Setup Menu
Alternative Setup Using 250Ω Resistor for Checking Current Output Signal Levels
Figure 5-6:
Table 5-8:
132
An alternative method for measuring the output of the Current Loop converter is to
connect a 250 ohm ±1% resistor across the current loop output in lieu of the current
meter (see Figure 3-9 for pin assignments and diagram of the analog output connector).
This allows the use of a voltmeter connected across the resistor to measure converter
output as VDC or mVDC.
V OUT +
V OUT -
ANALYZER
In this case, follow the procedure above but adjust the output for the following values:
Current Loop Output Check
% FS
0
100
Teledyne API – Model T300/T300M CO Analyzer
V
+ D C
G n d
Volt
Meter
V IN +
250 Ω
V IN -
Recording
Voltage across Resistor for
2-20 mA
500 mVDC
5000 mVDC
Device
Voltage across
Resistor for 4-20 mA
1000 mVDC
5000 mVDC
06864B DCN6314