Batch Configuration; Figure 5 Set Up Batch - G-Tek LM Pro T051 Operating Manual

User programmable temperature data logger
Table of Contents

Advertisement

5. Delay: Set the required alarm time delay in hr:mn for alarm trigger. The range for
alarm time delay is from 00:00 to 23:59 (hr:mn). Alarm time delay for cumulative event
should be more than or equal to that of single event.
Note: It is advisable to set all four alarm parameters, else the default parameters for the remaining
alarms will be applied.

5.1.2 Batch Configuration

After configuring the alarm parameters, "Setup Batch" window will pop up. Select the "Edit"
button to configure the batch parameters as shown in figure 5.
In Batch configuration of the data logger following terms are configured:
1. ID Name: User should assign an appropriate ID Name of maximum 10 alphanumeric
characters.
2. Batch Name: User should assign an appropriate batch name of maximum 10
alphanumeric characters.
3. Unit of Temperature: The unit for temperature is °C.
4. Store Interval: It is the interval between two successive logging of data. It can be set
between 5 seconds(minimum) to 18 hours(maximum).
5. No. of Readings: This shows no. of data that will be recorded for the given batch
configuration. This number is calculated based on Batch start, stop time and store
interval. Maximum no. of readings is 50000.
6. Time Zone: Select the appropriate time zone from the drop down. The time zone will
be as per UTC time.
7. Current Date and Time: It shows current date and time of the LM
Click on the "clock" button to sync the current date and time of the device with PC
time as per selected time zone.
Page 12

Figure 5 Set up Batch

Pro
T051 data logger.

Advertisement

Table of Contents
loading

This manual is also suitable for:

99962

Table of Contents