DVANCED OLLUTION NSTRUMENTATION Teledyne Advanced Pollution Instrumentation, Inc. (TAPI) is a worldwide market leader in the design and manufacture of precision analytical instrumentation used for air quality monitoring, continuous emissions monitoring, and specialty process monitoring applications. Founded in San Diego, California, in 1988,...
Page 4
Teledyne API – T200 NOx Analyzer Operation Manual This page intentionally left blank. 06858A DCN5847...
NEVER use a gas analyzer to sample any combustible gas(es)! Note Technical Assistance regarding the use and maintenance of this instrument or any other Teledyne API product can be obtained by contacting Teledyne API’s Customer Service Department: Telephone: 800-324-5190 Email: api-customerservice@teledyne.com or by accessing various service options on our website at http://www.teledyne-api.com/...
Page 6
Teledyne API – T200 NOx Analyzer Operation Manual This page intentionally left blank. 06858A DCN5847...
HEREUNDER, WHETHER FOR BREACH OF WARRANTY OR OTHERWISE. TERMS AND CONDITIONS All units or components returned to Teledyne API should be properly packed for handling and returned freight prepaid to the nearest designated Service Center. After the repair, the equipment will be returned, freight prepaid.
Page 8
Teledyne API – T200 NOx Analyzer Operation Manual This page intentionally left blank. 06858A DCN5847...
(Organization), a description of other information related to this manual (Related Information), and the conventions used to present the information in this manual (Conventions Used). STRUCTURE This T200 manual, PN 06858, is comprised of multiple documents, assembled in PDF format, as listed below. Part No. Rev Name/Description...
Page 10
Teledyne API – T200 NOx Analyzer Operation Manual instructions, and the initial calibration and functional checks. It also has a section with Frequently Asked Questions (FAQs) and a glossary. Part II comprises the operating instructions, which include basic, advanced and remote operation, calibration, diagnostics, testing, validating and verifying, and ends with specifics of calibrating for use in EPA monitoring.
REVISION HISTORY This section provides information regarding changes to this manual. 2010, T200 Manual, PN06858 Rev A, DCN5847 – Initial Release Document Change Summary 06858A DCN5847...
Page 12
Teledyne API – T200 NOx Analyzer Operation Manual This page intentionally left blank. 06858A DCN5847...
3.7.3. Functional Check ...........................55 3.8. Initial Calibration of the T200 Analyzer ......................56 3.8.1. Interferents for NOx, NO and NO2 Measurements ................56 3.8.2. Initial Calibration Procedure for T200 analyzers without Options............56 3.8.2.1. Verifying the T200 Reporting Range Settings................57 3.8.2.2. Verify the Expected NO and NO Span Gas Concentration............58...
Page 14
7.4.4.4. Analog Output Reporting Range Default Settings............... 103 7.4.4.5. RNGE MODE SNGL: Configuring the T200 Analyzer for Single Range Mode ....104 7.4.4.6. RNGE MODE IND: Configuring the T200 analyzer for Independent Range Mode ..105 7.4.4.7.
Page 15
9.4.2. Manually Configuring the Ethernet (Static IP Address) ..............169 9.4.3. Changing the Analyzer’s HOSTNAME ....................171 9.5. Using the T200 with a Hessen Protocol Network ..................172 9.5.1. General Overview of Hessen Protocol....................172 9.5.2. Hessen COMM Port Configuration ..................... 172 9.5.3.
Page 16
10.1.4. NO Conversion Efficiency (CE) ....................... 186 10.2. Manual Calibration Checks and Calibration of the T200 Analyzer in its Base Configuration ....187 10.2.1. Setup for Basic Calibration Checks and Calibration of the T200 analyzer........187 10.2.2. Performing a Basic Manual Calibration Check ................. 188 10.2.3.
Page 17
12.2.1.2. Sample Gas Flow Valves and Routing..................236 12.2.2. Flow Rate Control - Critical Flow Orifices ..................237 12.2.2.1. Critical Flow Orifice........................237 12.2.2.2. Locations and Descriptions of Critical Flow Orifices Inside the T200 ........238 12.2.3. Ozone Gas Generation and Air Flow....................239 12.2.3.1. The O Generator ........................
Page 18
Table of Contents Teledyne API – T200 NOx Analyzer Operation Manual 12.6. Pneumatic Sensor Board ......................... 262 12.7. Power Supply/Circuit Breaker ........................263 12.7.1. AC Power Switch/Circuit Breaker ..................... 264 12.7.2. AC Power Configuration ........................264 12.7.2.1. AC Configuration – Internal Pump (JP7) ................... 265 12.7.2.2.
Page 19
Teledyne API – T200 NOx Analyzer Operation Manual Table of Contents 14.5.3. Unstable Zero and Span ........................308 14.5.4. Inability to Span - No SPAN Button (CALS) ..................309 14.5.5. Inability to Zero - No ZERO Button (CALZ) ..................309 14.5.6.
Page 20
Energizing the T200 Control Inputs .....................45 Figure 3-11: Gas Line Connections from Calibrator – Basic T200 Configuration ..........50 Figure 3-12: Gas Line Connections from Bottled Span Gas – Basic T200 Configuration .......50 Figure 5-1: Current Loop Option Installed......................69 Figure 5-2: Concentration Alarm Relay ......................71...
Page 21
Set up for Manual Calibrations/Checks of T200’s in Base Configuration w/ a Gas Dilution Calibrator........................... 187 Figure 10-2: Set up for Manual Calibrations/Checks of T200’s in Base Configuration w/ Bottled Gas ..187 Figure 10-3: Pneumatic Connections for Precision Calibration of an T200 with the Internal Span Gas Generator ..........................
Voltage Tolerances for the TEST CHANNEL Calibration............140 Table 8-9: Current Loop Output Check ...................... 144 Table 8-10: Test Channels Functions available on the T200’s Analog Output..........147 Table 9-1: COMM Port Communication Modes ..................155 Table 9-2: Terminal Mode Software Commands ..................158 Table 9-3: Teledyne API's Serial I/O Command Types ................
Page 23
Analog Output Test Function - Nominal Values Voltage Outputs ..........316 Table 14-10: Status Outputs Check ....................... 318 Table 14-11: T200 Control Input Pin Assignments and Corresponding Signal I/O Functions....... 319 Table 15-1: Static Generation Voltages for Typical Activities ..............344 Table 15-2: Sensitivity of Electronic Devices to Damage by ESD ...............
Page 24
Table of Contents Teledyne API – T200 NOx Analyzer Operation Manual This page intentionally left blank. xxii 06858A DCN5847...
NO, which is then measured as above (including the original NO in the sample gas) and reported as NO Since the density of the sample gas effects the brightness of the chemiluminescence reaction, the T200 software compensates for temperature and pressure changes.
Page 28
Introduction Teledyne API – T200 NOx Analyzer Operation Manual This page intentionally left blank. 06858A DCN5847...
Provided here are the specifications, the EPA Equivalency Designation, and the emissions and safety compliance approvals for this instrument. 2.1. SPECIFICATIONS Table 2-1 presents the instrument’s parameters and the specifications that each meets. Table 2-1: T200 Basic Unit Specifications Parameter Specification Min/Max Range...
Defined as twice the zero noise level by the US EPA. 2.2. EPA EQUIVALENCY DESIGNATION Teledyne API’s T200 nitrogen oxides analyzer is designated as a reference method for NO measurement, as defined in 40 CFR Part 53, when operated under the following conditions: ...
2.3. CE MARK COMPLIANCE EMISSIONS COMPLIANCE The Teledyne API T200 nitrogen oxides analyzer was tested and found to be fully compliant with: EN61326 (1997 w/A1: 98) Class A, FCC Part 15 Subpart B Section 15.107 Class A, ICES-003 Class A (ANSI C63.4 1992) &...
Page 32
Specifications and Approvals Teledyne API – T200 NOx Analyzer Operation Manual This page intentionally left blank. 06858A DCN5847...
GETTING STARTED This section first introduces the instrument, then presents the procedures for getting started, i.e., unpacking and inspection, making electrical and pneumatic connections, and conducting an initial calibration check. 3.1. FRONT PANEL Figure 3-1 shows the analyzer’s front panel layout, followed by a close-up of the display screen in Figure 3-2, which is described in Table 3-1.
Getting Started Teledyne API – T200 NOx Analyzer Operation Manual Figure 3-2: Display Screen and Touch Control CAUTION – Avoid Damaging Touchscreen Do not use hard-surfaced instruments such as pens to touch the control buttons. The front panel liquid crystal display screen includes touch control. Upon analyzer start-up, the screen shows a splash screen and other initialization indicators before the main display appears, similar to Figure 3-2 above.
Teledyne API – T200 NOx Analyzer Operation Manual Getting Started Figure 3-3 shows how the front panel display is mapped to the menu charts that are used to illustrate the display throughout this manual. The Mode, Param (parameters), and Conc (gas concentration) fields in the display screen are represented across the top row of each menu chart.
Getting Started Teledyne API – T200 NOx Analyzer Operation Manual 3.2. REAR PANEL Figure 3-4: Rear Panel Layout – Base Unit Table 3-2 provides a description of each component on the rear panel. 06858A DCN5847...
Page 37
Teledyne API – T200 NOx Analyzer Operation Manual Getting Started Table 3-2: Rear Panel Description Component Function cooling fan Pulls ambient air into chassis through side vents and exhausts through rear. Connector for three-prong cord to apply AC power to the analyzer.
Teledyne API – T200 NOx Analyzer Operation Manual Getting Started Figure 3-6: Pneumatic Diagram – Basic T200 NOTE For pneumatic diagrams of Model T200’s with various calibration valve options see Section 5. 06858A DCN5847...
ENERAL AFETY AZARD To avoid personal injury, always use two persons to lift and carry the T200. CAUTION – Avoid Warranty Invalidation Printed circuit assemblies (PCAs) are sensitive to electro-static discharges too small to be felt by the human nervous system. Damage resulting from failure to use ESD protection when working with electronic assemblies will void the instrument warranty.
Page 41
Teledyne API – T200 NOx Analyzer Operation Manual Getting Started 5. Inspect the interior of the instrument to ensure all circuit boards and other components are in good shape and properly seated. 6. Check the connectors of the various internal wiring harnesses and pneumatic hoses to ensure they are firmly and properly seated.
Getting Started Teledyne API – T200 NOx Analyzer Operation Manual 3.4.1. VENTILATION CLEARANCE Whether the analyzer is set up on a bench or installed into an instrument rack, be sure to leave sufficient ventilation clearance. Table 3-3: Ventilation Clearance MINIMUM REQUIRED...
Page 43
Teledyne API – T200 NOx Analyzer Operation Manual Getting Started CAUTION ENERAL AFETY AZARD The T200 analyzer can be configured for both 100-130 V and 210-240 V at either 47 or 63 Hz. To avoid damage to your analyzer, ensure that the AC power voltage matches the voltage indicated on the analyzer’s model/specs label (Figure 3-4) before plugging...
See Section 8.1 for details on setting up the DAS. 3.5.3. ANALOG OUTPUT CONNECTIONS The T200 is equipped with several analog output channels accessible through a connector on the back panel of the instrument. Output channels A1, A2 and A3 are assigned to the NO , NO and NO concentration signals of the analyzer.
Teledyne API – T200 NOx Analyzer Operation Manual Getting Started To access these signals attach a strip chart recorder and/or data-logger to the appropriate analog output connections on the rear panel of the analyzer. Pin-outs for the analog output connector are:...
Getting Started Teledyne API – T200 NOx Analyzer Operation Manual 3.5.4. CONNECTING THE STATUS OUTPUTS The status outputs report analyzer conditions via optically isolated NPN transistors, which sink up to 50 mA of DC current. These outputs can be used interface with devices that accept logic-level digital inputs, such as Programmable Logic Controllers (PLCs).
Teledyne API – T200 NOx Analyzer Operation Manual Getting Started 3.5.5. CONNECTING THE CONTROL INPUTS The analyzer is equipped with three digital control inputs that can be used to activate the zero and span calibration modes remotely (see Section 10.1.2.4).
COM2 port is not in use except for RS-232 Multidrop communication. 3.5.9. CONNECTING TO A MULTIDROP NETWORK If your unit has a Teledyne API RS-232 multidrop card, see Section 5.6.3 for a description and Section 9.2 for instructions on setting it up.
If your analyzer is equipped with an external zero air scrubber option, it is capable of creating zero air from ambient air. For analyzers without the external zero air scrubber, a zero air generator such as the Teledyne API’s Model 701 can be used. Please visit the company website for more information. 3.6.1.2. Calibration (Span) Gas Calibration gas is a gas specifically mixed to match the chemical composition of the type of gas being measured at near full scale of the desired reporting range.
100 ppm NOTE The NO permeation tube included with the T200’s optional Internal Zero Air generator (IZS) has a limited accuracy of about ±5%. While NO permeation tubes may be sufficient for informal calibration checks, they are not approved by the US EPA as calibration sources for performing actual calibration of the analyzer.
If a dynamic dilution system such as the Teledyne API Model 700 is used to dilute high concentration gas standards to low, ambient concentrations, ensure that the NO concentration of the reference gas matches the dilution range of the calibrator.
Teledyne API – T200 NOx Analyzer Operation Manual 3.6.2. BASIC PNEUMATIC SETUP FOR THE T200 ANALYZER Figure 3-11: Gas Line Connections from Calibrator – Basic T200 Configuration Figure 3-12: Gas Line Connections from Bottled Span Gas – Basic T200 Configuration 06858A DCN5847...
Page 53
In applications where the sample gas is received from a pressurized manifold and the analyzer is not equipped with one of the T200’s pressurized span options, a vent must be placed on the sample gas line. This vent line must be: ...
Customer Service as soon as possible: 800-324-5190 or email: api-customerservice@teledyne.com If you are unfamiliar with the T200 theory of operation, we recommend that you read Section 12. For information on navigating the analyzer’s software menus, see the menu trees described in Appendix A.1.
Teledyne API – T200 NOx Analyzer Operation Manual Getting Started 3.7.2. WARNING MESSAGES Because internal temperatures and other conditions may be outside the specified limits during the analyzer’s warm-up period, the software will suppress most warning conditions for 30 minutes after power up. If warning messages persist after the 30 minutes warm up period is over, investigate their cause using the troubleshooting guidelines in Section 14.1.
The A/D or at least one D/A channel have not been calibrated. BOX TEMP WARNING The temperature inside the T200 chassis is outside the specified limits. CANNOT DYN SPAN Contact closure span calibration failed while DYN_SPAN was set to ON.
Teledyne API – T200 NOx Analyzer Operation Manual Getting Started 3.7.3. FUNCTIONAL CHECK After the analyzer’s components have warmed up for at least 30 minutes, verify that the software properly supports any hardware options that are installed. For information on navigating through the analyzer’s software menus, see the menu trees described in Appendix A.1.
See Section 10.4 for information regarding setup and calibration of T200 analyzers with Z/S Valve options. If you are using the T200 analyzer for EPA monitoring, only the calibration method described in Section 11 should be used. 3.8.1. INTERFERENTS FOR NOX, NO AND NO2 MEASUREMENTS...
Unit of Measure: PPB Reporting Range: 500 ppb Mode Setting: SNGL While these are the default setting for the T200 analyzer, it is recommended that you verify them before proceeding with the calibration procedure, by pressing: 06858A DCN5847...
Getting Started Teledyne API – T200 NOx Analyzer Operation Manual 3.8.2.2. Verify the Expected NO and NO Span Gas Concentration NOTE For this initial calibration, it is important to verify the PRECISE Concentration Value of the SPAN gases independently. If you supply NO gas to the analyzer, the values for expected NO and NO MUST be identical.
Teledyne API – T200 NOx Analyzer Operation Manual Getting Started 3.8.2.3. Initial Zero/Span Calibration Procedure To perform an initial Calibration of the T200 nitrogen oxides analyzer, press: The T200 Analyzer is now ready for operation. NOTE Once you have completed the above set-up procedures, please fill out the Quality Questionnaire that was shipped with your unit and return it to Teledyne API.
Page 62
Getting Started Teledyne API – T200 NOx Analyzer Operation Manual This page intentionally left blank. 06858A DCN5847...
Q: Why is the ZERO or SPAN button not displayed during calibration? A: The T200 disables certain these buttons expected span or zero value entered by the users is too different from the gas concentration actually measured value at the time. This is to prevent the accidental recalibration of the analyzer to an out-of-range response curve.
Page 64
Frequently Asked Questions and Glossary Teledyne API – T200 NOx Analyzer Operation Manual Q: How long does the sample pump last? A: The sample pump should last one to two years and the pump head should be replaced when necessary.
10 megabits per second (Mbps) 100BaseT same as 10BaseT except ten times faster (100 Mbps) APICOM name of a remote control program offered by Teledyne-API to its customers ASSY Assembly Code-Activated Switch Corona Discharge, a frequently luminous discharge, at the surface of a...
Page 66
Frequently Asked Questions and Glossary Teledyne API – T200 NOx Analyzer Operation Manual Term Description/Definition Dry Filter Unit DHCP Dynamic Host Configuration Protocol. A protocol used by LAN or Internet servers to automatically set up the interface protocols between themselves and...
Page 67
Teledyne API – T200 NOx Analyzer Operation Manual Frequently Asked Questions and Glossary Term Description/Definition Light Emitting Diode Liters Per Minute Mass Flow Controller Measure/Reference NDIR Non-Dispersive Infrared MOLAR MASS the mass, expressed in grams, of 1 mole of a specific substance. Conversely, one mole is the amount of the substance needed for the molar mass to be the same number in grams as the atomic mass of that substance.
Page 68
Frequently Asked Questions and Glossary Teledyne API – T200 NOx Analyzer Operation Manual Term Description/Definition RS-485 specification and standard describing a binary serial communication method among multiple devices at a data rate faster than RS-232 with a much longer distance between the host and the furthest device...
5.1. OPTIONAL PUMPS (OPT 11A, OPT 11B, OPT 12A, OPT 12B & OPT 12C) A variety of external pumps are available for the T200 analyzer. The range of available pump options meets all typical AC power supply standards while exhibiting the same pneumatic performance.
5.3. CARRYING STRAP/HANDLE (OPT 29) The chassis of the T200 analyzer allows the user to attach a strap handle for carrying the instrument. The handle is located on the right side and pulls out to accommodate a hand for transport. When pushed in, the handle is nearly flush with the chassis, only protruding out about 9 mm (3/8”).
Teledyne API – T200 NOx Analyzer Operation Manual Non-Gas Related Optional Hardware and Software Figure 5-1: Current Loop Option Installed 5.4.1. CONVERTING CURRENT LOOP ANALOG OUTPUTS TO STANDARD VOLTAGE OUTPUTS NOTE See Section 15 for more information on preventing ESD damage.
5.5. SPARE PARTS KITS 5.5.1. T200 EXPENDABLES KIT (OPT 42A) This kit includes a recommended set of expendables for one year of operation of the T200 including replacement sample particulate filters. See Appendix B for a detailed listing of the contents.
5.6.2. CONCENTRATION ALARM RELAY (OPTION 61) The Teledyne API “E” series analyzers have an option for four (4) “dry contact” relays on the rear panel of the instrument. This relay option is different from and in addition to the “Contact Closures” that come standard on all TAPI instruments.
In instruments that sample more than one gas type, there could be more than one gas type triggering the Concentration 1 Alarm (“Alarm 2” Relay). For example, the T200 instrument can monitor both NO & NO gas. The software for this instrument is flexible enough to allow you to configure the alarms so that you can have 2 alarm levels for each gas.
The dilution ratio feature is a software option that is designed for applications where the sample gas is diluted before being analyzed by the T200. Typically this occurs in continuous emission monitoring (CEM) applications where the quality of gas in a smoke stack is being tested and the sampling method used to remove the gas from the stack dilutes the gas.
Page 76
Non-Gas Related Optional Hardware and Software Teledyne API – T200 NOx Analyzer Operation Manual This page intentionally left blank. 06858A DCN5847...
These option packages include variations for handling pressurized and non-pressurized zero air and span gas supplied from external sources as well as internally generated span gas. Table 6-1: List of T200 Calibration Gas Options DESCRIPTION MANUAL...
Gas Inlet and Calibration Valve Options Teledyne API – T200 NOx Analyzer Operation Manual 6.1. AMBIENT ZERO/AMBIENT SPAN VALVES (OPT 50A) This valve package includes: Two solenoid valves located inside the analyzer that allow the user to switch either zero, span or sample gas to the instrument’s sensor.
This vent line must be no more than 10 meters long. CALIBRATION GAS SOURCES: SPAN GAS Attach a gas line from the source of calibration gas (e.g. a Teledyne API's M700E Dynamic Dilution Calibrator) to the SPAN1 inlet (see Figure 6-1). Use PTFE tubing; minimum O.D ¼”. ZERO AIR Zero air is supplied by the zero air generator such as a Teledyne API's M701.
Teledyne API – T200 NOx Analyzer Operation Manual Gas Inlet and Calibration Valve Options 6.2. AMBIENT ZERO/PRESSURIZED SPAN VALVES (OPT 50B) This calibration valve package is appropriate for applications where Span Gas is being supplied from a pressurized source such as bottled NIST SRM gases. This option includes: ...
Teledyne API – T200 NOx Analyzer Operation Manual Gas Inlet and Calibration Valve Options 6.2.2. PNEUMATIC SETUP (OPT 50B) Figure 6-6: Gas Line Connections for T200 with OPT 50B SAMPLE GAS SOURCE: Attach a sample inlet line to the SAMPLE inlet fitting.
Teledyne API – T200 NOx Analyzer Operation Manual 6.3. INTERNAL SPAN SOURCE - IZS (OPT 50G) The T200 nitrogen oxides analyzer can also be equipped with an internal NO span gas generator and calibration valve option. This option package is intended for applications where there is a need for frequent automated calibration checks without access to an external source of span gas.
(nitric acid) takes place whenever water and NO are present in the same gas mixture. In the T200 this is mitigated as much as possible by passing the air supply for the span gas generator through a special dryer, however the permeable membrane of the NO...
-20° C (~600 ppm H O) and effectively remove concentrations of ammonia ) up to about 1 ppm. An additional Sample Conditioner can be added to the T200’s sample gas stream. INSTRUMENT CHASSIS Sample Gas Dryer/Scrubber Orifice Dia.
An external zero air scrubber for Z/S valves can be used in place of a zero air generator The following pneumatic diagram illustrates the internal and external flow for a T200 analyzer with a Z/S valve option and the Zero Air...
Page 89
PART II – OPERATING INSTRUCTIONS 06858A DCN5847...
7.1. OVERVIEW OF OPERATING MODES The T200 analyzer software has a variety of operating modes. The most common mode that the analyzer will be operating in is the SAMPLE mode. In this mode, a continuous read-out of the NO concentrations can be viewed on the front panel and output as an analog voltage from rear panel terminals.
This information is particularly when troubleshooting a performance problem with the T200 (see Section 13). Figure 7-2 will display the Test Functions on the front panel screen. Table 7-2 lists the available TEST functions.
Teledyne API – T200 NOx Analyzer Operation Manual Basic Operation of the T200 Analyzer To view these TEST functions, press, Figure 7-2: Viewing T200 Test Functions NOTE A value of “XXXX” displayed for any of the TEST functions indicates an out-of-range reading or the analyzer’s inability to calculate it.
Basic Operation of the T200 Analyzer Teledyne API – T200 NOx Analyzer Operation Manual Table 7-2: Test Functions Defined DISPLAY PARAMETER UNITS DESCRIPTION The Full Scale limit at which the reporting range of the analyzer’s ANALOG OUTPUTS is currently set.
AZERO WARN in message indicates auto-zero reading at time warning was displayed. BOX TEMP WARNING The temperature inside the T200 chassis is outside the specified limits. CANNOT DYN SPAN Contact closure span calibration failed while DYN_SPAN was set to ON.
7.3. CALIBRATION MODE Pressing the CAL button, switches the T200 into calibration mode. In this mode the user can, in conjunction with introducing of zero or span gases of known concentrations into the analyzer, cause it to adjust and recalculate the slope (gain) and offset of the its measurement range.
Teledyne API – T200 NOx Analyzer Operation Manual Basic Operation of the T200 Analyzer 7.4. SETUP MODE The SETUP mode contains a variety of choices that are used to configure the analyzer’s hardware and software features, perform diagnostic procedures, gather information on the instruments performance and configure or access data from the internal data acquisition system (DAS).
Special instrument or software features or installed options may also be listed here. Use this information to identify the software and hardware installed in your T200 analyzer when contacting customer service. To access the configuration table, press: 06858A DCN5847...
Basic Operation of the T200 Analyzer 7.4.2. SETUP PASS: ENABLING/DISABLING PASSWORDS The T200 provides password protection of the calibration and setup functions to prevent unauthorized adjustments. When the passwords have been enabled in the PASS menu item, the system will prompt the user for a password anytime a password-protected function is requested.
Page 100
Basic Operation of the T200 Analyzer Teledyne API – T200 NOx Analyzer Operation Manual Example: If all passwords are enabled, the following touchscreen control sequence would be required to enter the VARS or DIAG submenus: NOTE The instrument still prompts for a password when entering the VARS and DIAG menus, even if passwords are disabled, but it displays the default password (818) upon entering these menus.
7.4.3.1. Setting the Internal Clock’s Time and Day The T200 has a time of day clock that supports the DURATION step of the automatic calibration (ACAL) sequence feature, time of day TEST function, and time stamps on for the DAS feature and most COMM port messages.
Basic Operation of the T200 Analyzer Teledyne API – T200 NOx Analyzer Operation Manual 7.4.3.2. Adjusting the internal Clock’s speed In order to compensate for CPU clocks which run faster or slower, you can adjust a variable called CLOCK_ADJ to speed up or slow down the clock by a fixed amount every day.
The T200 solves this problem by allowing the user to select a reporting range for the analog outputs that only includes that portion of the physical range that covers the specific application. This increases the reliability and accuracy of the analyzer by avoiding additional gain-amplification circuitry.
The A1, A2 and A3 channels output a signal that is proportional to the NO , NO and NO concentrations of the sample gas, respectively. The T200 can be set so that these outputs operate in one of the three following modes: ...
Basic Operation of the T200 Analyzer Teledyne API – T200 NOx Analyzer Operation Manual 7.4.4.5. RNGE MODE SNGL: Configuring the T200 Analyzer for Single Range Mode NOTE This is the default reporting range mode for the analyzer. When the single range mode is selected (SNGL), all analog NO...
Teledyne API – T200 NOx Analyzer Operation Manual Basic Operation of the T200 Analyzer 7.4.4.6. RNGE MODE IND: Configuring the T200 analyzer for Independent Range Mode The independent range mode (IND) assigns the three NO , NO and NO concentrations to individual analog output channels.
Page 108
Basic Operation of the T200 Analyzer Teledyne API – T200 NOx Analyzer Operation Manual To select the IND range mode, press the following buttons: SAMPLE RANGE=500.0 PPB NOX= XXXX <TST TST> CAL SETUP Concentration field displays all gases. SETUP X.X...
Page 109
Teledyne API – T200 NOx Analyzer Operation Manual Basic Operation of the T200 Analyzer To set the upper range limit for each independent reporting range, press: 06858A DCN5847...
Basic Operation of the T200 Analyzer Teledyne API – T200 NOx Analyzer Operation Manual 7.4.4.7. RNGE MODE AUTO: Configuring the T200 analyzer for Auto Range Mode In AUTO range mode, the analyzer automatically switches the reporting range between two user-defined ranges (LOW and HIGH).
Teledyne API – T200 NOx Analyzer Operation Manual Basic Operation of the T200 Analyzer 7.4.4.8. SETUP RNGE UNIT: Setting the Reporting Range Units of Measure The T200 can display and report concentrations in ppb, ppm, ug/m , mg/m units. Changing units affects all of the COMM port values, and all of the display values for all reporting ranges.
Basic Operation of the T200 Analyzer Teledyne API – T200 NOx Analyzer Operation Manual 7.4.5. RNGE DIL: USING THE OPTIONAL DILUTION RATIO FEATURE This feature is an optional software utility used to compensate for any dilution of the sample gas that may occur before it enters the SAMPLE inlet.
The DAS feature of the T200 can store up to one million data points, which can, depending on individual configurations, cover days, weeks or months of valuable measurements. The data records are stored in non-volatile memory and are retained even when the instrument is off.
Advanced Features of the T200 analyzer Teledyne API – T200 NOx Analyzer Operation Manual 8.1.2. DAS STRUCTURE The DAS is designed around the feature of a “record”. A record is a single data point. The type of data recorded in a record is defined by two properties: ...
The last 1100 daily averages are stored to cover more than four years of analyzer performance. HIRES: Records one-minute, instantaneous data of all active parameters in the T200. Short-term trends as well as signal noise levels can be detected and documented.
Advanced Features of the T200 analyzer Teledyne API – T200 NOx Analyzer Operation Manual List of Channels List of Parameters STORE NUM PARAMETER MODE PRECISION SAMPLES Name: NOXCNC1 Event: NOCNC1 Parameters: N2CNC1 O2CONC Report Period: STABIL No. of Records: RS-232 Report:...
The number of DAS objects are limited by the instrument’s finite storage capacity. For information regarding the maximum number of channels, parameters, and records and how to calculate the file size for each data channel, refer to the DAS manual downloadable from the T-API website at http://www.teledyne-api.com/manuals/ under Special Manuals.
Advanced Features of the T200 analyzer Teledyne API – T200 NOx Analyzer Operation Manual 8.1.5. SETUP DAS EDIT: ACCESSING THE DAS EDIT MODE DAS configuration is most conveniently done through the APICOM remote control program. The following list of button strokes shows how to edit using the front panel.
Teledyne API – T200 NOx Analyzer Operation Manual Advanced Features of the T200 analyzer 8.1.5.1. Editing DAS Data Channel Names To edit the name of an DAS data channel, follow the instruction shown in Section 8.1.5 then press: 06858A DCN5847...
Advanced Features of the T200 analyzer Teledyne API – T200 NOx Analyzer Operation Manual 8.1.5.2. Editing DAS Triggering Events Triggering events define when and how the DAS records a measurement of any given data channel. Triggering events are firmware-specific and a complete list of Triggers for this model analyzer can be found in Appendix A- 5.
Advanced Features of the T200 analyzer 8.1.5.3. Editing DAS Parameters Data parameters are types of data that may be measured and stored by the DAS. For each Teledyne API's analyzer model, the list of available data parameters is different, fully defined and not customizable. Appendix A-5 lists firmware specific data parameters for the T200.
Page 122
Advanced Features of the T200 analyzer Teledyne API – T200 NOx Analyzer Operation Manual To modify, add or delete a parameter, follow the instruction shown in Section 8.1.5 then press: Starting at the EDIT CHANNEL MENU DAS EDIT – Touchscreen Functions...
Teledyne API – T200 NOx Analyzer Operation Manual Advanced Features of the T200 analyzer 8.1.5.4. Editing Sample Period and Report Period The DAS defines two principal time periods by which sample readings are taken and permanently recorded: SAMPLE PERIOD: Determines how often DAS temporarily records a sample reading of the parameter in volatile memory.
Advanced Features of the T200 analyzer Teledyne API – T200 NOx Analyzer Operation Manual The SAMPLE PERIOD and REPORT PERIOD intervals are synchronized to the beginning and end of the appropriate interval of the instruments internal clock. If SAMPLE PERIOD is set for one minute the first reading would occur at the beginning of the next full minute according to the instrument’s internal clock.
Teledyne API – T200 NOx Analyzer Operation Manual Advanced Features of the T200 analyzer 8.1.5.6. Editing the Number of Records The number of data records in the DAS is limited to about a cumulative one million data points in all channels (one megabyte of space on the DOM).
Advanced Features of the T200 analyzer Teledyne API – T200 NOx Analyzer Operation Manual 8.1.5.7. RS-232 Report Function The DAS can automatically report data to the communications ports, where they can be captured with a terminal emulation program or simply viewed by the user using the APICOM software.
Teledyne API – T200 NOx Analyzer Operation Manual Advanced Features of the T200 analyzer 8.1.5.8. Enabling / Disabling the HOLDOFF Feature The DAS HOLDOFF feature prevents data collection during calibration operations. To enable or disable the HOLDOFF, follow the instruction shown in Section 8.1.5 then press: Starting at the EDIT CHANNEL MENU SETUP X.X...
Advanced Features of the T200 analyzer Teledyne API – T200 NOx Analyzer Operation Manual 8.1.5.9. The Compact Report Feature When enabled, this option avoids unnecessary line breaks on all RS-232 reports. Instead of reporting each parameter in one channel on a separate line, up to five parameters are reported in one line.
Refer to the APICOM manual for details on these procedures. The APICOM user manual (Teledyne API's P/N 039450000) is included in the APICOM installation file, which can be downloaded at http://www.teledyne-api.com/software/apicom/.
It is advised that you download and backup all data and the original DAS configuration before attempting any DAS changes. Refer to the next Section 9 for details on remote access to and from the T200 analyzer via the instrument’s COMM ports.
The following table lists all variables that are available within the 818 password protected level. See Appendix A2 for a detailed listing of all of the T200 variables that are accessible through the remote interface. Table 8-4: Variable Names (VARS)
Page 132
Advanced Features of the T200 analyzer Teledyne API – T200 NOx Analyzer Operation Manual To access and navigate the VARS menu, use the following button sequence: SAMPLE RANGE=500.0 PPB NOX= XXXX <TST TST> CAL SETUP Concentration field displays all gases.
8.2.1. Setting the Gas Type for the STABILITY Test Function One of the most common VARS used on the T200 is the stability test function ([Gas Type] STB). One of its primary uses is during calibration operations to judge when NO...
Advanced Features of the T200 analyzer Teledyne API – T200 NOx Analyzer Operation Manual 8.3. SETUP MORE DIAG :THE DIAGNOSTIC MENU A series of diagnostic tools is grouped together under the SETUPMOREDIAG menu. The parameters are dependent on firmware revision (see Appendix A). These tools can be used in a variety of troubleshooting and diagnostic procedures and are referred to in many places of the maintenance and troubleshooting sections of this manual.
Teledyne API – T200 NOx Analyzer Operation Manual Advanced Features of the T200 analyzer To access the various DIAG submenus, press the following buttons: Figure 8-4: Accessing the DIAG Submenus 06858A DCN5847...
Teledyne API – T200 NOx Analyzer Operation Manual 8.4. USING THE T200 ANALYZER’S ANALOG OUTPUTS The T200 analyzer comes equipped with four analog outputs. The first three outputs (A1 A2, & A3) carry analog signals that represent the currently measured concentrating of NO , NO and NO (see Section 7.4.4.3).
Teledyne API – T200 NOx Analyzer Operation Manual Advanced Features of the T200 analyzer To access the ANALOG I/O CONFIGURATION sub menu, press: Figure 8-5: Accessing the Analog I/O Configuration Submenus 06858A DCN5847...
Advanced Features of the T200 analyzer Teledyne API – T200 NOx Analyzer Operation Manual 8.4.2. ANALOG OUTPUT VOLTAGE / CURRENT RANGE SELECTION In its standard configuration the analog outputs is set to output a 0 – 5 VDC signals. Several other output ranges are available (see Table 8-7).
Teledyne API – T200 NOx Analyzer Operation Manual Advanced Features of the T200 analyzer 8.4.3. CALIBRATION OF THE ANALOG OUTPUTS Analog output calibration should be carried out on first startup of the analyzer (performed in the factory as part of the configuration process) or whenever recalibration is required.
Advanced Features of the T200 analyzer Teledyne API – T200 NOx Analyzer Operation Manual 8.4.3.2. Automatic Group Calibration of the Analog Outputs To calibrate the outputs as a group with the AOUTS CALIBRATION command, select the ANALOG I/O CONFIGURATION submenu (see Figure 8-5) then press: NOTE Before performing this procedure, ensure that the AUTO CAL for each analog output is enabled.
Teledyne API – T200 NOx Analyzer Operation Manual Advanced Features of the T200 analyzer 8.4.3.3. Automatic Individual Calibration of the Analog Outputs To use the AUTO CAL feature to initiate an automatic calibration for an individual analog output, select the...
Advanced Features of the T200 analyzer Teledyne API – T200 NOx Analyzer Operation Manual 8.4.3.4. Manual Calibration of the Analog Outputs Configured for Voltage Ranges For highest accuracy, the voltages of the analog outputs can be manually calibrated. NOTE The menu for manually adjusting the analog output signal level will only appear if the AUTO-CAL feature is turned off for the channel being adjusted (See Section 8.4.3.1).
Page 143
Teledyne API – T200 NOx Analyzer Operation Manual Advanced Features of the T200 analyzer To adjust the signal levels of an analog output channel manually, select the ANALOG I/O CONFIGURATION submenu (see Figure 8-5) then press: 06858A DCN5847...
Advanced Features of the T200 analyzer Teledyne API – T200 NOx Analyzer Operation Manual 8.4.3.5. Manual Adjustment of Current Loop Output Span and Offset A current loop option may be purchased for the A1, A2 and A3 Analog outputs of the analyzer. This option places circuitry in series with the output of the D-to-A converter on the motherboard that changes the normal DC voltage output to a 0-20 milliamp signal (See Section 5.4).
Page 145
Teledyne API – T200 NOx Analyzer Operation Manual Advanced Features of the T200 analyzer To adjust the zero and span signal levels of the current outputs, select the ANALOG I/O CONFIGURATION submenu (see Figure 8-5) then press: 06858A DCN5847...
Advanced Features of the T200 analyzer Teledyne API – T200 NOx Analyzer Operation Manual An alternative method for measuring the output of the Current Loop converter is to connect a 250 ohm 1% resistor across the current loop output in lieu of the current meter (see Figure 3-8 for pin assignments and diagram of the analog output connector).
8.4.4. TURNING AN ANALOG OUTPUT OVER-RANGE FEATURE ON/OFF In its default configuration, a ± 5% over-range is available on each of the T200’s analog outputs. This over- range can be disabled if your recording device is sensitive to excess voltage or current.
This can be achieved in the T200 by defining a zero offset, a small voltage (e.g., 10% of span). To add a zero offset to a specific analog output channel, select the ANALOG I/O CONFIGURATION submenu...
Advanced Features of the T200 analyzer 8.4.6. SELECTING A TEST CHANNEL FUNCTION FOR OUTPUT A4 The test functions available to be reported are listed in Table 8-10: Table 8-10: Test Channels Functions available on the T200’s Analog Output TEST CHANNEL DESCRIPTION...
Page 150
Advanced Features of the T200 analyzer Teledyne API – T200 NOx Analyzer Operation Manual To activate the TEST Channel and select a function, press: 06858A DCN5847...
Advanced Features of the T200 analyzer 8.4.7. AIN CALIBRATION This is the submenu to conduct a calibration of the T200 analyzer’s analog inputs. This calibration should only be necessary after major repair such as a replacement of CPU, motherboard or power supplies.
Advanced Features of the T200 analyzer Teledyne API – T200 NOx Analyzer Operation Manual 8.4.8. EXTERNAL ANALOG INPUTS (XIN1…XIN8) OPTION CONFIGURATION To configure the analyzer’s optional external analog inputs, define for each channel: gain (number of units represented by 1 volt) ...
9.1. USING THE ANALYZER’S COMMUNICATION PORTS The T200 is equipped with an Ethernet port, a USB port and two serial communication ports accessible via two DB-9 connectors on the back panel of the instrument. The COM1 connector (labeled RS232) is a male DB-9 connector and the COM2 is a female DB9 connector.
Remote Operation Teledyne API – T200 NOx Analyzer Operation Manual Data Bits: 8 data bits with 1 stop bit. Parity: None. Figure 9-1: Default Pin Assignments for Back Panel COMM Port connectors (RS-232 DCE & DTE) The signals from these two connectors are routed from the motherboard via a wiring harness to two 10-pin connectors on the CPU card, J11 and J12 (Figure 9-2).
NOTE Cables that appear to be compatible because of matching connectors may incorporate internal wiring that makes the link inoperable. Check cables acquired from sources other than Teledyne API for pin assignments before using. To assist in properly connecting the serial ports to either a computer or a modem, there are activity indicators just above the RS-232 port.
Remote Operation Teledyne API – T200 NOx Analyzer Operation Manual 9.1.4. COMM PORT BAUD RATE To select the baud rate of either one of the COMM Ports (use COM2 to view/configure USB port), press: SAMPLE RANGE=500.0 PPB NOX= XXXX <TST TST>...
Modes are listed in the order in which they appear in the SETUP MORE COMM COM[1 OR 2] MODE menu The default setting for this feature is ON. Do not disable unless instructed to by Teledyne API's Customer Service personnel.
Remote Operation Teledyne API – T200 NOx Analyzer Operation Manual 9.1.6. COMM PORT TESTING The serial ports can be tested for correct connection and output in the COMM menu. This test sends a string of 256 ‘w’ characters to the selected COMM port. While the test is running, the red LED on the rear panel of the analyzer should flicker.
Remote Operation 9.1.7. MACHINE ID Each type of Teledyne API's analyzer is configured with a default ID code. The default ID code for all T200 analyzers is 200. The ID number is only important if more than one analyzer is connected to the same communications channel such as when several analyzers are: ...
Teledyne API – T200 NOx Analyzer Operation Manual 9.1.8. TERMINAL OPERATING MODES The T200 can be remotely configured, calibrated or queried for stored data through the serial ports. As terminals and computers use different communication schemes, the analyzer supports two communicate modes specifically designed to interface with these two types of devices.
Teledyne API – T200 NOx Analyzer Operation Manual Remote Operation 9.1.8.2. Command Syntax Commands are not case-sensitive and all arguments within one command (i.e. ID numbers, key words, data values, etc.) must be separated with a space character. All Commands follow the syntax: X [ID] COMMAND <CR>...
9.1.8.5. COMM Port Password Security In order to provide security for remote access of the T200, a LOGON feature can be enabled to require a password before the instrument will accept commands. This is done by turning on the SECURITY MODE (Mode 4, Section 9.1.5).
9.1.9. REMOTE ACCESS BY MODEM The T200 can be connected to a modem for remote access. This requires a cable between the analyzer’s COMM port and the modem, typically a DB-9F to DB-25M cable (available from Teledyne API with P/N WR0000024).
Page 164
Remote Operation Teledyne API – T200 NOx Analyzer Operation Manual To initialize the modem press: SAMPLE RANGE=500.0 PPB NOX= XXXX <TST TST> CAL SETUP Concentration field displays all gases. SETUP X.X PRIMARY SETUP MENU CFG DAS RNGE PASS CLK MORE EXIT SETUP X.X...
Teledyne API – T200 NOx Analyzer Operation Manual Remote Operation 9.2. MULTIDROP RS-232 SET UP When the RS-232 Multidrop option is installed, the instrument designated as last in the chain must be terminated. This requires installing a shunt between two pins on the multidrop printed circuit assembly (PCA) inside the instrument.
4-digit identification number, press the button of the digit to be changed). NOTE Teledyne API recommends setting up the first link, between the Host and the first instrument and testing it before setting up the rest of the chain.
9.4.1. CONFIGURING THE ETHERNET INTERFACE USING DHCP The Ethernet for your T200 uses Dynamic Host Configuration Protocol (DHCP) to configure its interface with your LAN automatically. This requires your network servers also to be running DHCP. The analyzer will do this the first time you turn the instrument on after it has been physically connected to your network.
HOST NAME addressed from other computers on the LAN or via the Internet. To change, see Section 9.4.3. Do not change the setting for this property unless instructed to by Teledyne API’s Customer Service personnel. NOTE If the gateway IP, instrument IP and the subnet mask are all zeroes (e.g. “0.0.0.0”), the DCHP was not successful in which case you may have to configure the analyzer’s Ethernet properties manually.
Page 170
Remote Operation Teledyne API – T200 NOx Analyzer Operation Manual To view the above properties listed in Table 9-5, press: 06858A DCN5847...
Teledyne API – T200 NOx Analyzer Operation Manual Remote Operation 9.4.2. MANUALLY CONFIGURING THE ETHERNET (STATIC IP ADDRESS) To configure Ethernet communication manually: 1. Connect a cable from the analyzer’s Ethernet port to a Local Area Network (LAN) or Internet port.
Page 172
Remote Operation Teledyne API – T200 NOx Analyzer Operation Manual Next, refer to Section 9.4 for the Ethernet configuration settings and configure the INSTRUMENT IP, GATEWAY IP and SUBNET MASK addresses by pressing: 06858A DCN5847...
The HOSTNAME is the name by which the analyzer appears on your network. The initial default Hostname is blank. To change this name (particularly if you have more than one T200 analyzer on your network, where each must have a different Hostname), press:...
API's web site: http://www.teledyne-api.com/manuals/index.asp . 9.5.2. HESSEN COMM PORT CONFIGURATION Hessen protocol requires the communication parameters of the T200’s COMM ports to be set differently than the standard configuration as shown in the table below. Table 9-6: RS-232 Communication Parameters for Hessen Protocol...
9.5.3. ACTIVATING HESSEN PROTOCOL Once the COMM port has been properly configured, the next step in configuring the T200 in order to operate over a Hessen protocol network is to activate the Hessen mode for COMM ports and configure the communication parameters for the port(s) appropriately.
TYPE 2 that has more flexibility when operating with instruments that can measure more than one type of gas. For more specific information about the difference between TYPE 1 and TYPE 2 download the Manual Addendum for Hessen Protocol from the Teledyne API's web site: http://www.teledyne-api.com/manuals/.
Teledyne API – T200 NOx Analyzer Operation Manual Remote Operation 9.5.5. SETTING THE HESSEN PROTOCOL RESPONSE MODE The Teledyne API's implementation of Hessen Protocol allows the user to choose one of several different modes of response for the analyzer. Table 9-7: Teledyne API's Hessen Protocol Response Modes...
Teledyne API – T200 NOx Analyzer Operation Manual 9.5.6. HESSEN PROTOCOL GAS LIST ENTRIES 9.5.6.1. Gas List Entry Format and Definitions The T200 analyzer keeps a list of available gas types. Each entry in this list is of the following format. [GAS TYPE],[RANGE],[GAS ID],[REPORTED] WHERE: GAS TYPE = The type of gas to be reported (e.g.
Teledyne API – T200 NOx Analyzer Operation Manual Remote Operation 9.5.6.2. Editing or Adding HESSEN Gas List Entries To add or edit an entry to the Hessen Gas List, press: 06858A DCN5847...
Remote Operation Teledyne API – T200 NOx Analyzer Operation Manual 9.5.6.3. Deleting HESSEN Gas List Entries To delete an entry from the Hessen Gas list, press: 06858A DCN5847...
9.5.7. SETTING HESSEN PROTOCOL STATUS FLAGS Teledyne API's’ implementation of Hessen protocols includes a set of status bits that the instrument includes in responses to inform the host computer of its condition. Each bit can be assigned to one operational and warning message flag.
Each instrument on a Hessen Protocol network must have a unique ID code. If more than one T200 analyzer is on the Hessen network, you will have to change this code for all but one of the T200 analyzer’s on the Hessen network (see Section 9.1.7).
Teledyne API's’ main line of ambient and stack-gas instruments from a remote connection through direct cable, modem or Ethernet. Running APICOM, a user can: Establish a link from a remote location to the T200 through direct cable connection via RS-232 modem or Ethernet. ...
Page 184
Remote Operation Teledyne API – T200 NOx Analyzer Operation Manual This page intentionally left blank. 06858A DCN5847...
T200 CALIBRATION PROCEDURES This section contains a variety of information regarding the various methods for calibrating a T200 NO Analyzer as well as other supporting information. For information on EPA protocol calibration, please refer to Section 11. This section is organized as follows: SECTION 10.1 –...
Teledyne API – T200 NOx Analyzer Operation Manual NOTE Throughout this Section are various diagrams showing pneumatic connections between the T200 and various other pieces of equipment such as calibrators and zero air sources. These diagrams are only intended to be schematic representations of these connections and do not reflect actual physical locations of equipment and fitting location or orientation.
Teledyne API – T200 NOx Analyzer Operation Manual T200 Calibration Procedures For analyzers without the external zero air scrubber, a zero air generator such as the Teledyne API's M701 can be used. Please visit the company website for more information.
Data recording devices should be capable of bi-polar operation so that negative readings can be recorded. For electronic data recording, the T200 provides an internal data acquisition system (DAS), which is described in detail in Section 8.1. APICOM, a remote control program, is also provided as a convenient and powerful tool for data handling, download, storage, quick check and plotting (see Section 9.6).
CALIBRATION OF THE T200 ANALYZER. Connect the sources of zero air and span gas as shown below in one of the following ways:. Figure 10-1: Set up for Manual Calibrations/Checks of T200’s in Base Configuration w/ a Gas Dilution Calibrator Figure 10-2: Set up for Manual Calibrations/Checks of T200’s in Base Configuration w/ Bottled Gas...
T200 Calibration Procedures Teledyne API – T200 NOx Analyzer Operation Manual 10.2.2. PERFORMING A BASIC MANUAL CALIBRATION CHECK NOTE If the ZERO or SPAN buttons are not displayed, the measurement made is out of the allowable range allowed for a reliable calibration.
T200 Calibration Procedures 10.2.3. PERFORMING A BASIC MANUAL CALIBRATION The following section describes the basic method for manually calibrating the T200 NO analyzer. If the analyzer’s reporting range is set for the AUTO range mode, a step will appear for selecting which range is to be calibrated (LOW or HIGH).
T200 Calibration Procedures Teledyne API – T200 NOx Analyzer Operation Manual 10.2.3.2. Zero/Span Point Calibration Procedure NOTE If the ZERO or SPAN buttons are not displayed, the measurement made during is out of the allowable range allowed for a reliable calibration. See Section 13 for troubleshooting tips.
INTERNAL SPAN GAS GENERATOR While the T200 can be calibrated using the internal span gas generators permeation tube as a span gas source, it is still necessary to perform a precision calibration of the instrument using more accurate zero and span gas standards that the internal generator can provide.
Generator 10.3.3. CAL ON NO FEATURE When using the IZS option to calibrate the T200, the analyzer’s CAL_ON_NO feature must be turned on. This feature enables a continuous zero gas flow across the IZS permeation tube and through the NO converter.
Page 195
Teledyne API – T200 NOx Analyzer Operation Manual T200 Calibration Procedures TO turn the CAL_ON_NO feature ON/OFF, press, SAMPLE RANGE=500.0 PPB NO=XXXX <TST TST> SETUP SETUP X.X PRIMARY SETUP MENU CFG DAS RNGE PASS CLK MORE EXIT SETUP X.X SECONDARY SETUP MENU...
T200 Calibration Procedures Teledyne API – T200 NOx Analyzer Operation Manual 10.3.4. PERFORMING A MANUAL CALIBRATION CHECK WITH THE INTERNAL SPAN GAS GENERATOR SAMPLE RANGE=500.0 PPB NOX= XXXX Set the Display to show <TST TST> CAL CALZ CALS SETUP the NOX STB test function.
Teledyne API – T200 NOx Analyzer Operation Manual T200 Calibration Procedures 10.3.5. PERFORMING A MANUAL CALIBRATION WITH THE INTERNAL SPAN GAS GENERATOR If the analyzer’s reporting range is set for the AUTO range mode, a step will appear for selecting which range is to be calibrated (LOW or HIGH).
T200 Calibration Procedures Teledyne API – T200 NOx Analyzer Operation Manual 10.3.5.2. Zero/Span Point Calibration Procedure with Internal Span Gas Generator NOTE If the ZERO or SPAN buttons are not displayed, the measurement made during is out of the allowable range allowed for a reliable calibration.
10.4. MANUAL CALIBRATION AND CAL CHECKS WITH THE VALVE OPTIONS INSTALLED There are a variety of valve options available on the T200 for handling calibration gases (see Section 6 for descriptions of each). Generally performing calibration checks and zero/span point calibrations on analyzers with these options installed is similar to the methods discussed in the previous sections of this Section.
T200 Calibration Procedures Teledyne API – T200 NOx Analyzer Operation Manual 10.4.2. MANUAL CALIBRATION CHECKS WITH VALVE OPTIONS INSTALLED SAMPLE Set the Display to show <TST CAL CALZ CALS SETUP test function. Analyzer display This function calculates continues to cycle...
T200 Calibration Procedures 10.4.3. MANUAL CALIBRATION USING VALVE OPTIONS The following section describes the basic method for manually calibrating the T200 NO analyzer. If the analyzer’s reporting range is set for the AUTO range mode, a step will appear for selecting which range is to be calibrated (LOW or HIGH).
T200 Calibration Procedures Teledyne API – T200 NOx Analyzer Operation Manual 10.4.3.2. Zero/Span Point Calibration Procedure NOTE If the ZERO or SPAN buttons are not displayed, the measurement made during is out of the allowable range allowed for a reliable calibration. See Section 13 for troubleshooting tips.
If contact closures are being used in conjunction with the analyzer’s AutoCal (see Section 10.5) feature and the AutoCal attribute “CALIBRATE” is enabled, the T200 will not re-calibrate the analyzer UNTIL when the contact is opened. At this point, the new calibration values will be recorded before the instrument returns to SAMPLE mode.
T200 Calibration Procedures Teledyne API – T200 NOx Analyzer Operation Manual For each mode, there are seven parameters that control operational details of the SEQUENCE. They are: Table 10-3: AutoCal Attribute Setup Parameters ATTRIBUTE ACTION TIMER ENABLED Turns on the Sequence timer.
Teledyne API – T200 NOx Analyzer Operation Manual T200 Calibration Procedures The following example sets sequence #2 to do a zero-span calibration every other day starting at 1:00 AM on September 4, 2008, lasting 15 minutes, without calibration. This will start ½ hour later each iteration.
T200 Calibration Procedures Teledyne API – T200 NOx Analyzer Operation Manual 10.5.1. SETUP ACAL: PROGRAMMING AND AUTO CAL SEQUENCE To program the example sequence shown in Table 10-4, press: SAMPLE RANGE = 500.0 PPB NOX=XXX.X < TST TST > CAL CALZ CZLS SETUP SETUP X.X...
Page 207
Teledyne API – T200 NOx Analyzer Operation Manual T200 Calibration Procedures CONTINUED FROM PREVIOUS PAGE - STARTING DATE SETUP X.X STARTING DATE: 04–SEP–03 <SET SET> EDIT EXIT SETUP X.X STARTING TIME:00:00 <SET SET> EDIT EXIT Toggle buttons to set time: SETUP X.X...
Page 208
T200 Calibration Procedures Teledyne API – T200 NOx Analyzer Operation Manual CONTINUED FROM PREVIOUS PAGE DELTA TIME SETUP DURATION:15.0 MINUTES <SET SET> EDIT EXIT Toggle buttons to set SETUP DURATION 15.0MINUTES duration for each iteration of the sequence: ENTR EXIT Set in Decimal minutes from 0.1 –...
Teledyne API – T200 NOx Analyzer Operation Manual T200 Calibration Procedures 10.6. CALIBRATION QUALITY ANALYSIS After completing one of the calibration procedures described above, it is important to evaluate the analyzer’s calibration SLOPE and OFFSET parameters. These values describe the linear response curve of the analyzer, separately for NO and NO .
Teledyne API – T200 NOx Analyzer Operation Manual 10.7. GAS FLOW CALIBRATION Rate of sample gas and O flow through the T200 is a key part of the NO , NO and NO concentration calculations. The FLOW CALIBRATION submenu located under the DIAG menu allows the calibration/ adjustment of these calculations.
11.1. T200 CALIBRATION – GENERAL GUIDELINES In general, calibration is the process of adjusting the gain and offset of the T200 against a standard with certified, traceable concentration. The reliability of data derived from the analyzer depends primarily upon its state of calibration.
EPA Protocol Calibration Teledyne API – T200 NOx Analyzer Operation Manual Calibrations should be carried out at the field monitoring site. The analyzer should be in operation for at least several hours (preferably overnight) before calibration so that it is fully warmed up and its operation has stabilized.
Either a strip chart recorder, data acquisition system, digital data acquisition system should be used to record the data from the T200 RS-232 port or analog outputs. If analog readings are being used, the response of that system should be checked against a NIST referenced voltage source or meter. Data recording device should be capable of bi-polar operation so that negative readings can be recorded.
These checks are also used in data reduction and system validation. A Level 1 Span check is used to document that the T200 is within control limits and must be conducted every 2 weeks. A Level 2 Span Check is to be conducted between the Level 1 Checks on a schedule to be determined by the user.
EPA Protocol Calibration 11.5. GAS PHASE TITRATION (GPT) 11.5.1. GPT PRINCIPLE OF OPERATION Gas phase titration (GPT) is recommended during calibration of the T200. Those using a NO permeation tube should refer to the CFR The principle of GPT is based on the rapid gas phase reaction between NO and O...
Page 216
EPA Protocol Calibration Teledyne API – T200 NOx Analyzer Operation Manual 4. Calculate the NO concentration [NO] needed to approximate 90% of the URL of the NO analyzer to be calibrated: Equation 11-3 [NO] = URL of analyzer (ppm) x 90/100 5.
Teledyne API – T200 NOx Analyzer Operation Manual EPA Protocol Calibration 11.5.2.1. Example GPT Calculation The following is an example calculation that can be used to determine whether an existing GPT calibrator will meet the required conditions for a specific calibration. For this example, it is assumed that only the...
It is also necessary to calibrate the NO channel. This can be done simultaneously with the NO calibration. During the calibration the T200 should be operating in its normal sampling mode, and the test atmosphere should pass through all filters, scrubbers, conditioners, and other components used during normal ambient sampling and as much of the ambient air inlet system as is practicable.
Page 219
Teledyne API – T200 NOx Analyzer Operation Manual EPA Protocol Calibration Figure 11-1: GPT Calibration System NOTE The analyzer should be calibrated on the same range used for monitoring. Ensure that the GPT calibration system can supply the range of concentrations at a sufficient flow over the whole range of concentrations that will be encountered during calibration.
EPA Protocol Calibration Teledyne API – T200 NOx Analyzer Operation Manual 11.6.2. ZERO CALIBRATION The zero point calibration procedure is identical to that described in Sections 10.1, 10.2, 10.4 or 10.5. 11.6.3. SPAN CALIBRATION 1. Adjust the NO concentration to approximately 80% of the URL of the NO channel.
NO from the NO concentration. Unlike analog instruments, this difference is calculated by the T200's internal computer software. It is extremely unlikely that the NO concentration will be in error. Therefore this procedure is a confirmation that the NO subtraction algorithm in the computer is operating correctly.
11.8.1. SUMMARY OF QUALITY ASSURANCE CHECKS The following items should be checked on a regularly scheduled basis to assure high quality data from the T200. See Table 11-3 for a summary of activities; also the QA Handbook should be checked for specific procedures.
Teledyne API – T200 NOx Analyzer Operation Manual EPA Protocol Calibration 11.8.2. SHORT CALIBRATION CHECKS A system of Level 1 and Level 2 zero/span checks (Table 11-2) is recommended. These checks must be . Level 1 zero conducted in accordance with the specific guidance given in Section 12 of the Q.A. Handbook and span checks must be conducted every two weeks.
EPA Protocol Calibration Teledyne API – T200 NOx Analyzer Operation Manual NO and NO precision checks should also be made if those data are being reported. Information from the check procedure is used to assess the precision of the monitoring data;...
Teledyne API – T200 NOx Analyzer Operation Manual EPA Protocol Calibration NOTE If the NO impurity in the NO cylinder, [NO is greater than the 1 ppm value allowed in the calibration procedure, check that the NO delivery system is not the source of contamination before discarding the NO standard.
Page 226
EPA Protocol Calibration Teledyne API – T200 NOx Analyzer Operation Manual This page intentionally left blank. 06858A DCN5847...
12.1.1. CHEMILUMINESCENCE CREATION IN THE T200 REACTION CELL The T200’s measures the amount of NO present in a gas by detecting the chemiluminescence which occurs when nitrogen oxide (NO) is exposed to ozone (O This reaction is a two-step process: ...
12.1.2. CHEMILUMINESCENCE DETECTION IN THE T200 REACTION CELL 12.1.2.1. The Photo Multiplier Tube (PMT) The T200 uses a special kind of vacuum tube, called a photo-multiplier tube (PMT), to detect the amount of light created by the NO and O reaction in the reaction cell.
Reaction Cell with PMT Tube and Optical Filter The narrowness of this band of sensitivity allows the T200 to ignore extraneous light and radiation that might interfere with the T200’s measurement. For instance, some oxides of sulfur can also be chemiluminescent...
Once the chamber is completely dark, the T200 records the output of the PMT and keeps a running average of these AZERO values. This average offset value is subtracted from the raw PMT readings while the instrument is measuring NO and NO to arrive at an Auto Zero corrected reading.
It should be noted that the chemiluminescence method is subject to interferences from a number of sources. The T200 has been successfully tested for its ability to reject interference from most of these sources. Table 12- 1 list the most common types of interferents that could affect the performance of your T200.
. 12.1.5.3. Light Leaks The T200 sensitivity curve includes a small portion of the visible light spectrum (see Figure 12-2), therefore it is important to ensure that the reaction cell is completely sealed with respect to light. To ensure this: ...
Teledyne API – T200 NOx Analyzer Operation Manual Theory of Operation Table 12-1: List of Interferents Interference Type Rejection Method Dilution: Viscosity of CO molecules causes them to If high concentrations of CO are suspected, collect in aperture of Critical Flow Orifice altering flow special calibration methods must be performed to rate of NO.
In this section of the manual vacuum readings are given in inches of mercury absolute (In-Hg-A). This pressure value is referenced against zero (a perfect vacuum). The gas flow for the T200 is created by a pump that is pneumatically downstream from the rest of the instrument’s components. This is either: ...
Page 237
Teledyne API – T200 NOx Analyzer Operation Manual Theory of Operation Figure 12-5: Internal Gas Flow for Basic T200 with External Pump Figure 12-6: Basic Internal Gas Flow for Basic T200 with Internal Pump 06858A DCN5847...
Teledyne API – T200 NOx Analyzer Operation Manual 12.2.1.1. Vacuum Manifold The vacuum created by the analyzer’s pump is supplied to all of the gas streams for the T200 analyzer through the vacuum manifold (also called the exhaust manifold). Figure 12-7: Vacuum Manifold, Standard Configuration Configurations will vary depending on the optional equipment that is installed.
12.2.2. FLOW RATE CONTROL - CRITICAL FLOW ORIFICES Sample gas flow in the T200 analyzer is created via the use of several flow control assemblies (see Figure 12-8 for an example) located in various places in the gas streams of the instrument. These assemblies consist of: ...
12.2.2.2. Locations and Descriptions of Critical Flow Orifices Inside the T200 The T200 uses several of the following critical flow orifices (Figure 12-9) to create and maintain the proper flow rate of gas through its various components. (Please note that not all features displayed in Figure 12-9 are standard components of T200 analyzers).
NOTE The diameter of a critical flow orifice may change with temperature because of expansion of the orifice material and, hence, the most crucial critical flow orifices in the T200 (those controlling the sample gas and O flow into the cell itself) are located in the reaction cell where they can be maintained at a constant temperature.
Theory of Operation Teledyne API – T200 NOx Analyzer Operation Manual Figure 12-10: Ozone Generator Principle 12.2.3.2. Ozone Generator Dry Air Supply Ambient air usually contains enough water vapor to greatly diminishes the yield of ozone produced by the ozone generator.
Page 243
Dryer To provide a dry purge gas for the outer side of the Nafion tube, the T200 returns some of the dried air from the inner tube to the outer tube. This means that any time the analyzer is turned on after having been OFF for 30 minutes or more, the humidity gradient between the inner and outer tubes is not very large and the dryer’s...
, the break-down is not quite fast enough to ensure that it is completely removed from the exhaust gas stream of the T200 by the time the gas exits the analyzer. Due to the high toxicity and reactivity of O , a highly efficient catalytic converter scrubs or converts all of the O from the gas exiting the reaction cell.
12.2.4.3. Sample Gas Flow Calculation Sample gas flow in the T200 analyzer is not a directly measured value, but is rather calculated based on the measured pressure differential across the sample gas critical flow orifice. Specifically, the upstream reading of the sample pressure sensor is compared to the downstream pressure reading of the vacuum pressure sensor and this differential is used, by the analyzer’s CPU, to derive the gas flow rate through the reaction cell.
Theory of Operation Teledyne API – T200 NOx Analyzer Operation Manual 12.3. ELECTRONIC OPERATION 12.3.1. OVERVIEW Figure 12-13 shows a block diagram of the major electronic components of the analyzer. ANALOG COM2 USB COM RS232 Ethernet Female port Male C Bus)
Page 247
Teledyne API. It communicates with the user as well as receives data from and issues commands to a variety of peripheral devices via a separate printed circuit assembly onto which the CPU is mounted: the motherboard.
The DOM is a 44-pin IDE flash drive with storage capacity to 128 MB. It is used to store the computer’s operating system, the Teledyne API firmware, and most of the operational data generated by the analyzer’s internal data acquisition system (DAS).
CPU. The A/D can be configured for several different input modes and ranges but in the T200 it is used in unipolar mode with a +5V full scale. The converter includes a 1% over and under-range. This allows signals from –0.05V to +5.05V to be fully converted.
Theory of Operation Teledyne API – T200 NOx Analyzer Operation Manual PMT TEMPERATURE: The PMT temperature is measured with a thermistor inside the PMT cold block. Its signal is amplified by the PMT temperature feedback circuit on the preamplifier board and is digitized and sent to the CPU where it is used to calculate the current temperature of the PMT.
Page 251
Teledyne API – T200 NOx Analyzer Operation Manual Theory of Operation the tube at a high constant temperature (necessary to ensure that the permeation rate of NO is constant). It is stored and reported as test function IZS TEMP. ...
In its standard configuration, the T200 comes with all four of these channels set up to output a DC voltage. However, 4-20mA current loop drivers can be purchased for the first three of these outputs, A1, A2 and A3.
Teledyne API – T200 NOx Analyzer Operation Manual Theory of Operation PMT PREAMPLIFIER RANGE CONTROL: The CPU uses this control switch the instrument between its LOW and HIGH physical ranges (see Section 7.4.4.1). The instrument can be forced into its HIGH physical range setting the SIGNAL I/O function PREAMP_RANGE_HI to ON.
HOCK AZARD Only those relays actually required by the configuration of the T200 are populated. A protective retainer plate is installed over the ac power relay to keep them securely seated in their sockets and prevent accidental contact with those sockets that are not populated see Figure 12-16).
Teledyne API – T200 NOx Analyzer Operation Manual Theory of Operation Retainer Mounting Screws AC Relay Retainer Plate Figure 12-16: Relay PCA P/N 045230100 with AC Relay Retainer in Place 12.3.4.1. Status LED’s Eight LED’s are located on the Analyzer’s relay PCA to show the current status on the various control functions performed by the relay PCA (see Figure 12-17).
These two valves are actuated with 12 V supplied from the relay board and under the control of the CPU through the I C bus. Additional valves sets also controlled by the CPU via the I C bus and the relay PCA can be included in the T200 (see Section 6 for descriptions of these valve sets). 06858A DCN5847...
CPU. All of the heaters used in the T200 are AC powered which are turned ON/OFF by AC Relays located on the relay PCA in response to commands issued by the CPU.
Teledyne API – T200 NOx Analyzer Operation Manual 12.3.4.5. Thermocouple Inputs and Configuration Jumper (JP5) Although the relay PCA supports two thermocouple inputs, the current T200 analyzers only utilize one. It is used NO converter. to sense the temperature of the NO ...
Figure 12-19: Thermocouple Configuration Jumper (JP5) Pin-Outs 12.4. SENSOR MODULE, REACTION CELL The T200 sensor assembly consists of several subassemblies, each with different tasks: The Photo Multiplier Tube (PMT) detects the intensity of the light from the chemiluminescence reaction between NO and O in the reaction cell.
Figure 12-20: T200 Sensor Module Assembly 12.5. PHOTO MULTIPLIER TUBE (PMT) The T200 uses a photo multiplier tube (PMT) to detect the amount of chemiluminescence created in the Reaction Cell. A typical PMT is a vacuum tube containing a variety of specially designed electrodes. Photons from the reaction are filtered by an optical high-pass filter, enter the PMT and strike a negatively charged photo cathode causing it to emit electrons.
(this is referred to as “dark noise”). The gain voltage of the PMT used in the T200 is usually set between 400 V and 800 V. This parameter is viewable through the front panel as test function HVPS (see Section 7.2.1).
Page 262
Theory of Operation Teledyne API – T200 NOx Analyzer Operation Manual Figure 12-22: PMT Preamp Block Diagram The PMT preamplifier PCA also operates two different tests used to calibrate and check the performance of the sensor module. The electrical test (ETEST) circuit generates a constant, electronic signal intended to simulate the output of the PMT (after conversion from current to voltage).
Figure 12-23: Typical Thermo-Electric Cooler In the case of the T200, the current flow is controlled by the TEC Control PCA which adjusts the amount of current applied to the TEC based on the temperature sensed by a thermistor embedded in the PMT’s cold block.
12.6. PNEUMATIC SENSOR BOARD The flow and pressure sensors of the T200 are located on a printed circuit assembly just behind the PMT sensor. Refer to Section 14.7.6.1 for a figure and on how to test this assembly. The signals of this board are supplied to the motherboard for further signal processing.
Teledyne API – T200 NOx Analyzer Operation Manual Theory of Operation 12.7. POWER SUPPLY/CIRCUIT BREAKER The analyzer operates on 100 VAC, 115 VAC or 230 VAC power at either 50 Hz or 60Hz. Individual instruments are set up at the factory to accept any combination of these five attributes.
A 6.75 amp circuit breaker is built into the ON/OFF switch. In case of a wiring fault or incorrect supply power, the circuit breaker will automatically turn off the analyzer. Under normal operation, the T200 draws about 1.5 A at 115 V and 2.0 A during start-up. WARNING LECTRICAL...
Connects pump pin 1 to 220 / 240VAC power line 3 to 8 A jumper between pins 5 and 10 may be present on the jumper plug assembly, but has no function on the Model T200. 110 VAC /115 VAC...
Theory of Operation Teledyne API – T200 NOx Analyzer Operation Manual 12.7.2.2. AC Configuration – Standard Heaters (JP2) Power configuration for the AC the standard heaters is set using Jumper set JP2 (see Figure 12-28 for the location of JP2).
Teledyne API – T200 NOx Analyzer Operation Manual Theory of Operation 12.7.2.3. AC Configuration – Heaters for Option Packages (JP6) The IZS valve option includes an AC heaters that maintain an optimum operating temperature for key components of those options. Jumper set JP6 is used to connect the heaters associated with those options to AC power.
Theory of Operation Teledyne API – T200 NOx Analyzer Operation Manual 12.8. FRONT PANEL TOUCHSCREEN/DISPLAY INTERFACE Users can input data and receive information directly through the front panel touchscreen display. The LCD display is controlled directly by the CPU board. The touchscreen is interfaced to the CPU by means of a touchscreen controller that connects to the CPU via the internal USB bus and emulates a computer mouse.
Analyzer has a high performance, VortexX86-based microcomputer running WINDOWS CE. Inside the WINDOWS CE shell, special software developed by Teledyne API interprets user commands via the various interfaces, performs procedures and tasks, stores data in the CPU’s various memory devices and calculates the concentration of the sample gas.
Unlike other analyzers that average the output signal over a fixed time period, the T200 averages over a defined number of samples, with samples being about 8 seconds apart (reflecting the switching time of 4 s each for NO and NO ).
Teledyne API – T200 NOx Analyzer Operation Manual Theory of Operation Note that, as RCEL TEMP, BOX TEMP, RCELL and SAMP pressure increase, the value of TP_FACTOR increases and, hence, the PMTDET value decreases. These adjustments are meant to counter-act changes in the concentrations caused by these parameters.
Page 274
Theory of Operation Teledyne API – T200 NOx Analyzer Operation Manual This page intentionally left blank. 06858A DCN5847...
MAINTENANCE SCHEDULE & PROCEDURES For the most part, the T200 analyzer is maintenance free. However it is recommended that a minimal number of simple procedures be performed regularly to ensure that the T200 continues to operate accurately and reliably over its lifetime. In general, the exterior can be wiped down with a lightly damp cloth; avoid spraying anything directly onto any part of the analyzer.
Teledyne API – T200 NOx Analyzer Operation Manual 13.2. MAINTENANCE SCHEDULE Table 13-2 shows a typical maintenance schedule for the T200. Please note that in certain environments (i.e. dusty, very high ambient pollutant levels) some maintenance procedures may need to be performed more often than shown.
Page 277
Table 13-2: T200 Maintenance Schedule DATE PERFORMED ITEM ACTION FREQ CHECK REQ’D TEST Review and Weekly functions evaluate Particulate Change Weekly filter particle filter Zero/span Evaluate Weekly check offset and slope Zero/span Zero and Every 3 months calibration span calibration...
Teledyne API – T200 NOx Analyzer Operation Manual 13.3. MAINTENANCE PROCEDURES The following procedures are to be performed periodically as part of the standard maintenance of the T200. 13.3.1. REPLACING THE SAMPLE PARTICULATE FILTER The particulate filter should be inspected often for signs of plugging or contamination.
Performing this procedure improperly or with incorrect tools creates a risk of causing a significant leak. 4. Take off the old filter element and replace it with a suitable equivalent (Teledyne API P/N FL-3). Figure 13-2: Particle Filter on O Supply Air Dryer 5.
Maintenance Schedule & Procedures Teledyne API – T200 NOx Analyzer Operation Manual 13.3.3. CHANGING THE OZONE FILTER CHEMICAL The O filter is located next to the O generator (see Figure 3-5) and cleans the O stream from solid and liquid contaminants that are created inside the O generator.
Page 281
AZARD Immediately wash your hands after contact. 8. Using a small powder funnel, fill the cartridge with about 10 g new silica gel desiccant (Teledyne API P/N CH43) so that it is level on both legs of the cartridge. ...
NOTE ONLY USE TELEDYNE API’S REFILL KITS! Teledyne API's refill kits have been properly conditioned to prevent a significant increase of the T200’s Auto Zero value which can cause large negative offsets on that may take 2-3 weeks to disappear.
Teledyne API – T200 NOx Analyzer Operation Manual Maintenance Schedule & Procedures 13.3.5. CHANGING THE PUMP DFU FILTER The exhaust air from the analyzer passes a small particle filter (Dry Filter Unit (DFU - filter), P/N FL3) before entering the pump. It should be replaced when: ...
, EVEN AT ROOM TEMPERATURE AND WILL CONTAMINATE THE ENTIRE INSTRUMENT. 13.3.7. CHANGING THE EXTERNAL ZERO AIR SCRUBBER (OPT 86C) The external zero air scrubber that is included with several of the T200’s optional calibration valve packages contains two chemicals: ...
Maintenance Schedule & Procedures Teledyne API – T200 NOx Analyzer Operation Manual Figure 13-4: Zero Air Scrubber Assembly 13.3.7.1. Changing the External Scrubber’s DFU Filter There is also a DFU filter on the inlet of the external zero air scrubber that is included in several of the optional calibration valve packages.
Teledyne API – T200 NOx Analyzer Operation Manual Maintenance Schedule & Procedures 13.3.8. CHANGING THE NO CONVERTER The NO converter is located in the center of the instrument, Figure 3-5 for the location, and Figure 13-5 for the assembly. The converter is designed for replacement of the cartridge only; the heater with built-in thermocouple is to be reused.
Page 288
Converter Assembly 10. Wrap the band heater around the new replacement converter cartridge and tighten the screws using a high-temperature anti-seize agent (Teledyne API P/N CH42) such as copper paste. Ensure to use proper alignment of the heater with respect to the converter tubes.
2. Disconnect the black 1/4" exhaust tube and the 1/8” sample and ozone air tubes from the reaction cell. Disconnect the heater/thermistor cable. 3. Remove two screws (Teledyne API P/N SN144) and two washers holding the reaction cell to the PMT housing and lift the cell and manifold out.
13.3.10. REPLACING CRITICAL FLOW ORIFICES There are several critical flow orifices installed in the T200 (see Figure 12-9 for a pneumatic location of each orifice). Despite the fact that these flow restrictors are protected by sintered stainless steel filters, they can, on occasion, clog up, particularly if the instrument is operated without sample filter or in an environment with very fine, sub-micron particle-size dust.
13.3.11. CHECKING FOR LIGHT LEAKS When re-assembled or operated improperly, the T200 can develop small gaps around the PMT, which let stray light from the analyzer surrounding into the PMT housing. To find such light leaks, follow the procedures below.
Maintenance Schedule & Procedures Teledyne API – T200 NOx Analyzer Operation Manual CAUTION LECTRICAL HOCK AZARD Take extra care not to touch any of the inside wiring with the metal cover or your body. Do not drop screws or tools into a running analyzer! 4.
13.3.12.2. Detailed Pressure Leak Check If a leak cannot be located by the above procedure, obtain a leak checker similar to Teledyne API’s P/N 01960, which contains a small pump, shut-off valve, and pressure gauge to create both over-pressure and vacuum.
Maintenance Schedule & Procedures Teledyne API – T200 NOx Analyzer Operation Manual 13.3.12.3. Performing a Sample Flow Check AUTION ECHNICAL NFORMATION Use a separate, calibrated flow meter capable of measuring flows between 0 and 1000 cm³/min to measure the gas flow rate though the analyzer. Do not use the built in flow measurement viewable from the front panel of the instrument.
Use common sense when operating inside a running analyzer. 14.1. GENERAL TROUBLESHOOTING The T200 Nitrogen oxide analyzer has been designed so that problems can be rapidly detected, evaluated and repaired. During operation, it continuously performs diagnostic tests and provides the ability to evaluate its key operating parameters without disturbing monitoring operations.
General Troubleshooting & Repair Teledyne API – T200 NOx Analyzer Operation Manual See Figure 3-5 or the general layout of components and sub-assemblies in the analyzer. See the wiring interconnect diagram and interconnect list in Appendix D. 14.1.1. FAULT DIAGNOSIS WITH WARNING MESSAGES The most common and/or serious instrument failures will result in a warning message being displayed on the front panel.
Page 297
Teledyne API – T200 NOx Analyzer Operation Manual General Troubleshooting & Repair To view or clear the various warning messages press: If a warning message persists after several attempts to clear it, the message Suppresses the may indicate a real problem and not an MSG CLR SETUP warning messages.
Page 298
General Troubleshooting & Repair Teledyne API – T200 NOx Analyzer Operation Manual Table 14-1: Front Panel Warning Messages WARNING FAULT CONDITION POSSIBLE CAUSES ● AZERO valve not working ● Valve control driver failed Auto-zero reading above 200 mV. ● Bad Relay PCA...
Page 299
Teledyne API – T200 NOx Analyzer Operation Manual General Troubleshooting & Repair WARNING FAULT CONDITION POSSIBLE CAUSES If Sample Pressure is < 15 in-HG: Sample Pressure is <15 in-Hg or > Blocked Particulate Filter 35 in-Hg Blocked Sample Inlet/Gas Line Normally 29.92 in-Hg at sea level...
General Troubleshooting & Repair Teledyne API – T200 NOx Analyzer Operation Manual 14.1.2. FAULT DIAGNOSIS WITH TEST FUNCTIONS Besides being useful as predictive diagnostic tools, the test functions viewable from the analyzers front panel can be used to isolate and identify many operational problems when combined with a thorough understanding of the analyzers Theory of Operation (see Section 12).
Teledyne API – T200 NOx Analyzer Operation Manual General Troubleshooting & Repair 14.1.3. DIAG SIGNAL I/O: USING THE DIAGNOSTIC SIGNAL I/O FUNCTION The signal I/O diagnostic mode allows access to the digital and analog I/O in the analyzer. Some of the digital signals can be controlled through the touchscreen.
Teledyne API – T200 NOx Analyzer Operation Manual 14.2. USING THE ANALOG OUTPUT TEST CHANNEL The signals available for output over the T200’s analog output channel can also be used as diagnostic tools. See Section 8.4 for instruction on activating the analog output and selecting a function.
If characters are written to the front panel display but DS5 does not flash then the program files have become corrupted, contact Teledyne API's Customer Service Department (see Section 14.9) because it may be possible to recover operation of the analyzer. If after 30 – 60 seconds, neither DS5 is flashing nor have any characters been written to the front panel display then the CPU is bad and must be replaced.
14.4. GAS FLOW PROBLEMS The T200 has two main flow paths, the sample flow and the flow of the ozone supply air. With IZS or zero/span valve option installed, there is a third (zero air) and a fourth (span gas) flow path, but either one of those is only controlled by critical flow orifices and not displayed on the front panel or stored to the DAS.
1” below ambient atmospheric pressure) and that the RCEL pressure is below 10 in-Hg-A. The T200 will calculate a sample flow up to about 14 in-Hg-A RCEL pressure but a good pump should always provide less than 10 in. ...
Also, check the flow assembly o-rings and replace as needed. 14.4.1.4. Sample Flow is Zero or Low but Analyzer Reports Correct Flow Note that the T200 analyzer can report a correct flow rate even if there is no or a low actual sample flow through the reaction cell.
Page 308
Teledyne API – T200 NOx Analyzer Operation Manual The T200 features a new orifice holder, which makes switching sample and ozone flow orifices very easy, refer to Section 13.3.10 on how to change the sample orifices and Appendix B for part numbers of these assemblies.
by a broken PMT temperature control circuit, allowing high zero offset (repair the faulty PMT cooler). After fixing the cause of a high Auto Zero filter reading, the T200 will take 15 minutes for the filter to clear itself, or ...
14.5.3. UNSTABLE ZERO AND SPAN Leaks in the T200 or in the external gas supply and vacuum systems are the most common source of unstable and non-repeatable concentration readings. Check for leaks in the pneumatic systems as described in Section 13.3.12.
Check to ensure that there is no ambient air leaking into zero air line. Check for leaks in the pneumatic systems as described in Section 13.3.12. 14.5.6. NON-LINEAR RESPONSE The T200 was factory calibrated to a high level of NO and should be linear to within 1% of full scale. Common causes for non-linearity are: ...
General Troubleshooting & Repair Teledyne API – T200 NOx Analyzer Operation Manual The standard gases may be mislabeled as to type or concentration. Labeled concentrations may be outside the certified tolerance. The sample delivery system may be contaminated.
Teledyne API – T200 NOx Analyzer Operation Manual General Troubleshooting & Repair 14.5.8. DISCREPANCY BETWEEN NO AND NO SLOPES If the slopes for NO and NO are significantly different after software calibration (more than 1%), consider the following three problems: ...
General Troubleshooting & Repair Teledyne API – T200 NOx Analyzer Operation Manual Dirty or plugged critical flow orifices. Check flows, pressures and, if necessary, change orifices (Section 13.3.10). Wrong materials in contact with sample - use glass, stainless steel or Teflon materials only. Porous materials, in particular, will cause memory effects and slow changes in response.
ON/OFF Switch on the front panel will trip to the OFF position immediately after power is switched on. T200’s without internal pumps that are configured for 230 V will still turn on at 115 V, but the heaters may burn out or not heat up fast enough. CAUTION...
Page 316
General Troubleshooting & Repair Teledyne API – T200 NOx Analyzer Operation Manual To assist tracing DC Power Supply problems, the wiring used to connect the various printed circuit assemblies and DC Powered components and the associated test points on the relay PCA follow a standard color-coding scheme as defined in the following table.
Teledyne API – T200 NOx Analyzer Operation Manual General Troubleshooting & Repair AGND -15V -13.5 -16.0 AGND AGND DGND -0.05 +0.05 Chassis DGND Chassis -0.05 +0.05 +12V Ret +12V +11.8 +12.5 DGND +12V Ret DGND -0.05 +0.05 14.7.3. I C BUS Operation of the I C bus can be verified by observing the behavior of D1 on the relay PCA &...
The ANALOG OUTPUT submenu, located under the SETUP MORE DIAG menu is used to verify that the T200 analyzer’s three analog outputs are working properly. The test generates a signal on all three outputs simultaneously as shown in the following table:...
Teledyne API – T200 NOx Analyzer Operation Manual General Troubleshooting & Repair 14.7.6.2. A/D Functions The simplest method to check the operation of the A-to-D converter on the motherboard is to use the Signal I/O function under the DIAG menu to check the two A/D reference voltages and input signals that can be easily measured with a voltmeter.
General Troubleshooting & Repair Teledyne API – T200 NOx Analyzer Operation Manual 14.7.6.3. Status Outputs Figure 14-5: Typical Set Up of Status Output Test To test the status output electronics: 1. Connect a jumper between the “D" pin and the “” pin on the status output connector.
5. Connect a second jumper from the “” pin on the connector to the “B” pin. 6. The status of EXT_ZERO_CAL should change to read “ON”. Table 14-11: T200 Control Input Pin Assignments and Corresponding Signal I/O Functions INPUT CORRESPONDING I/O SIGNAL...
Teledyne API – T200 NOx Analyzer Operation Manual 14.7.7. PRESSURE / FLOW SENSOR ASSEMBLY The flow and pressure sensors of the T200 are located on a PCA just behind the PMT sensor (see Figure 3-5) can be checked with a Voltmeter.
14.7.9. RS-232 COMMUNICATIONS 14.7.9.1. General RS-232 Troubleshooting Teledyne API's analyzers use the RS-232 communications protocol to allow the instrument to be connected to a variety of computer-based equipment. RS-232 has been used for many years and as equipment has become more advanced, connections between various types of hardware have become increasingly difficult.
3. Check to ensure that the set up command is correct (see Section 9.1.9). 4. Verify that the Ready to Send (RTS) signal is at logic high. The T200 sets pin 7 (RTS) to greater than 3 volts to enable modem transmission.
1.0000. Over time, the molybdenum in the NO converter oxidizes and it becomes less efficient at converting NO into NO. To ensure accurate operation of the T200, it is important to check the NO conversion efficiency periodically and to update this value as necessary.
General Troubleshooting & Repair Teledyne API – T200 NOx Analyzer Operation Manual 14.7.10.3. Calculating / Checking Converter Efficiency The T200 to automatically calculate the current NO conversion efficiency by comparing a known starting concentration of NO gas to the measured NO output of the converter. There are three steps to performing this operation.
Page 327
Teledyne API – T200 NOx Analyzer Operation Manual General Troubleshooting & Repair STEP TWO Input the starting NO concentration value into the T200 by pressing: SAMPLE <TST TST> SETUP This step only appears if the analyzer’s reporting range is set for range mode.
Page 328
General Troubleshooting & Repair Teledyne API – T200 NOx Analyzer Operation Manual STEP THREE NO converter efficiency, press: To cause the analyzer to calculate and record the NO Starting from CONVERTER EFFICIENCY MENU (see preceding steps) M-P CAL CONVERTER EFFICIENCY MENU...
Teledyne API – T200 NOx Analyzer Operation Manual General Troubleshooting & Repair NO Converter Performance 14.7.10.4. Evaluating NO If the converter appears to have performance problems (conversion efficiency is outside of allowed range of 96- 102%), check the following: ...
Page 330
8. Subtract line 2 from line 3 & write that number down on line 4. Also write the NO value on line 8 of the data sheet. The specification shown on the data sheet is the value that is used by Teledyne API when NO converters. performing the procedure on new NO ...
General Troubleshooting & Repair Teledyne API – T200 NOx Analyzer Operation Manual 14.7.12. SENSOR MODULE 14.7.12.1. PMT Sensor The photo multiplier tube detects the light emitted by the reaction of NO with ozone. It has a gain of about 500000 to 1000000. It is not possible to test the detector outside of the instrument in the field. The basic method to diagnose a PMT fault is to eliminate the other components using ETEST, OTEST and specific tests for other sub-assemblies.
Teledyne API – T200 NOx Analyzer Operation Manual General Troubleshooting & Repair NOTE This is a coarse test for functionality and not an accurate calibration tool. The resulting PMT signal can vary significantly over time and also changes with low-level calibration.
General Troubleshooting & Repair Teledyne API – T200 NOx Analyzer Operation Manual 14.7.13. PMT PREAMPLIFIER BOARD To check the correct operation of the preamplifier board, perform an the optics test (OTEST) and an electrical test (ETEST) described in Sections 14.7.12.2 and 14.7.12.3 above.
Teledyne API – T200 NOx Analyzer Operation Manual General Troubleshooting & Repair 14.7.14. PMT TEMPERATURE CONTROL PCA The TEC control PCA is located on the sensor housing assembly, under the slanted shroud, next to the cooling fins and directly above the cooling fan.
General Troubleshooting & Repair Teledyne API – T200 NOx Analyzer Operation Manual NOTE This is one of the two settings in the DIAG menu that is retained after you exit the menu. 14.7.16. INTERNAL SPAN GAS GENERATOR AND VALVE OPTIONS The zero/span valves and Internal span gas generator options need to be enabled in the software (contact the factory on how to do this).
Teledyne API – T200 NOx Analyzer Operation Manual General Troubleshooting & Repair The permeation tube of the internal span gas generator option is heated with a proportional heater circuit and the temperature is maintained at 50°C ±1C. Check the front panel display or the IZS_TEMP signal voltage using the SIGNAL I/O function under the DIAG Menu (Section 8.3).
Maintenance procedures (e.g., replacement of regularly changed expendables) are discussed in Section 13 (Maintenance Schedule & Procedures are not listed here). Also, there may be more detailed service notes for some of the below procedures. Contact Teledyne API's Customer Service Department (see Section 14.9).
) DRYER REPLACEMENT The T200 standard configuration is equipped with a dryer for the ozone supply air. An optional dryer is available for the sample stream and a combined dryer for both gas streams can also be purchased. To change one or both of these dryers: 1.
6. Feed NO gas into the analyzer. This should be 90% of the upper limit setting for the T200’s reporting range: EXAMPLE: if the reporting range is set at 500 ppb, use 450 ppb NO. 7. Wait until the STB value is below 0.5 ppb...
Page 341
Teledyne API – T200 NOx Analyzer Operation Manual General Troubleshooting & Repair Figure 14-8: Pre-Amplifier Board Layout 8. Scroll to the NORM PMT test function on the analyzer’s front panel. 9. With the NO gas concentrations mentioned in Step 5 above, the NORM PMT value should be 900 mV.
General Troubleshooting & Repair Teledyne API – T200 NOx Analyzer Operation Manual 14.8.5. REPLACING THE PMT, HVPS OR TEC The photo multiplier tube (PMT) should last for the lifetime of the analyzer, however, the high voltage power supply (HVPS) or the thermo-electric cooler (TEC) components may fail. Replacing any of these components requires opening the sensor module.
Page 343
Teledyne API – T200 NOx Analyzer Operation Manual General Troubleshooting & Repair NOTE If the black PMT housing end plate for the Sensor Assembly is removed, ensure to replace the 5 desiccant bags inside the housing. 8. Remove the dryer packages inside the PMT housing.
General Troubleshooting & Repair Teledyne API – T200 NOx Analyzer Operation Manual 23. Replace the sensor assembly into the chassis and fasten with four screws and washers. 24. Reconnect all electrical and pneumatic connections. 25. Leak check the system (see Section 13.3.12).
San Diego, California 92121-5201USA Toll-free Phone: 800-324-5190 Phone: 858-657-9800 Fax: 858-657-9816 Email: api-customerservice@teledyne.com Website: http://www.teledyne-api.com/ Before you contact Teledyne API's Customer Service, fill out the problem report form in Appendix C, which is also available online for electronic submission at http://www.teledyne-api.com/manuals/. 06858A DCN5847...
A PRIMER ON ELECTRO-STATIC DISCHARGE Teledyne API considers the prevention of damage caused by the discharge of static electricity to be extremely important part of making sure that your analyzer continues to provide reliable service for a long time. This section describes how static electricity occurs, why it is so dangerous to electronic components and assemblies as well as how to prevent that damage from occurring.
Teledyne API – T200 NOx Analyzer Operation Manual A Primer on Electro-Static Discharge 15.2. HOW ELECTRO-STATIC CHARGES CAUSE DAMAGE Damage to components occurs when these static charges come into contact with an electronic device. Current flows as the charge moves along the conductive circuitry of the device and the typically very high voltage levels of the charge overheat the delicate traces of the integrated circuits, melting them or even vaporizing parts of them.
A Primer on Electro-Static Discharge Teledyne API – T200 NOx Analyzer Operation Manual 15.3. COMMON MYTHS ABOUT ESD DAMAGE I didn’t feel a shock so there was no electro-static discharge: The human nervous system isn’t able to feel a static discharge of less than 3500 volts. Most devices are damaged by discharge levels much lower than that.
Page 349
Teledyne API – T200 NOx Analyzer Operation Manual A Primer on Electro-Static Discharge For technicians that work in the field, special lightweight and portable anti-ESD kits are available from most suppliers of ESD protection gear. These include everything needed to create a temporary anti-ESD work area anywhere.
A Primer on Electro-Static Discharge Teledyne API – T200 NOx Analyzer Operation Manual 15.4.2. BASIC ANTI-ESD PROCEDURES FOR ANALYZER REPAIR AND MAINTENANCE 15.4.2.1. Working at the Instrument Rack When working on the analyzer while it is in the instrument rack and plugged into a properly grounded power supply: 1.
A Primer on Electro-Static Discharge 15.4.2.3. Transferring Components from Rack to Bench and Back When transferring a sensitive device from an installed Teledyne API's analyzer to an Anti-ESD workbench or back: 1. Follow the instructions listed above for working at the instrument rack and workstation.
NOTE If you do not already have an adequate supply of anti-ESD gags or containers available, Teledyne API's Customer Service department will supply them (see Section 14.9 for contact information). Follow the instructions listed above for working at the instrument rack and workstation.
Page 353
INDEX Manual-Current Loop, 144, 146 MANUAL-VOLTAGE, 142 Converting Voltage to Current Output, 71 Current Loop, 70, 105 :CriticalflowOrifices, 241 Electrical Connections, 42 :CriticalFlowOrifices, 62, 240 Electronic Range Selection, 106, 138 IND Mode Assignments, 107 OUTPUT LOOP-BACK, 252 Over-Range Feature, 147 100 VAC, 265, 315 Recorder Offset, 148 115 VAC, 265, 267, 315...
Page 361
Teledyne API - T200 and M200E Series (04503D DCN5847) APPENDIX A: Version Specific Software Documentation APPENDIX A: Version Specific Software Documentation APPENDIX A-1: SOFTWARE MENU TREES AND INDEX, SOFTWARE VERSION K.3 ....3 APPENDIX A-2: SETUP VARIABLES FOR SERIAL I/O.............9 APPENDIX A-3: WARNINGS AND TEST MEASUREMENTS .............22 APPENDIX A-4: SIGNAL I/O DEFINITIONS.................26...
Page 362
APPENDIX A: Version Specific Software Documentation Teledyne API - T200 and M200E Series (04503D DCN5847) 06858 DCN5847...
Teledyne API - T200 and M200E Series (04503D DCN5847) APPENDIX A-1: Software Menu Trees and Index, Software Version K.3 APPENDIX A-1: Software Menu Trees and Index, Software Version K.3 Figure A-1: Basic Sample Display Menu 06858 DCN5847...
APPENDIX A-1: Software Menu Trees and Index, Software Version K.3 Teledyne API - T200 and M200E Series (04503D DCN5847) Figure A-2: Primary Setup Menu (Except DAS) 06858 DCN5847...
Page 365
APPENDIX A-1: Software Menu Trees and Index, Software Version K.3 Teledyne API - T200 and M200E Series (04503D DCN5847) Figure A-3: Secondary Setup Menu (COMM & VARS) 06858 DCN5847...
Page 366
APPENDIX A-1: Software Menu Trees and Index, Software Version K.3 Teledyne API - T200 and M200E Series (04503D DCN5847) SAMPLE SETUP MORE RNGE PASS ACAL COMM DIAG VARS HESN INET COM1 COM2 ENTER PASSWORD: 818 ENTER PASSWORD: 818 ENTER PASSWORD: 818 <SET...
Page 367
APPENDIX A-1: Software Menu Trees and Index, Software Version K.3 Teledyne API - T200 and M200E Series (04503D DCN5847) Press Press to start to start test Calibration Press Press to start to start Calibration Calibration (see Appendix A) Sets the...
Page 368
APPENDIX A-1: Software Menu Trees and Index, Software Version K.3 Teledyne API - T200 and M200E Series (04503D DCN5847) Figure A-6: Internal Data Acquisition (DAS) Menu 06858 DCN5847...
T2 Setup Variables, Software Version K.3 Please note that variables listed below under “Setup Variables Available Through serial I/O” should not be changed unless specifically instructed by Teledyne-API customer service or engineering! Changing those variables may cause significant problems with analyzer performance.
Page 370
APPENDIX A-2: Setup Variables For Serial I/O Teledyne API - T200 and M200E Series (04503D DCN5847) Setup Variable Numeric Default Value Value Description Units Range REM_CAL_DURATI Minutes 1–120 Duration of automatic calibration initiated from TAI protocol. CLOCK_ADJ Sec./Day -60–60 Time-of-day clock speed adjustment.
Page 371
Teledyne API - T200 and M200E Series (04503D DCN5847) Appendix A3: Warnings and Test Measurements, Software Version K.3 Setup Variable Numeric Default Value Value Description Units Range concentration to shorten filter. 5–100 0.015 0.1–100 FILT_SHORT_PCT 1–100 Percent change in concentration to shorten filter.
Page 372
APPENDIX A-2: Setup Variables For Serial I/O Teledyne API - T200 and M200E Series (04503D DCN5847) Setup Variable Numeric Default Value Value Description Units Range O2_FILT_DELAY Seconds 0–300 Delay before leaving O adaptive filter mode. Dilution factor for O . Used only O2_DIL_FACTOR —...
Page 373
Teledyne API - T200 and M200E Series (04503D DCN5847) Appendix A3: Warnings and Test Measurements, Software Version K.3 Setup Variable Numeric Default Value Value Description Units Range SG_AZERO_SAMP Samples 1–10 Number of auto-zero samples to average in single-gas measure modes.
Page 374
APPENDIX A-2: Setup Variables For Serial I/O Teledyne API - T200 and M200E Series (04503D DCN5847) Setup Variable Numeric Default Value Value Description Units Range NO2_SPAN2 Conc. 400, 0.01–9999.99 Target NO concentration during converter efficiency calibration of range 2. NOX_SLOPE2 PPM/mV 0.25–4...
Page 375
Teledyne API - T200 and M200E Series (04503D DCN5847) Appendix A3: Warnings and Test Measurements, Software Version K.3 Setup Variable Numeric Default Value Value Description Units Range RCELL_SET ºC 30–70 Reaction cell temperature set point and warning limits. Warnings: 45–55 35–45...
Page 376
APPENDIX A-2: Setup Variables For Serial I/O Teledyne API - T200 and M200E Series (04503D DCN5847) Setup Variable Numeric Default Value Value Description Units Range OFLOW_SET cc/m 0–500, Ozone flow warning limits. Set point is not used. 4, 9 4, 9 100–1000...
Teledyne API - T200 and M200E Series (04503D DCN5847) Appendix A3: Warnings and Test Measurements, Software Version K.3 Setup Variable Numeric Default Value Value Description Units Range MODEM_INIT — “AT Y0 &D0 Any character RS-232 COM1 modem &H0 &I0 S0=2 in the allowed initialization string.
Page 378
APPENDIX A-2: Setup Variables For Serial I/O Teledyne API - T200 and M200E Series (04503D DCN5847) Setup Variable Numeric Default Value Value Description Units Range TEST_CHAN_ID — NONE NONE, Diagnostic analog output ID. Enclose value in double quotes PMT DE- (") when setting from the RS-...
Page 379
Teledyne API - T200 and M200E Series (04503D DCN5847) Appendix A3: Warnings and Test Measurements, Software Version K.3 Setup Variable Numeric Default Value Value Description Units Range RCELL_PROP 1/ºC 0–10 Reaction cell PID temperature control proportional coefficient. Reaction cell PID temperature RCELL_INTEG —...
Page 380
APPENDIX A-2: Setup Variables For Serial I/O Teledyne API - T200 and M200E Series (04503D DCN5847) Setup Variable Numeric Default Value Value Description Units Range SERIAL_NUMBER — “00000000 ” Any character Unique serial number for in the allowed instrument. Enclose value in character set.
Page 381
Teledyne API - T200 and M200E Series (04503D DCN5847) Appendix A3: Warnings and Test Measurements, Software Version K.3 Setup Variable Numeric Default Value Value Description Units Range FACTORY_OPT — 0–65535 Factory option flags. Add values to combine flags. 1 = enable dilution factor...
APPENDIX A-3: Warnings and Test Measurements Teledyne API - T200 and M200E Series (04503D DCN5847) APPENDIX A-3: Warnings and Test Measurements Table A-2: Warning Messages, Software Version K.3 Name Message Text Description Warnings WSYSRES SYSTEM RESET Instrument was power-cycled or the CPU was reset.
Page 383
Teledyne API - T200 and M200E Series (04503D DCN5847) Appendix A3: Warnings and Test Measurements, Software Version K.3 Name Message Text Description WAUTOZERO AZERO WRN XXX.X MV Auto-zero reading above limit specified by AZERO_LIMIT variable. Value shown in message indicates auto-zero reading at time warning was displayed.
Page 384
APPENDIX A-3: Warnings and Test Measurements Teledyne API - T200 and M200E Series (04503D DCN5847) Table A-3: Test Measurements, Software Version K.3 Name Message Text Description Test Measurements NONOXCONC NO=396.5 NOX=396.5 Simultaneously displays NO and NOX concentrations. not 6 RANGE RANGE=500.0 PPB...
Page 385
Teledyne API - T200 and M200E Series (04503D DCN5847) Appendix A3: Warnings and Test Measurements, Software Version K.3 Name Message Text Description NO2=0.0 PPB concentration for current range. NO2_1 NO2_1=0.0 PPB concentration for range #1. NO2_2 NO2_2=0.0 PPB concentration for range #2.
APPENDIX A-4: Signal I/O Definitions Teledyne API - T200 and M200E Series (04503D DCN5847) APPENDIX A-4: Signal I/O Definitions Table A-4: Signal I/O Definitions, Software Version K.3 Signal Name Bit or Channel Description Number Internal inputs, U7, J108, pins 9–16 = bits 0–7, default I/O address 322 hex 0–7...
Page 387
Teledyne API - T200 and M200E Series (04503D DCN5847) Appendix A3: Warnings and Test Measurements, Software Version K.3 Signal Name Bit or Channel Description Number Control inputs, U14, J1006, pins 1–6 = bits 0–5, default I/O address 325 hex 0–5 Spare 6–7...
APPENDIX A-4: Signal I/O Definitions Teledyne API - T200 and M200E Series (04503D DCN5847) Signal Name Bit or Channel Description Number A status outputs, U24, J1017, pins 1–8 = bits 0–7, default I/O address 323 hex ST_SYSTEM_OK 0 = system OK...
Page 389
Teledyne API - T200 and M200E Series (04503D DCN5847) Appendix A3: Warnings and Test Measurements, Software Version K.3 Signal Name Bit or Channel Description Number Relay board digital output (PCF8575), default I C address 44 hex RELAY_WATCHDOG Alternate between 0 and 1 at least every 5...
Page 390
APPENDIX A-4: Signal I/O Definitions Teledyne API - T200 and M200E Series (04503D DCN5847) Signal Name Bit or Channel Description Number Rear board primary MUX analog inputs, MUX default I/O address 32A hex PMT_SIGNAL PMT detector HVPS_VOLTAGE HV power supply output...
Teledyne API - T200 and M200E Series (04503D DCN5847) Appendix A3: Warnings and Test Measurements, Software Version K.3 Signal Name Bit or Channel Description Number Rear board analog outputs, default I/O address 327 hex CONC_OUT_1 Concentration output #1 (NO DATA_OUT_1...
APPENDIX A-5: Trigger Events and DAS Functions Teledyne API - T200 and M200E Series (04503D DCN5847) APPENDIX A-5: Trigger Events and DAS Functions Table A-5: DAS Trigger Events, Software Version K.3 Name Description ATIMER Automatic timer expired EXITZR Exit zero calibration mode...
Page 393
Teledyne API - T200 and M200E Series (04503D DCN5847) Appendix A3: Warnings and Test Measurements, Software Version K.3 Table A-6: DAS Functions, Software Version K.3 Name Description Units PMTDET PMT detector reading RAWNOX Raw PMT detector reading for NOX RAWNO...
Page 394
APPENDIX A-5: Trigger Events and DAS Functions Teledyne API - T200 and M200E Series (04503D DCN5847) Name Description Units N2ZSC3 concentration for range #3 during zero/span calibration, just before computing new slope and offset CO2ZSC concentration during zero/span calibration, just...
Page 395
Teledyne API - T200 and M200E Series (04503D DCN5847) Appendix A3: Warnings and Test Measurements, Software Version K.3 Name Description Units XIN1 Channel 1 Analog In XIN1SLPE Channel 1 Analog In Slope XIN1OFST Channel 1 Analog In Offset XIN2 Channel 2 Analog In...
APPENDIX A-6: Terminal Command Designators Teledyne API - T200 and M200E Series (04503D DCN5847) APPENDIX A-6: Terminal Command Designators Table A-7: Terminal Command Designators, Software Version K.3 Command Additional Command Syntax Description Display help screen and this list of ? [ID]...
Page 397
Teledyne API - T200 and M200E Series (04503D DCN5847) Appendix A3: Warnings and Test Measurements, Software Version K.3 The command syntax follows the command type, separated by a space character. Strings in [brackets] are optional designators. The following key assignments also apply.
Teledyne API - T200 and M200E Series (04503D DCN5847) APPENDIX A-7: MODBUS Register Map APPENDIX A-7: MODBUS Register Map MODBUS Register Description Units Address (decimal, 0-based) MODBUS Floating Point Input Registers (32-bit IEEE 754 format; read in high-word, low-word order; read-only)
Page 399
Teledyne API - T200 and M200E Series (04503D DCN5847) Appendix A3: Warnings and Test Measurements, Software Version K.3 MODBUS Register Description Units Address (decimal, 0-based) PMT temperature C Sample flow rate cc/m Sample pressure “Hg Internal box temperature C High voltage power supply output...
Page 400
APPENDIX A-7: MODBUS Register Map Teledyne API - T200 and M200E Series (04503D DCN5847) MODBUS Register Description Units Address (decimal, 0-based) offset sensor cell temperature °C MODBUS Floating Point Holding Registers (32-bit IEEE 754 format; read/write in high-word, low-word order; read/write) Maps to NOX_SPAN1 variable;...
Page 401
Teledyne API - T200 and M200E Series (04503D DCN5847) Appendix A3: Warnings and Test Measurements, Software Version K.3 MODBUS Register Description Units Address (decimal, 0-based) IZS temperature warning In low span calibration mode NO concentration alarm limit #1 exceeded NO concentration alarm limit #2 exceeded...
Page 402
APPENDIX A-7: MODBUS Register Map Teledyne API - T200 and M200E Series (04503D DCN5847) This page intentionally left blank. A-42 06858 DCN5847...
APPENDIX B - Spare Parts Use of replacement parts other than those supplied by T-API may result in non compliance with European standard EN 61010-1. Note Due to the dynamic nature of part numbers, please refer to the Website or call Note Customer Service for more recent updates to part numbers.
Page 404
This page intentionally left blank. 06858A DCN5847...
Page 405
T200 Spare Parts List PN 06847A DCN5809 08/18/2010 1 of 3 page(s) Part Number Description 000940100 ORIFICE, 3 MIL, DILUTION & VACUUM MANIFOLDS & IZS 000940400 ORIFICE, 4 MIL, OZONE FLOW & O2 OPTION 000940600 ORIFICE, 10 MIL, SAMPLE FLOW & DILUTION & VACUUM MANIFOLDS...
Page 406
T200 Spare Parts List PN 06847A DCN5809 08/18/2010 2 of 3 page(s) Part Number Description 045230200 PCA, RELAY CARD W/RELAYS, E SERIES, S/N'S >467 045500100 ASSY, ORIFICE HOLDER, 4 MIL, OZONE FLOW 045500300 ASSY, ORIFICE HOLDER, 10 MIL, SAMPLE FLOW & DIL MANIFOLD...
Page 407
T200 Spare Parts List PN 06847A DCN5809 08/18/2010 3 of 3 page(s) Part Number Description FM0000004 FLOWMETER (KB) FT0000010 FITTING, FLOW CONTROL HW0000005 FOOT, CHASSIS/PUMP PACK HW0000020 SPRING, FLOW CONTROL HW0000030 ISOLATOR, SENSOR ASSY HW0000031 FERRULE, SHOCKMOUNT HW0000099 STANDOFF, #6-32X.5, HEX SS M/F...
Page 408
Recommended Spare Parts Stocking Levels Models T200 and M200E (Reference: 04414Q DCN5480) Recommended Spare Parts Stocking Level: Standard Units Part Number Description 6-10 11-20 21-30 011310000 ASSY, DRYER, NOX 011930000 CD, PMT (R928), NOX, M200A, M200E(KB) 014080100 ASSY, HVPS, SOX/NOX...
Appendix C Warranty/Repair Questionnaire T200 and M200E (04503D, DCN5798) CUSTOMER: ____________________________________ PHONE: ___________________________________________ CONTACT NAME: _______________________________ FAX NO. ___________________________________________ SITE ADDRESS: ______________________________________________________________________________________ MODEL SERIAL NO.: ____________________________ FIRMWARE REVISION: _______________________________ 1. ARE THERE ANY FAILURE MESSAGES? ____________________________________________________________ _____________________________________________________________________________________________________ _____________________________________________________________________________________________________ PLEASE COMPLETE THE FOLLOWING TABLE: (NOTE: DEPENDING ON OPTIONS INSTALLED, NOT ALL TEST...
Page 412
5. IF POSSIBLE, PLEASE INCLUDE A PORTION OF A STRIP CHART PERTAINING TO THE PROBLEM. CIRCLE PERTINENT DATA. 6. THANK YOU FOR PROVIDING THIS INFORMATION. YOUR ASSISTANCE ENABLES TELEDYNE API TO RESPOND FASTER TO THE PROBLEM THAT YOU ARE ENCOUNTERING.
APPENDIX D – Wire List and Electronic Schematics 06858A DCN5847...
Page 414
This page intentionally left blank. 06858A DCN5847...
Page 415
T200 INTERCONNECT LIST (Reference: 0691101A) Revision Description Checked Date Initial Release 9/3/2010 5833 CONNECTION FROM CONNECTION TO Cable Part Signal Assembly Assembly 0364901 CBL, AC POWER AC Line Power Entry CN0000073 Power Switch SW0000025 AC Neutral Power Entry CN0000073 Power Switch...
Page 416
T200 INTERCONNECT LIST (Reference: 0691101A) CONNECTION FROM CONNECTION TO Cable Part Signal Assembly Assembly 0402603 CBL, IZS & O2 SENSOR HEATERS/THERMISTORS, REACTION CELL & MANIFOLD THERMISTORS Rcell thermistor A Reaction cell thermistor 041920000 Motherboard 058021100 Rcell thermistor B Reaction cell thermistor...
Page 417
T200 INTERCONNECT LIST (Reference: 0691101A) CONNECTION FROM CONNECTION TO Cable Part Signal Assembly Assembly 04671 CBL, MOTHERBOARD TO XMITTER BD (MULTIDROP OPTION) Motherboard 058021100 Xmitter bd w/Multidrop 069500000 Motherboard 058021100 14 Xmitter bd w/Multidrop 069500000 RTS0 Motherboard 058021100 13 Xmitter bd w/Multidrop...
Page 418
T200 INTERCONNECT LIST (Reference: 0691101A) CONNECTION FROM CONNECTION TO Cable Part Signal Assembly Assembly 06746 CBL, MB TO 06154 CPU Motherboard 058021100 Shield Motherboard 058021100 14 CPU PCA 067240000 COM1 RTS0 Motherboard 058021100 13 CPU PCA 067240000 COM1 Motherboard 058021100...
Page 420
VERSION TABLE 016680000 - CE MARK VERSION STD PROD. VERSION UP TO 10/99 016680100 - NON CE MARK (OBSOLETE) +15V +15V 016680200 - SUB PS 17 SWITCHER FOR LINEAR SUPPLY +15V DELETE COMPONENTS T1, D1, D2, C9, C11, PTC1, PTC2, U2 ADD COMPONENTS 016680300 - LOW OUTPUT + FIXED FREQ 1.2K...
Page 423
General Trace Width Requirements 1. Vcc (+5V) and I2C VCC should be 15 mil 2. Digitial grounds should be at least 20 mils 3. +12V and +12V return should be 30 mils AC_Line 4. All AC lines (AC Line, AC Neutral, RELAY0 - 4, All signals on JP2) should be 30 mils wide, with 120 mil isolation/creepage distance around them 5.
Page 424
AC_Line RELAY3 RELAY4 Aux Relay Connector RELAY3 RELAY4 MOLEX6 AC_Neutral I2C_Vcc I2C_Vcc SLD-RLY SLD-RLY IO10 IO11 IO12 +12V IO13 Valve4 IN 4 OUT4 IN 3 SN74HC04 Valve5 VLV_ENAB ENABLE OUT 3 IN 2 OUT 2 Valve6 IN 1 OUT 1 Valve7 UDN2540B(16) CON10...
Page 425
+15V 2.55K VDD_TC +15V 5.6V LTC1050 TYPE J J TC Connector OPA2277 -15V MICROFIT-4 249K TYPE k K TC Connector 332K JUMPER VEE_TC 5.6V 2.55K VDD_TC -15V LTC1050 OPA2277 TYPE J 249K J TC Connector 1 uF TOUT 676K JUMPER LT1025 VEE_TC Title...
Page 426
+15V 1.1K ASCX PRESSURE SENSOR 1.0UF LM4040CIZ S1/S4_OUT S2_OUT S3_OUT 10V_REF +15V ASCX PRESSURE SENSOR MINIFIT6 +15V FLOW SENSOR FM_4 CN_647 X 3 1.0UF +15V LM4040CIZ CON4 SCH, PCA 04003, PRESS/FLOW, 'E' SERIES The information herein is the APPROVALS DATE property of API and is DRAWN submitted in strictest con-...
Page 427
Interconnections preamp cktry HVPS Cktry 04181H-1-m100e200e.sch 04181H-2-m100e200e.SCH 04181H-3-m100e200e.SCH Title M100E/200E PMT Preamp PCA Size Number Revision 04181 Date: 10-May-2007 Sheet of File: N:\PCBMGR\04179cc\Source\RevG\04179.ddb Drawn By: 06858A DCN5847 D-15...
Page 428
ON JP2: +15V PMT TEMPERATURE FEEDBACK FOR 100E/200E : SHORT PINS 2 &5 ONLY. +12V_REF FOR 200EU: SHORT PINS 3 & 6 and PINS 2 & 5. +15V +15V TJP1A OPTIC TEST 6.2V ZENER SEE TABLE PMT TEMP CONFIG JUMPER TJP2A 6.2V 150K...
Page 431
Not Used Title SCH, E-Series Analog Output Isolator, PCA 04467 Size Number Revision 04468 Date: 6/28/2004 Sheet of File: N:\PCBMGR\..\04468B.sch Drawn By: 06858A DCN5847 D-19...
Page 432
Name Name 04524-p1.sch 04524-p2.sch Name 04524-p3.sch Title M100E/200E/400E RELAY PCA SCHEMATIC Size Number Revision 04522 Date: 16-May-2007 Sheet of File: N:\PCBMGR\04522cc\source\04522.ddb Drawn By: D-20 06858A DCN5847...
Page 433
General Trace Width Requirements 1. Vcc (+5V) and I2C VCC should be 15 mil 2. Digitial grounds should be at least 20 mils 3. +12V and +12V return should be 30 mils AC_Line 4. All AC lines (AC Line, AC Neutral, RELAY0 - 4, All signals on JP2) should be 30 mils wide, with 120 mil isolation/creepage distance around them AC_Neutral 5.
Page 438
J1010 DB9 FEMALE RS-GND1 RX for Com1 RTS1 TX for Com1 CTS1 TV ARRAY SMDA15LCC LED, GRN, smt 1206 LED, RED R111 -15V 4.9K, 5% 4.9K, 5% 10k, 1% J1013 DCE side of switch is side towards pin 1, RTS0 CTS0 RS-GND0 RTS1...
Page 439
U20B ISOLATED 0-20MA OPTIONAL BOARDS DACV DACV TP26 DAC, 12 BIT +15V 74HC32 DOUT 0.15 uF, ceramic 0.15 uF, ceramic U20C CSDACA ANALOG VOLTAGE & CURRENT OUTPUTS DUAL DAC A1 J1020 -15V +15V OPA4277UA FE BEAD POT, DIGITAL WRDAC 74HC32 U35A SOCKET U31 R64 22...
Page 440
CH14 IC0000127 CH13 OPA2277UA CH12 CH11 +15V -15V +15V +15V 0.15 uF, ceramic ANALOG INPUTS 10 uF, 35V, TANTALUM 0.15 uF, ceramic VREF AGND J110 J109 RS0000282 MICROFIT-12 MICROFIT-12 ADCVREF 1.0K IN 1 R45 induces an 1M, 1%, 1206 CHIP IN 2 U53A offset in analog...
Page 441
+15V +5VANA BYPASS CAPS MUST BE WITHIN ON/OFF 1/2" OF THE + C60 REGULATOR 10 uF, 35V, TANTALUM LP2981IM5 INPUT/OUTPUT PINS 1 uF D[0..7] THERMISTER +15V Install XT1 (through hole) +5VANA OR XT2 (SMD), MAX382CWN but not both. TEMPMUX THERMISTER1 +VSS THERMISTER1 IN 1...
Page 442
CONTROL INPUTS 470x8 15Kx8 DIGIO0 PS2702-4 J1004 EXTERNAL CONTROL L22 FE BEAD 74HC541 D[0..7] TERMBLOCK-10 10000 pF EXT_+5V_OUT 10000 pF 330 pF, 50V 330 pF, 50V PS2702-4 330 pF, 50V 330 pF, 50V L25 FE BEAD Place these termination resistors at the end of each data line.
Page 443
DIGITAL OUTPUTS RN10 470x8 PS2702-4 10000 pF TP19 SHDN SHDN 10000 pF 74HC574 DIGIO2 74HC32 L46 FE BEAD J1017 A STATUS OUTPUTS PS2702-4 D[0..7] L47 FE BEAD TERMBLOCK-12 10000 pF FE BEAD 10000 pF RESETTABLE FUSE, 0.3A, 60V FE BEAD DIODE, SCHOTTKY EXT_+5V_OUT Title...
Page 444
DIGITAL OUTPUTS 510x8 PS2702-4 SHDN SHDN 10000 pF 74HC574 DIGIO0 U59C 10000 pF 74HC32 L31 FE BEAD J1008 PS2702-4 D[0..7] CONTROL OUTPUTS L35 FE BEAD CO_EXT_RET TERMBLOCK-14 L59 FE BEAD EXTERNAL CONNECTOR SOLDER SIDE 10000 pF 510x8 10000 pF 10000 pF PS2702-4 SHDN 74HC574...