4. PNEUMATIC CONNECTIONS

To prevent dust from getting into the analyzer, it was shipped with small plugs
inserted into each of the pneumatic fittings on the rear panel. Make sure that all dust
plugs are removed before attaching exhaust and supply gas lines.
Table 4-1. Table 3-3:
REAR PANEL LABEL
SAMPLE
EXHAUST
SPAN
ZERO AIR
TO CONVERTER
FROM CONVERTER
REAR PANEL LABEL
FROM ANALYZER
TO ANALYZER
07267B DCN6485
This section supersedes the information contained in the Pneumatic Connections section
in the T101 Manual - PN 07266.
Sample and calibration gases should only come into contact with PTFE
(Teflon) or glass materials. They should not come in contact with FEP or
stainless steel materials.
Figure 4-1 and Figure 4-2 show the most common configurations for gas supply and
exhaust lines to the Model T102 Analyzer. Figure 4-3, Figure 4-4, and Figure 4-5 show
the connections for units with valve options installed.
Please refer to Figure 3-4 and Figure 3-5 for the location of pneumatic connections at the
rear panel of the T102 and the M501-TRS instruments.
Inlet / Outlet Connector Labels and Functions
T102 PNEMATIC CONNECTERS
Connects the sample gas to the analyzer. When operating the analyzer without zero/span
option, this is also the inlet for any calibration gases.
Exhausts the gas sampled by the analyzer. Connect to an outside area away from people.
On units with zero/span/shutoff valve options installed, connect a gas line to the source of
calibrated span gas here.
On Units with zero/span valve or IZS option installed, this port connects the zero air gas or
the zero air cartridge to the analyzer.
Sample gas leaves the T102 to be conditioned by the M501-TRS via this port.
Sample gas returns to the T102 after being conditioned by the M501-TRS via this port.
M501-TRS PNEUMATIC CONNECTERS
Sample gas enters the M501-TRS from the T102 via this port.
Sample gas leaves the M501-TRS to return to the T102 via this port.
NOTE
CAUTION
FUNCTION
FUNCTION
Pneumatic Connections
25