Setup Menu
Avoid accidentally setting the LOW
range (RANGE1) of the instrument
with a higher span limit than the
HIGH range (RANGE2).
This will cause the unit to stay in
the low reporting range perpetually
and defeat the function of the
AUTO range mode.
The LOW and HIGH
ranges have separate
slopes and offsets for
computing the CO
concentration.
The two ranges must
be independently
calibrated
.
Toggle these buttons
to select the upper
SPAN limit for the
reporting range.
106
To set individual ranges press the following control button sequence.
SAMPLE
<TST
TST> CAL
SETUP X.X
CFG DAS
SETUP X.X
MODE
SET UNIT
SETUP X.X
SNGL DUAL
SETUP X.X
SNGL DUAL AUTO
SETUP X.X
MODE
SET
SETUP X.X
0
0
SETUP X.X
0
0
Teledyne API – Model T300/T300M CO Analyzer
RANGE=50.0 PPM
CO=XX.XX
PRIMARY SETUP MENU
RNGE
PASS CLK MORE
RANGE MODE MENU
DIL
RANGE MODE:SNGL
AUTO
DIL
ENTR EXIT
RANGE MODE:AUTO
ENTR
RANGE CONTROL MENU
UNIT
DIL
LOW RANGE:50.0 Conc
0
5
0
.0
ENTR EXIT
HIGH RANGE:200.0 Conc
2
0
0
.0
ENTR EXIT
SETUP
EXIT
EXIT
EXIT
EXIT
EXIT discards the new
setting.
ENTR accepts the
new setting.
06864B DCN6314