Calibration Procedures
9.7.2.2. SET CO2 SPAN GAS CONCENTRATION:
SAMPLE
RANGE=50.0 PPM
< TST TST >
CAL
SAMPLE
GAS TO CAL:CO
CO
CO2
O2
SAMPLE
GAS TO CAL:O2
CO
CO2
O2
228
Set the expected CO
span gas concentration.
2
This should be equal to the percent concentration of the CO
reporting range (default factory setting = 12%).
CO= XX.XX
SETUP
M-P CAL
ENTR EXIT
M-P CAL
ENTR
EXIT
The CO
span concentration value automatically defaults
2
to 12 %.
If this is not the the concentration of the span gas being
used, toggle these buttons to set the correct concentration
of the CO
calibration gases.
2
Teledyne API – Model T300/T300M CO Analyzer
RANGE=50.0 PPM
<TST TST> ZERO SPAN
CONC
C O2 SPAN CONC:12.00%
0
1
2
.0
0
span gas of the selected
2
CO= XX.XX
EXIT
EXIT ignores the new
ENTR EXIT
setting and returns to
the previous display.
ENTR accepts the new
setting and returns to
the previous menu.
06864B DCN6314