Setup Menu

5.9.3.1. ANALOG OUTPUT VOLTAGE / CURRENT RANGE SELECTION

Table 5-6: Analog Output Voltage Ranges

RANGE NAME
0.1V
1V
5V
10V
• The default offset for all VDC ranges is 0-5 VDC.
CURR
• While these are the physical limits of the current loop modules, typical applications use 2-20 or 4-20 mA for the lower and upper
limits. Please specify desired range when ordering this option.
The default offset for all current ranges is 0 mA.
Current outputs are available only on A1-A3.
122
In its standard configuration, each of the analog outputs is set to output a 0–5 VDC
signals. Several other output ranges are available. Each range has is usable from -5% to
+ 5% of the rated span.
RANGE SPAN
0-100 mVDC
0-1 VDC
0-5 VDC
0-10 VDC
0-20 mA
Teledyne API – Model T300/T300M CO Analyzer
MINIMUM OUTPUT
-5 mVDC
-0.05 VDC
-0.25 VDC
-0.5 VDC
0 mA
MAXIMUM OUTPUT
105 mVDC
1.05 VDC
5.25 VDC
10.5 VDC
20 mA