T3113, T3117 TRANSMITTERS
Programmable transmitter of temperature, relative humidity
and other derived humidity values
with 4-20 mA outputs
Instruction Manual

Advertisement

Table of Contents
loading
Need help?

Need help?

Do you have a question about the T3117 and is the answer not in the manual?

Questions and answers

Summary of Contents for Comet T3117

  • Page 1 T3113, T3117 TRANSMITTERS Programmable transmitter of temperature, relative humidity and other derived humidity values with 4-20 mA outputs Instruction Manual...
  • Page 2: Device Setting From The Manufacturer

    Instruction manual for use of T3113, T3117 transmitters Transmitter is designed for measurement of air temperature at °C or °F and relative humidity without aggressive ingredients with computing one of the following values: dew point temperature, absolute humidity, specific humidity, mixing ratio and specific enthalpy. Measuring temperature and relative humidity sensors are non-removable instrument parts.
  • Page 3: Typical Application Wiring

    0.5 m, otherwise undesirable induction of interference signals can appear. Outside diameter of the cable for T3113(D) and T3117(D) devices must be from 3,5 to 8 mm (e.g. SYKFY), for devices T3113L and T3117L with respect to the female connector. Do NOT connect shielding at connector side.
  • Page 4 Dimensions – T3113D, T3117D Dimensions – T3113L, T3117L Connection: see Appendix A IE-SNC-T3113(7)-07...
  • Page 5: Readings On Lcd Display

    LCD Info mode Several settings of installed transmitter is possible to verify without a use of the computer. It is necessary to connect at least power of current loop I1. Unscrew the transmitter lid and shortly press button between display and interconnection terminals by means of a tool (e.g.
  • Page 6: Technical Parameters

    Technical parameters: Accuracy data shown are for value displayed on LCD display. For value on analog output is valid too, if selected output range is set inside measuring range. Analog outputs: Two galvanic isolated current loops with range from 4 to 20 mA Current output in case of error: <3.8 mA or >24 mA Power: 9 to 30 V dc, maximum ripple 0.5 % Measuring parameters:...
  • Page 7 0 to 100 %RH without condensation Dimensions: see dimensional drawings Weight: approximately T3113(L)(D) 225 g, T3117(L)(D) 580 g Material of the case: ABS, stem from stainless steel This value depends on the atmospheric pressure. For computing is used constant value stored inside device memory.
  • Page 8: Optional Accessory

    Optional accessory Mounting flanges: PP90 Procedure of modification of transmitter adjustment: • Device adjustment is performed by means of the optional SP003 communication cable, connected to USB port of the PC. • It is necessary to have installed configuration program Tsensor on the PC. It is free to download at www.cometsystem.cz.
  • Page 9: End Of Operation

    • Measured temperature is higher than approximately 600°C (i.e. high non-measurable resistance of temperature sensor, probably opened circuit). • Relative humidity is higher than 100%, i.e. damaged humidity sensor, or humidity calculation of humidity is not possible (due to error during temperature measurement). •...