TROUBLESHOOTING and REPAIR
11.5.3. Replacing the PMT, HVPS or TEC
Any time that the black PMT housing end plate cover for the Sensor Assembly is
removed, always replace the 5 desiccant bags inside the housing.
74
6. Install the UV filter. Handle carefully and never touch the filter's
surface. The UV filter's wider ( ring ) side should be facing out.
Install the UV filter retainer and tighten screws.
7. Install the shutter cover and minifit connector. Tighten 4
screws.
8. Re-plug J4 connector into the motherboard.
Figure 11-3. PMT Assembly - Exploded View
The PMT should last for the lifetime of the analyzer. However, in some cases, the high
voltage power supply (HVPS) or the thermo-electric cooler (TEC) may fail.
To replace the PMT, the HVPS or the TEC:
Teledyne API - T102/501 TRS, Addendum to T101 Operation Manual
NOTE
07267B DCN6485