Displaying Of Reading In Normal And Alarm Condition; Figure 9 Insert Manual Tag - G-Tek LM Pro T051 Operating Manual

User programmable temperature data logger
Table of Contents

Advertisement

5.6 Displaying of Reading in Normal and Alarm Condition

Consider the following alarm configuration of LM
temperature condition and alarm status when data recording is ON:
Alarm No.
Alarm Type
A1
A2
A3
A4
1) Reading is within all setpoints; high/ low.
-
OK sign, battery, reading and unit will be seen on display.
2) Reading is outside the cumulative alarm setpoint high/ low for time duration less than
alarm high/low delay.
-
OK sign, battery, reading with UP/ Down arrow and unit will be seen on display.
3) Reading is outside the cumulative alarm setpoint high/ low for time duration greater
than alarm high/low delay.
-
Alarm sign, bell sign, battery, reading with Up/ Down arrow, single event indication
and unit will be seen on display.
Page 15

Figure 9 Insert Manual Tag

Setpoint (°C)
High
15.0
High
10.0
Low
00.0
Low
-05.0
Pro
T051 data logger, for explanation of
Event Type
Single
Cumulative
Cumulative
Single
Delay (Hr:Mn)
01:00
10:00
05:00
00:30