5.3 Batch Start of the Device
Once the batch configuration is completed from LMViewE051 application, recording of data
will be started by one of the following criteria:
1. Once the configured start time has been reached.
2. User can manually start the device after batch configuration, if required by pressing
"Start" key for about 10 seconds as shown in figure 7.
When the recording of data is started, STP LED turns off and RUN LED starts flashing and
"REC" message is seen on display.
Note: If not needed, then let the batch start at pre-configured time only.
5.4 View Min/Max and Current Temperature Data
As described in key functions
running condition, the display will show Min/Max temperature for the device and current
temperature data respectively as shown in figure 8.
5.5 Inserting a Tag Event
Whenever user wants to record a specific data other than at the logging interval, it can be
done by Manual Tag feature of the device.
While recording of data is going on, user can enter special event with time stamp by pressing
"Min/Max" key for more than 4 seconds. "Tag" message is seen on device display as shown
in figure 9.
Page 14
Figure 7 Start batch manually
(section
4.3), by pressing "Min/Max" key for 1 sec in normal
Figure 8 View Min/Max and current Temperature