Quick Start Guide
Figure 1-7: Non Incendive Field Wiring Installation (CSA) 1056-27/37
A. Sensor cable is shielded. Max cable length is 50 ft. (15.2 m).
B. Sensor cable is shielded. Max cable length is 50 ft. (15.2 m).
C. Metal conduit
D. Metal conduit
E. Ground connection may be made in hazardous area.
F. Metal conduit
WARNING
Flammable
Use with non-flammable process media only.
Note
A. Installation must conform to the CEC.
B. Seal required at each conduit entrance.
C. During installation, leave maximum amount of jacket insulation possible
on N.I. field wiring within instrument enclosure. After termination, wrap
N.I. field wiring within enclosure with mylar tape to ensure adequate
double insulation remains.
Unless otherwise specified.