html). Give special attention to specific regulations regarding the use and operation of ambient sulfur dioxide (fluorescence-based) analyzers. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 103: Maintenance And Service

    OFFSET and SLOPE and alter the instrument’s calibration. Alternatively, use the Auto Cal feature described in Section 4.3 with the CALIBRATE attribute set to OFF (not enabled). 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 104: Table 5-1. Maintenance Schedule

    As necessary orifice & sintered filters Desiccant bags Replace Any time PMT housing is opened for maintenance These Items are required to maintain full warranty; all other items are strongly recommended. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 105: Predictive Diagnostics

    Degradation of IZS permeation tube day to day. SO2 Conc (Concentration) (Standard configuration at span): Drift of instrument response; UV Lamp output is excessively Decreasing over time Stable for low. constant concentration 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 106: Operational Health Checks

    The report is generated every 24 hours to a Web services “cloud” where it is available for viewing by Teledyne API technical support personnel. To download the report for your own viewing on a computer or to send to others, insert a flash drive into a front panel USB port and press the Download button, which is enabled when the instrument detects the flash drive.
  • Page 107: Remote Updates

    Once it’s been determined that the firmware is new, the Install button will be enabled; if the firmware version on the flash drive is the 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 108: Touchscreen Display Calibration (For Earlier Instruments)

    Test the accuracy of the calibration by touching parts of the screen and see that the mouse pointer follows your touches. 9. If you press the CANCEL button, the calibration won’t be altered. Otherwise, press the ACCEPT button. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 109: Time Zone Changes

    6. After the Time Zone is implemented first (Steps 1 through 5), then other changes to the date and/or time can be made, and recycling the power is not necessary. Figure 5-5. Time Zone Change Requirements 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 110: Hardware Maintenance Procedures

    2. Open the hinged front panel and unscrew the retaining ring on the filter assembly. Figure 5-6. Replacing the Particulate Filter 3. Carefully remove the retaining ring, PTFE o-ring, glass window and filter element. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 111: Changing/Removing The Izs Permeation Tube

    Do not ship the instrument without removing the permeation tube. The tube continues to emit gas, even at room temperature and will contaminate the entire instrument. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 112: Changing The External Zero Air Scrubber

    10. Adjust the scrubber cartridge such that it does not protrude above or below the analyzer in case the instrument is mounted in a rack. If necessary, squeeze the clips for a tighter grip on the cartridge. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 113: Servicing Critical Flow Orifices

    1. Turn off power to the instrument and vacuum pump. 2. Locate the critical flow orifice on the pressure sensor assembly. 3. Disconnect the pneumatic line. 4. Unscrew the NPT fitting. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 114: Figure 5-8. Critical Flow Orifice Assembly

    8. Using a new filter and o-rings, reassemble the parts as shown in Figure 5-7. 9. Reinstall the NPT fitting and connect all tubing 10. Power up the analyzer and allow it to warm up for 60 minutes. 11. Perform a leak check. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 115: Checking For Light Leaks

    Assembly is removed, ensure to replace the 5 desiccant bags inside the housing. 7. Carefully replace the analyzer cover. If tubing was changed, carry out a pneumatic leak check (Section 5.6.6). 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 116: Checking For Pneumatic Leaks

    3. Pressurize the instrument with the leak checker or tank gas, allowing enough time to fully pressurize the instrument through the critical flow orifice. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 117: Performing Flow Checks/Calibrations

    Ensure that the inlet to the flow meter is at atmospheric pressure. • 3. The sample flow measured with the external flow meter should be within ± 10% of the analyzer’s Flow specification (Table 1-1). 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 118: Checking The Hydrocarbon Scrubber (Kicker)

    This procedure requires a pressurized source of air (chemical composition is unimportant) capable of supplying up to 15 psiA and a leak checking fixture such as the one illustrated in Figure 5-10. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 119: Figure 5-10. Simple Leak Check Fixture

    8. Close the shut-off valve. 9. Wait 5 minutes. If the gauge pressure drops >1 psi within 5 minutes, then the hydrocarbon scrubber has an internal leak and must be replaced. Contact Teledyne API’s Technical Support. 083730100 DCN8060 Teledyne API T100 SO...
  • Page 120: Service And Troubleshooting

    PCA. Note that the analyzer’s DC power wiring is color-coded and these colors • match the color of the corresponding test points on the relay PCA. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 121: Fault Diagnosis With Alerts

    DARK CAL WARNING indicating that too much stray • I C bus failure light is present in the sample • Loose connector/wiring chamber. • PMT preamp board bad or out of cal 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 122 Failure of motherboard • I C Bus failure The CPU unable to communicate RELAY BOARD WARN • Failed Relay Board with the Relay PCA. • Loose connectors/wiring Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 123 • Shutter solenoid stuck in closed position Clears the next time successful span calibration is performed. Clears the next time successful zero calibration is performed. Only appears if the IZS option is installed. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 124: Fault Diagnosis With Dashboard Functions

    It decreases about 1 in-Hg per 1000 ft gain in altitude. A variety of factors such as air conditioning systems, passing storms, and air temperature, can also cause changes in the absolute atmospheric pressure. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 125: Table 5-4. Dashboard Functions - Possible Causes For Out-Of-Range Values

    UV Lamp output decay. High offset could be due to: incorrect span gas concentration/contaminated zero air/leak; Offset < 250 mV low-level calibration off; light leak; aging UV filter; contaminated reaction cell; pneumatic leak. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 126: Using The Diagnostic Signal I/O Functions

    If the front panel displays properly but DS5 does not flash, then the program files may have become corrupted or the CPU may be defective; contact Teledyne API's Technical Support Department (see Section 5.9) because it may be possible to recover operation of the analyzer.
  • Page 127: Relay Pca Watchdog And Status Leds

    Permeation Tube Heater Dark Shutter Zero/Span and IZS Options Zero/Span Valve Zero/Span and IZS Options Sample/CAL Valve Sample Chamber Heater I2C Watchdog LED Figure 5-13. Relay PCA Status LED Locations 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 128: Flow Problems

    The use of an independent, external flow meter to perform a flow check as described in Section 5.6.6.3 is essential. Refer to the pneumatic flow diagrams as needed for reference. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 129: Sample Flow Is Zero Or Low

    If the sample and vacuum pressures are around their nominal values (28 and • <10 in-Hg-A, respectively) and the flow still displays no numerical value, carry out a leak check as described in Section 5.6.6. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 130: High Flow

    Taken together with a zero or low actual flow, this could indicate a clogged sample orifice. Again, monitoring the pressures and flows regularly will reveal such problems, because the pressures would slowly or suddenly change from their nominal, mean values. Teledyne Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 131: Calibration Problems

    3. Confirm the lack of response by supplying SO range value to the analyzer. 4. Check the sample flow for proper value. 5. Check for disconnected cables to the sensor module. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 132: Unstable Zero And Span

    If the Zero button is grayed out, the actual gas concentration must be significantly different from the actual zero point (as per last calibration), which may be for any of several reasons. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 133: Non-Linear Response

    The following section provides an itemized list of the most common dynamic problems with recommended troubleshooting checks and corrective actions. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 134: Excessive Noise

    This section describes how to determine if a certain component or subsystem is actually the cause of the problem being investigated. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 135: Ac Main Power

    If the Jumper block is WHITE the heaters are configured for 115 VAC at 60 Hz. • • If the Jumper block is BLUE the heaters are configured for 220, 240 VAC at 50 Hz. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 136: Dc Power Supply

    Configuration Jumper Figure 5-14. Location of Relay Board Power Configuration Jumper AC Configuration of the pump is accomplished via an in-line, hard wired, set of connections. Call Teledyne API’s Technical Support Department for more information. DC P OWER UPPLY If the analyzer’s AC mains power is working, but the unit is still not operating properly, there may be a problem with one of the instrument’s switching power supplies, which...
  • Page 137: I 2 C Bus

    USB bus may be malfunctioning. You can verify this failure by logging on to the instrument remotely with Teledyne API’s NumaView™ Remote software. If the analyzer responds to remote commands and the display changes accordingly, the touch-screen interface may be faulty.
  • Page 138: Motherboard

    ±10 mV between the measured and displayed voltage, the motherboard may be faulty TATUS UTPUTS + D C G n d SYSTEM_OK 1000 Ω Figure 5-15. Typical Set Up of Status Output Test Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 139: Pneumatic Pressure / Flow Sensor Assembly

    With the sample pump disconnected or turned off, this voltage should be 4500 ± 250 • With the pump running, it should be about 0.2 V less as the sample pressure drops by • about 1 in-Hg-A from ambient pressure. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 140: Cpu

    OMMUNICATIONS RS-232 T ENERAL ROUBLESHOOTING Teledyne API's analyzers use the RS-232 communications protocol to allow the instrument to be connected to a variety of computer-based equipment. Problems with RS-232 connections usually center around 4 general areas: • Incorrect cabling and connectors. See Section 2.3.1.8 under RS-232 Connection for connector and pin-out information.
  • Page 141 LED should flicker as the instrument is receiving data. 8. Ensure that the communications software or terminal emulation software is functioning properly. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 142: Photomultiplier Tube (Pmt) Sensor Module

    The first is +15 V, which powers the supply. • The second is the programming voltage which is generated on the preamplifier board. Adjustment of the HVPS is covered in the factory calibration procedure in Section • 5.7.9.3. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 143: Pmt Temperature Control Pca

    If the red LED located on the top edge of this assembly is not glowing, the control circuit is not receiving power. Check the analyzer’s power supply, the relay board’s power distribution circuitry and the wiring that connects them to the PMT temperature control PCA. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 144: Internal Span Gas (Izs) Generator And Valve Options

    50°C ±1°C. Check the IZS Temp in the Dashboard or the IZS Temp Raw signal in the Utilities>Diagnostics>Analog Inputs menu. At 50° C, the temperature signal from the IZS thermistor should be around 2500 mV. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 145: Temperature Sensors

    If the temperature fails to drop after 20 minutes or to reach this point after 60 • minutes, there is a problem in the cooler circuit. If the control circuit on the preamplifier board is faulty, a temperature of –1° • C will be reported. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 146: Service Procedures

    1. Document all analyzer parameters that may have been changed, such as range, auto-cal, analog output, serial port and other settings before replacing the DOM. 2. Turn off power to the instrument, fold down the rear panel by loosening the mounting screws. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 147: Sensor Module Repair And Cleaning

    Do not look at the UV lamp while the unit is operating. UV light can cause eye damage. Always use safety glasses made from UV blocking material when working with the UV Lamp Assembly. (Generic plastic glasses are not adequate). 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 148: Figure 5-16. Sensor Module Wiring And Pneumatic Fittings

    SO measurement. After any repair or service has been performed on • the sensor module, the T100 should be allowed to warm up for 60 minutes. Always perform a leak check and calibrate the analyzer • before placing it back into service.
  • Page 149: Figure 5-17. Sensor Module Mounting Screws

    5. Remove the three sensor module mounting screws. Mounting Screw Housing Sample Mounting Chamber Mounting Screw Screw Figure 5-17. Sensor Module Mounting Screws Follow the above steps in reverse order to reinstall the sensor module. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 150: Figure 5-18. Hex Screw Between Lens Housing And Sample Chamber

    2. Remove the sample chamber from the PMT lens and filter housing by unscrewing the 4 hex screws that fasten the chamber to the housing. 3. Remove the four lens cover screws. Figure 5-19. UV Lens Housing / Filter Housing Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 151: Figure 5-20. Pmt Uv Filter Housing Disassembled

    Use gloves and a clean plastic covered surface during assembly. Cleanliness of the inside of the light shield, the UV lens filter housing and the PMT lens is especially important. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 152: Figure 5-21. Disassembling The Shutter Assembly

    10. Install UV filter retainer and tighten screws. 11. Install the shutter cover and minifit connector. Tighten 4 shutter cover screws. 12. Reinstall the sensor module and Plug J4 connector into the motherboard. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 153 UV lamp is adjusted to optimize the amount of UV light shining through the UV filter/lens and into the reaction cell. Changes to the physical alignment of the lamp can affect the analyzers ability to accurately measure SO 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 154: Figure 5-22. Shutter Assembly

    Figure 5-22. Shutter Assembly Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 155: Figure 5-23. Uv Lamp Adjustment

    LAMP reads approximately 3600mV before continuing to adjust the lamp position. Adjust the UV reference detector potentiometer (Figure 5-24) until UV >3700mV or < 3300mV LAMP reads as close to 3500mV as possible. .< 600mV Replace the lamp 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 156: Figure 5-24. Location Of Uv Reference Detector Potentiometer

    Assembly. (Generic plastic glasses are not adequate). CAUTION – GENERAL SAFETY HAZARD DO NOT over tighten the thumbscrew, as over-tightening can cause breakage to the lamp and consequently release mercury into the area. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 157 CAUTION – GENERAL SAFETY HAZARD DO NOT over tighten the thumbscrew, as over-tightening can cause breakage to the lamp and consequently release mercury into the area. 11. Calibrate the lamp (Utilities>Diagnostics menu). 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 158: Figure 5-25. Sensor Assembly

    Light from Reaction PMT Temperature Chamber shines Sensor (thermistor) through hole in side of Cold Block Thermo-Electric Cooler (TEC) PMT Heat Exchange Fins TEC Driver PCA Cooling Fan Housing Figure 5-25. Sensor Assembly Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 159 The thermo-electric cooler needs to be mounted flat to the heat sink. If there is any significant gap, the TEC might burn out. Ensure to apply heat sink paste before mounting it and tighten the screws evenly and cross-wise. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 160: Pmt Sensor Hardware Calibration ("Factory Cal")

    Home page meter (Section 2.5.6) or in the Dashboard (Section 2.5.3). 3. Locate the preamplifier board (PMT Preamp PCA) (see Figure 2-3). Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 161: Figure 5-26. Pre-Amplifier Board Layout

    400 ppb and the calculation would be: Target NORM PMT = (2 x 400) + 33 mV Target NORM PMT = 833 mV 10. Set the HVPS coarse adjustment to its minimum setting (0). 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 162: Removing / Replacing The Relay Pca From The Instrument

    The Relay retainer plate installed on the relay PCA covers the lower right mounting screw of the relay PCA. Therefore, when removing the relay PCA, the retainer plate must be removed first. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 163: Frequently Asked Questions

    Figure 5-28. Relay PCA Mounting Screw Locations FREQUENTLY ASKED QUESTIONS The following list was compiled from the Teledyne API's Technical Support Department’s ten most commonly asked questions relating to the analyzer. QUESTION ANSWER Why does the ENTR button Sometimes the ENTR button will disappear if you select a setting that is...
  • Page 164: Technical Assistance

    Teledyne API Technical Support 9970 Carroll Canyon Road San Diego, California 92131-1106 USA Toll-free Phone: 800-324-5190 Phone: +1 858-657-9800 Fax: +1 858-657-9816 Email: api-techsupport@teledyne.com Website: http://www.teledyne-api.com/ Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 165: Principles Of Operation

    The first stage (Equation 6-1) occurs when SO molecules are struck by photons of the appropriate ultraviolet wavelength. In the case of the T100, a band pass filter between the source of the UV light and the affected gas limits the wavelength of the light to approximately 214 nm.
  • Page 166: Figure 6-1. Uv Absorption

    Intensity of the excitation UV light. The absorption coefficient of SO (a constant). Concentration of SO in the sample chamber. The distance between the UV source and the SO molecule(s) being affected (path length). Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 167 (k). and; no interfering conditions are present (such as interfering gases or stray light); the amount of fluorescent light emitted (F) is directly related to the concentration of the SO in the Sample Chamber. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 168: The Uv Light Path

    SO in the sample gas. THE UV LIGHT PATH The optical design of the T100’s sample chamber optimizes the fluorescent reaction between SO and UV Light (refer to Figure 6-2) and assure that only UV light resulting...
  • Page 169: Uv Source Lamp

    Figure 6-2. UV Light Path UV SOURCE LAMP The source of excitation UV light for the T100 is a low pressure zinc-vapor lamp. An AC voltage heats up and vaporizes zinc contained in the lamp element creating a light- producing plasma arc. Zinc-vapor lamps are preferred over the more common mercury-...
  • Page 170: The Reference Detector

    The degree of offset differs from detector to detector and from PMT to PMT and can change over time as these components age. To account for these offsets the T100 includes a shutter, located between the UV Lamp and the source filter that periodically cuts off the UV light from the sample chamber. This happens every 30 minutes.
  • Page 171: Optical Filters

    OPTICAL FILTERS The T100 analyzer uses two stages of optical filters to enhance performance. The first stage conditions the UV light used to excite the SO by removing frequencies of light that are not needed to produce SO *. The second stage protects the PMT detector from reacting to light not produced by the SO * returning to its ground state.
  • Page 172: Pmt Optical Filter

    PMT O PTICAL ILTER The PMT used in the T100 reacts to a wide spectrum of light which includes much of the visible spectrum and most of the UV spectrum. Even though the 214 nm light used to excite the SO is focused away from the PMT, some of it scatters in the direction of the PMT as it interacts with the sample gas.
  • Page 173: Optical Lenses

    SO * being measured by the PMT, eliminating a possible source of measurement offset. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 174: Measurement Interferences

    UV light. While this can be a significant problem for some analyzers, the design of the T100 is very tolerant of variations in sample gas flow rate and therefore does not suffer from this type of interference.
  • Page 175: Light Pollution

    PARAMAGNETIC MEASUREMENT OF O The oxygen sensor used in the T100 utilizes the fact that oxygen is attracted into strong magnetic field while most other gases are not, to obtain fast, accurate oxygen measurements.
  • Page 176: O 2 Sensor Operation Within The T100 Analyzer

    50°C by means of a PID loop. The O sensor assembly itself does not have any serviceable parts and is enclosed in an insulated canister. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 177: Carbon Dioxide

    CO present in the sensor chamber. This dual wavelength method the CO measurement allows the instrument to compensate for ancillary effects like sensor aging and contamination. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 178: Co 2 Operation Within The T100 Analyzer

    Probe Orange wire 22 awg OVDC 5VDC Analog Output Relay PCA Black wire Grey wire 22 awg 22 awg P110 Motherboard Figure 6-9. CO Sensor Option PCA Layout and Electronic Connections Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 179: Pneumatic Operation

    Hg-A), i.e. indicate an absolute pressure referenced against zero (a perfect vacuum). SAMPLE GAS FLOW The Flow of gas through the T100 UV Fluorescence SO Analyzer is created by a small internal pump that pulls air though the instrument. Chassis...
  • Page 180: Flow Rate Control

    FLOW RATE CONTROL The T100 uses a special flow control assembly located in the exhaust vacuum manifold (refer to Figure 6-10) to maintain a constant flow rate of the sample gas through the instrument. This assembly consists of: A critical flow orifice.
  • Page 181: Sample Particulate Filter

    Figure 6-12. T100 Hydrocarbon Scrubber (Kicker) In the T100 some of the cleaned air from the inner tube is returned to be used as the purge gas in the outer tube (refer to Figure 6-12). This means that when the analyzer is first started, the concentration gradient between the inner and outer tubes is not very large and the scrubber’s efficiency is relatively low.
  • Page 182: Pneumatic Sensors

    PNEUMATIC SENSORS The T100 uses two pneumatic sensors to verify gas streams. These sensors are located on a printed circuit assembly, called the pneumatic pressure/flow sensor board. The flow simultaneously enters the sample pressure sensor and the flow sensor from the outlet of the reaction cell.
  • Page 183: Electronic Operation

    Heater UV Reference Detector Sample Cal IZS Option Valve Permeation Option Tube Heater Temperature Sensor IZS Valve TEC Drive PREAMP PCA Option PMT TEC Figure 6-13. Electronic Block Diagram 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 184 CPU and the analyzer’s other major components. Concentration data of the T100 are generated by the Photo Multiplier Tube (PMT), which produces an analog current signal corresponding to the brightness of the fluorescence reaction in the sample chamber.
  • Page 185: Cpu

    Figure 6-14. CPU Board SENSOR MODULE Electronically, the T100 sensor module is a group of components that: create the UV light that initiates the fluorescence reaction between SO and O ;...
  • Page 186: Sample Chamber

    The main electronic components of the sample chamber are the reference detector (refer to Section 6.1.4); the UV Lamp (refer to Section 6.1.3) and its electronically operated shutter (refer to Section 6.1.6); and the sample chamber heating circuit. Figure 6-16. T100 Sample Chamber Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 187: Sample Chamber Heating Circuit

    CPU through the thermistor interface circuitry of the motherboard. PHOTO MULTIPLIER TUBE (PMT) The T100 uses a photo multiplier tube (PMT) to detect the amount of fluorescence created by the SO and O reaction in the sample chamber.
  • Page 188: Figure 6-18. Basic Pmt Design

    PMT, an LED used by the instrument’s optical test function, a thermistor that measures the temperature of the PMT and various components of the PMT cooling system including the Thermo-Electric Cooler (TEC). Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 189: Pmt Cooling System

    HERMOELECTRIC OOLER The core of the T100 PMT cooling system is a solid state heat pump called a thermoelectric cooler (TEC). Thermoelectric coolers transfer heat from a one side of a special set of semiconductor junctions to the other when a DC current is applied. The heat is pumped at a rate proportional to the amount of current applied.
  • Page 190: Tec Control Board

    PMT. Secondly, the gain of the amplified signal can further be adjusted through a potentiometer. These adjustments should only be performed when encountering problems with the software calibration that cannot be rectified otherwise. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 191: Figure 6-20. Pmt Preamp Block Diagram

    E-Test Generator PMT Temp Analog Signal Signal to Motherboard Offset PMT Temp Sensor Temperature Feedback TEC Control Circuit PMT Output Signal (PMT) to Motherboard Figure 6-20. PMT Preamp Block Diagram 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 192: Pneumatic Sensor Board

    The Relay PCA is located in the right-rear quadrant of the analyzer and is mounted vertically on the backside of the same bracket as the instrument’s DC power supplies. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 193: Figure 6-21. Relay Pca Layout (P/N 045230100)

    Never remove this retainer while the instrument is plugged in and turned on. The contacts of the AC relay sockets beneath the shield carry high AC voltages even when no relays are present. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 194: Heater Control

    EATER ONTROL The T100 uses a variety of heaters for its individual components. All heaters are AC powered and can be configured for 100/120 VAC or 220/230VAC at 50-60 Hz. The two sample chamber heaters are electronically connected in parallel for analyzers at 100/120 VAC line power and in series for units configured for 220/230 VAC.
  • Page 195: Heaters For Izs Option (Jp6) And Ac Configuration

    FUNCTION PINS 1 to 8 Common Internal Permeation Tube Oven Heater 2 to 7 Neutral to Load Permeation Tube Heater Figure 6-23. Typical Jumper Set (JP2) Set Up of Heaters 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 196: Valve Control

    The relay board also hosts two valve driver chips, each of which can drive up four valves. In its basic configuration the T100 requires no valve control to operate. However, on units with either the zero/span valve or the IZS option installed, the valve control is used.
  • Page 197: Sensor Inputs

    (see Section 6.5.10.2). The second is located on the PMT Preamplifier PCA and is used only as a reference for the preamplifier circuitry. Its output is neither reported nor stored. 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 198: Analog Outputs

    WARNING – ELECTRICAL SHOCK HAZARD Should the AC power circuit breaker trip, investigate and correct the condition causing this situation before turning the analyzer back on. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 199: Figure 6-24. Power Distribution Block Diagram

    Sample Sample/Cal IZS Option AC POWER for Z/S and Lamp Lamp Permeation Chamber IZS Valve Shutter Power Tube Heaters DC POWER Options Supply Heater Figure 6-24. Power Distribution Block Diagram 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 200: Front Panel Touchscreen/Display Interface

    CPU by means of a touchscreen controller that connects to the CPU via the internal USB bus and emulates a computer mouse. Figure 6-25. Front Panel and Display Interface Block Diagram Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 201: Lvds Transmitter Board

    SOFTWARE OPERATION The analyzer has a high performance, 86-based microcomputer running Windows CE. Inside the WINDOWS CE shell, special software developed by Teledyne API interprets user commands via the various interfaces, performs procedures and tasks, stores data in the CPU’s various memory devices and calculates the concentration of the sample gas.
  • Page 202: Adaptive Filter

    The instrument slope and offset values recorded during the last calibration can be viewed in the Dashboard. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 203: Temperature/Pressure Compensation (Tpc)

    UV light generated in the instrument’s sample chamber. To negate this effect the T100 maintains the sample gas at a stable, raised temperature. Pressure changes can also have a noticeable, if more subtle, effect on the SO concentration calculation.
  • Page 204: Glossary

    Disk On Module, a 44-pin IDE flash drive with up to 128MB storage capacity for instrument’s firmware, configuration settings and data Disk Operating System DRAM Dynamic Random Access Memory Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 205 Per-Fluoro-Alkoxy, an inert polymer; one of the polymers that Du Pont markets as Teflon ® Programmable Logic Controller, a device that is used to control instruments based on a logic level signal coming from the analyzer 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 206 VARS Variables, the variable settings of the instrument Voltage-to-Frequency Zero / Span Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 207: Appendix A - Modbus Registers

    4096 mV reference (REF_4096_MV) Sample flow cc/m IZS temperature °C Vacuum pressure “Hg Pre-amplified UV lamp intensity reading concentration concentration during zero/span calibration, just before computing new slope and offset slope — offset 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 208 (32-bit IEEE 754 format; read/write in high-word, low-word order; read/write) Maps to SO2_SPAN1 variable; target conc. for range #1 Conc. units Maps to SO2_SPAN2 variable; target conc. for range #2 Conc. units Maps to O2_TARG_SPAN_CONC variable Maps to CO2_TARG_SPAN_CONC variable Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 209 10+3 concentration alarm limit #1 exceeded 10+3 concentration alarm limit #2 exceeded In CO calibration mode 11+3 concentration alarm limit #1 exceeded concentration alarm limit #2 exceeded 11+3 083730100 DCN8060 Teledyne API T100 SO Analyzer with NumaView™ Software...
  • Page 210 Concentration alarm option. option. option. with O correction option. Set DYN_ZERO or DYN_SPAN variables to ON to enable calculating new slope or offset. Otherwise a calibration check is performed. External analog input option. Teledyne API T100 Analyzer with NumaView™ Software 083730100 DCN8060...
  • Page 211 Appendix B - Interconnect Diagram 083730100A DCN8060...

Table of Contents