Working with measurement tools
13.9.4.3 Setting up an alarm for a difference calculation
Note
• You can set up an alarm for a difference calculation when defining user presets, or
when editing an image in the archive.
• This procedure assumes that you have previously set up a difference calculation.
To set up an alarm for a difference calculation, do the following:
1. If you are editing an image in the archive, select
2. Select
(Select) and push the navigation pad. This displays a dialog box.
3. Select Delta and push the navigation pad. This displays a context menu.
4. Select
(Set alarm on delta) and push the navigation pad. This displays a dialog
box.
5. In the dialog box, you can define the settings for the alarm.
• Alarm condition: The condition that triggers the alarm. Applicable values are
Above, Below, or Off.
• Alarm limit: The temperature value that will be the critical condition when an alarm
is triggered or not.
• Alarm sound: Applicable values are Beep or No sound.
6. Push the navigation pad. This closes the dialog box.