Summary of Contents for SMAR TT481W

  • Page 3 Smar provides specific training to instruct and qualify such professionals. However, each country must comply with the local safety procedures,...
  • Page 4 – Operation & Maintenance, Instruction Manual TT481 WirelessHART...
  • Page 5: Table Of Contents

    Table of Contents TABLE OF CONTENTS INTRODUCTION ............................. VII INSTALLATION FLOWCHART ........................XIII SECTION 1 - INSTALLATION ........................1.1 GENERAL ..................................1.1 MOUNTING ................................. 1.1 BATTERY MODULE WIRING ............................. 1.4 INSTALLATION IN HAZARDOUS LOCATIONS ......................1.6 INTRINSICALLY SAFE ..............................1.6 SECTION 2 - OPERATION ..........................
  • Page 6 – Operation & Maintenance, Instruction Manual TT481 WirelessHART...
  • Page 7: Introduction

    Installation Flowchart INTRODUCTION WirelessHART technology overview The WirelessHART technology is based on a wireless mesh network communication protocol used in process automation applications. It adds wireless capabilities to the HART protocol, while maintaining compatibility with existing HART devices, commands and already known and used tools. WirelessHART network Basically, a WirelessHART network, defined in the HART specifications, consists of a host, a WirelessHART Gateway and one or more field devices and/or WirelessHART adapters.
  • Page 8 – Operation & Maintenance, Instruction Manual TT481 WirelessHART Network Manager The Network Manager is an application that can be embedded in the WirelessHART Gateway. On a WirelessHART network is only allowed to have one Network Manager. Among its responsibilities, the Network Manager distributes network identity (advertisement) publishing its existence, manages and authenticates the addition (joining) of devices to the network.
  • Page 9 TT400 WirelessHART - Manual de Instrução, Operação e Manutenção Planning an WirelessHART network The planning of a WirelessHART network is a task that is very similar to the activities that currently we perform with conventional wired devices. Furthermore, due to the simplicity of a mesh WirelessHART network, is exempt, in general, detailed field surveys, which are usually needed when we plan networks based on other wireless technologies.
  • Page 10 – Operation & Maintenance, Instruction Manual TT481 WirelessHART o Reposition the gateway as needed ; o Check the coverage area of each device; o Ensure that each device has three neighbors within its coverage area; o Place the repeaters as needed. Installation As mentioned before, WirelessHART devices should be connected to the process and configured the same way as conventional wired HART devices.
  • Page 11 TT400 WirelessHART - Manual de Instrução, Operação e Manutenção o If there are high devices, does not exceed 45 ° viewing angles between them; o Make sure that the gateway is integrated to the host system as planned. Commissioning Commissioning consists of testing the device and verifying its configuration data. The TT481 WirelessHARTTM can be commissioned both before and after installation.
  • Page 12 Always install the equipment with the antenna in the vertical direction. If the device is installed horizontally, consult Smar to purchase the antenna for horizontal assemblies, in order to be 90 ° with the equipment. No WirelessHART device should be located at the highest point of the plant, preventing it from lightning;...
  • Page 13 (when available). When replacing the Battery Module (Smar code 400-1273), the replacement must be configured using a configurator that will cause the equipment to reset the estimated lifetime count for the new module. (Search for “Battery Life Remaining”).
  • Page 14: Installation Flowchart

    – Operation & Maintenance, Instruction Manual TT481 WirelessHART INSTALLATION FLOWCHART Start Was the transmitter configured on the bench to match the application? Configure the sensor type Check the area classification and its practices Configure the sensor connection Install the transmitter preferably on weather- protected areas Configure the engineering unit Install the transmitter according...
  • Page 15: Section 1 - Installation

    Section 1 INSTALLATION General The overall accuracy of temperature and other measurements depends on several variables. Although the transmitter has an outstanding performance, proper installation is essential, in order to maximize its performance. Among all factors, which may affect transmitter accuracy, environmental conditions are the most difficult to control.
  • Page 16 – Operation & Maintenance, Instructions Manual TT481 WirelessHART MOUNTING BRACKET FOR TUBE 2" Figure 1.1 (a) – Dimensional Drawing - TT481 WirelessHART with 4 Channels...
  • Page 17 Installation MOUNTING BRACKET FOR TUBE 2" Figure 1.1 (b) – Dimensional Drawing - TT481 WirelessHART with 8 Channels WARNING The TT481 WirelessHART should always be installed with the antenna positioned upward.
  • Page 18: Battery Module Wiring

    – Operation & Maintenance, Instructions Manual TT481 WirelessHART Minimum 0.5 m Legend: 1 - Vertical Obstacle 2 - Floor 3 - Minimum of three neighboring equipment Minimum 1.5 m 4 - Advised five neighboring transmitters Figure 1.2 – Wiring Diagram for Wireless Transmitters Battery Module Wiring The device go out of factory with battery module off, for safety and standard shipping.
  • Page 19 Installation Figure 1.3 – Connecting the Battery Module to Electronic Board The maintenance port allows initial configuration of the transmitter. To do this, it must connect a HART configurator in communication terminals "CN7" and "CN8", shown in Figure 1.4.. Figure 1.4 – Maintenance Port...
  • Page 20: Installation In Hazardous Locations

    The instrument modification or parts replacement supplied by other than authorized representative of Smar is prohibited and will void the certification. The transmitters are marked with options of the protection type. The certification is valid only when the protection type is indicated by the user.
  • Page 21: Section 2 - Operation

    Section 2 OPERATION The TT481 WirelessHART accepts signals from mV generators such as thermocouples or resistive sensors such as RTDs. The criterium is that the signal is within the range of the input. For mV, the range is -50 to 500 mV and for resistance, 0 to 2000 Ohm. For current the range is 4 to 20 mA. Functional Description-Hardware Refer to the block diagram (Figure 2.1).
  • Page 22: Temperature Sensor

    - Operation & Maintenance, Instruction Manual WARNING By no means should be used other than the power supplied by batteries Module Smar (code 400- 1273). When you replace the Battery Module (code Smar 400-1209) to set up the replacement via a configurator that will cause the device to reboot count the estimated lifespan for the new module.
  • Page 23: Thermoresistances (Rtds)

    Operation  NBS (B, E, J, K, N, R, S e T)  DIN (L, U) THERMORESISTANCES (RTDs) Resistance Temperature Detectors, most commonly known as RTD's, are based on the principle that the resistance of a metal increases as its temperature increases. Standardized RTDs, whose tables are stored in the memory of the TT400 WirelessHART , are the following:...
  • Page 24: Currents (4 To 20Ma)

    ® TT481 WirelessHART - Operation & Maintenance, Instruction Manual A differential connection is similar to the two-wire connection and gives the same problem (see Figure 2.4). The resistance of the other two wires will be measured and does not cancel each other out in a temperature measurement, since linearization will affect them differently.
  • Page 25: Writting Protection

    Operation Writting Protection The write protection function can be activated by two means: hardware (switch on the main board) and software. The writing of any parameter is possible only if both guards are disabled. Another way to protect the writing, but in a partial way, is the Hardware Lock option (Lock Device for ®...
  • Page 26 ® TT481 WirelessHART - Operation & Maintenance, Instruction Manual...
  • Page 27: Section 3 - Maintenance

    Section 3 MAINTENANCE General All servicing of Smar transmitters must be done by a qualified person and exchange of components (provided by Smar) must be made only by individuals certified to do so. Diagnostics with the Display The display can show failure messages in the alphanumeric segment. These messages are shown in Table 3.1.
  • Page 28 – Operation & Maintenance, Instruction Manual TT481 WirelessHART Antenna If necessary disassembly the antena set, it is ESSENTIAL to remove the cover device to disconnect the cable from the antenna circuit board. Remove the front screws (2) from the b) Remove cover (1) of the device. gear box cover.
  • Page 29: Reassembly Procedure

    Maintenance Reassembly Procedure WARNING This operation type must be done in a safety area and with the transmitter no energized. Figure 3.1 shows the components position mentioned in this description. With the aid of an Allen wrench, attach Attach the electronic board (6) to the box the battery module (13) using screw using their screws (5).
  • Page 30: Returning Materials

    If it becomes necessary to return the transmitter and/or configurator to Smar, simply contact our office, informing the defective instrument's serial number, and return it to our factory. In order to speed up analysis and solution of the problem, the defective item should be returned with the Service Request Form (SRF –...
  • Page 31 Maintenance Figure 3.1 (b) – Exploded View - TT481 WirelessHART with 8 Channels SPARE PARTS LIST FOR TRANSMITTER Parts Code Antenna 400-1258 Box with Cover - Option C4: 4 channels 400-1268 - Option C8: 8 channels Battery Pack 400-1273 Electronic Board GLL1428 400-1272 Cable Gland Kit 400-1269...
  • Page 32: Section 4 - Technical Characteristics

    - Burst mode to 8s, @25 ºC, network with at least 3 devices neighbor: 5 years Battery PS: The module used batteries in the transmitters must be provided exclusively by Smar (PACK BATTERY - Code 400-1273). 4 1/2 -digit numerical and 5-character alphanumerical LCD indicator (optional). Function and status icon.
  • Page 33 – Operation & Maintenance, Instruction Manual TT481 WirelessHART Human Machine Interface Specifications Item Icon Description Primary Variable Sensor 1 Blinking when the transmitter is seeking wireless network Blinking when connecting to the wireless network Indication of the state Transmitter operating wireless in the display network...
  • Page 34 Technical Characteristics RANGE MINIMUM * DIGITAL SENSOR SPAN mA ACCURACY % 4 to ± 0.1% or ± 4 µA Table 4.4 - 4-20 mA Sensor Characteristics *Reading accuracy on the display and accessed by communication. **Not applicable for the first 20% of range (up to 440 º C). NA Non applicable.
  • Page 35: Ordering Code

    – Operation & Maintenance, Instruction Manual TT481 WirelessHART Ordering Code MODEL INTELLIGENT TEMPERATURE TRANSMITTER TT481 WirelessHART COD. Channels Number 4 channels 8 channels COD. Telecommunication Certification Certificate Anatel COD. Certification Type Without Certification Ex-ia (Intrinsic Safe) (pending) COD. Organ Certifier Without Organ Certifier CEPEL COD.
  • Page 36 Technical Characteristics MODEL INTELLIGENT TEMPERATURE TRANSMITTER (CONTINUATION) CÓD. Sensor Type (Sensor 5) (Not applicable for option with 4 channels) RTD Cu10 – GE 400 Ohm Thermocouple type S - NBS RTD Ni120 – Edison Curve #7 2K Ohm Thermocouple type T - NBS RTD Pt50 –...
  • Page 37 – Operation & Maintenance, Instruction Manual TT481 WirelessHART...
  • Page 38: Appendix A - Srf - Service Request Form For Temperature Transmitter

    Date: Signature: For warranty or non-warranty repair, please contact your representative. Further information about address and contacts can be found on www.smar.com/contactus.asp. NOTA (1) This field should be filled out by the Smar. (3) Required for Wireless HART ® devices.
  • Page 39 – Operation, Maintenance and Instructions Manual TT400 WirelessHART...
  • Page 40: Appendix B - Battery Safety Datasheet

    Appendix B BATTERY SAFETY DATASHEET Section 1 – Identification Manufacturer: Tadiran Model: TL-5920 US office address: 2001 Marcus Avenue, Suite 125E, Lake Success, NY 11040 Emergency Telephone: 1-800-424-9300 Information Telephone: 1-516-621-4980 Section 2 – Composition Ingredients Lithium Metal (Li) <5% Thionyl Chloride (SOCl2) <47% Carbon (C)
  • Page 41 – Instruction, Operation and Maintenance Manual TT400 WirelessHART Batteries may explode when exposed to: excessive heat (above 150 °C), recharged, discharged below 0V, punched and crushed. Hydrogen Chloride (HCl) and sulfur dioxide (SO ) can be formed during thermal decomposition of Cl Section 6 –...