Comm Port Testing; 131; Figure 6-3: Comm - Com1 Test Port - Teledyne t100 Operation Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for t100:
Table of Contents

Advertisement

Teledyne API - T100 UV Fluorescence SO
6.2.3.

COMM PORT TESTING

Test runs
automatically
6.3.
RS-232
06807F DCN7335
Analyzer
2
The serial ports can be tested for correct connection and output in the COMM menu.
This test sends a string of 256 'w' characters to the selected COM port. While the test is
running, the red LED on the rear panel of the analyzer should flicker.
To initiate the test press, access the COMMUNICATIONS Menu (refer to Figure 5-13),
then press:
SETUP X.X
ID
INET
COM1
SETUP X.X
SET>
EDIT
SETUP X.X
COM1 BAUD RATE:19200
<SET
SET>
EDIT
SETUP X.X
COM1 : TEST PORT
<SET
TEST
SETUP X.X
TRANSMITTING TO COM1
<SET
TEST
Figure 6-3: COMM – COM1 Test Port
The RS232 and COM2 communications (COMM) ports operate on the RS-232 protocol
(default configuration). Possible configurations for these two COMM ports are
summarized as follows:
RS232 port can also be configured to operate in single or RS-232 Multidrop mode
(Option 62); refer to Section 3.3.1.8.
COM2 port can be left in its default configuration for standard RS-232 operation
including multidrop, or it can be reconfigured for half-duplex RS-485 operation
(please contact the factory for this configuration).
Note that when the rear panel COM2 port is in use, except for multidrop
communication, the rear panel USB port cannot be used. (Alternatively, when the USB
port is enabled, COM2 port cannot be used except for multidrop).
Communications Setup and Operation
COMMUNICATIONS MENU
COM2
COM1 MODE:0
Select which
COMM port to
test.
EXIT
EXIT
EXIT
EXIT
EXIT returns to
COMM menu
EXIT

131