Environmental protection............................. 5 Use in explosion risk areas pursuant to Directive 94/9/EC (ATEX) ................5 T5500/T6500 without switching contact ......................5 T5500/T6500 I#### with inductive proximity switch KE##I##0C0H2 ..............5 Technical data ................................6 Labeling on the device ..............................6 Labeling on the device for explosion risk areas (ATEX) ..................
11.3 Disposal ................................. 12 Appendix ................................12 12.1 Data sheet for Bourdon tube pressure gauge T5500/T6500 ................. 12 12.2 Declaration of Conformity for device without switching contact ..............13 12.3 Declaration of Conformity for device with inductive proximity switch ............14 12.4...
2 Safety 2.1 General sources of hazards Pressure gauges are pressurized parts where failure can result in hazardous situations. The selection of pressure gauge should be made in accordance with the rules set out in EN 837-2. 2.2 Use in accordance with intended purpose The devices are only to be used for the intended purpose as described by the manufacturer.
The documentation has been filed with TÜV-Nord-Cert (see declaration of conformity). Labeling: II 2 GD c IIC Tx 3.2 T5500/T6500 I#### with inductive proximity switch KE##I##0C0H2 Area of use: Explosion risk areas Zone 1 and 2, risk from gases Permitted temperatures: The maximum occurring surface temperature of 95 °C was determined with no covering of dust and with no safety factor.
5.1 Labeling on the device for explosion risk areas (ATEX) The label with the marking for explosion risk areas is located on the outside of the housing. Device without switching contacts: ###T5500####ATEX or ###T6500####ATEX Page 6 of 21...
Erläuter ung, z usätzlich ei n Hi nweis : 6.3.2 Instrument connection The instrument connection is located on the bottom side (T5500 and T6500) or at the back side (T5500) of the pressure gauge and is a threaded connection. 6.3.3 Vent valve...
6.3.4 Rear wall / plug with blow-out capability The pressure gauge has a plug capable of blowing out on the rear wall of the housing (Model T5500) or a rear wall capable of blowing out (Model T6500). These act as a safety feature pursuant to EN 837-1 and simultaneously allow for temperature compensation for the housing, via a rubber membrane.
When connecting the device, the pipes must be depressurized. The pressure metering pipe must be laid inclined in such a way that, for example, for measurements of fluids no air pockets can form, and for measurements of gases no water pockets. If the necessary incline is not achieved, then at suitable points water separators or air separators must be installed.
Depressurize the pressure metering pipe. Open the bayonet ring and remove ring and window. Hold the pointer in place while turning the screw, until the pointer (after releasing again) has the correct position. Close the case again and pay attention to correct fit of window, gasket and bayonet ring.
When undertaking servicing work on the device, the pressure lines must be depressurized and the plant secured against being switched on again. 9.2 Check on function, and recalibration The check on function and recalibration is carried out at regular intervals, depending on the application. The precise testing cycles should be adjusted in line with the operating conditions and ambient conditions.
Please help to protect the environment and dispose of or recycle the devices and components used in accordance with the applicable regulations. 12 Appendix 12.1 Data sheet for Bourdon tube pressure gauge T5500/T6500 Detailed data sheet is available from supplier’s website (see 1.6 Manufacturer’s address, customer services) This table refers to specific documents:...
Page 13
12.2 Declaration of Conformity for device without switching contact Page 13 of 21...
Page 14
12.3 Declaration of Conformity for device with inductive proximity switch Page 14 of 21...
Page 15
12.4 EU design type test certification (dusts) for inductive proximity switches of types SJ KE##I##0C0H2 Page 15 of 21...
Need help?
Do you have a question about the T5500 and is the answer not in the manual?
Questions and answers