T0213 TRANSMITTER
Programmable transmitter of temperature, relative humidity
and other derived humidity values
with 0 to 10 V outputs
Instruction Manual

Advertisement

Table of Contents
loading
Need help?

Need help?

Do you have a question about the T0213 and is the answer not in the manual?

Questions and answers

Summary of Contents for Comet System T0213

  • Page 1 T0213 TRANSMITTER Programmable transmitter of temperature, relative humidity and other derived humidity values with 0 to 10 V outputs Instruction Manual...
  • Page 2: Device Setting From The Manufacturer

    Don’t connect transmitter while power supply voltage is on. Interconnection terminals of T0213(D) are accessible after unscrewing four screws and removing the lid. Lace the cable through a gland at the case wall. Connect the cable to terminals with respecting the signal polarity (see figure).
  • Page 3: Typical Application Wiring

    0.5 m, otherwise undesirable induction of interference signals can appear. Outside diameter of the cable for T0213(D) device must be from 3,5 to 8 mm (e.g. SYKFY), for device T0213L with respect to the female connector. Do NOT connect shielding at connector side.
  • Page 4 Dimensions – T0213D Dimensions – T0213L Connection: see Appendix A IE-SNC-T0213-08...
  • Page 5: Readings On Lcd Display

    Reading next to this symbol is calculated absolute humidity or error state of value. g/kg Reading next to this symbol is calculated specific humidity or mixing ratio (depends on device setting) or error state of value. If specific enthalpy is selected, there is shown only value (number) without corresponding unit! IE-SNC-T0213-08...
  • Page 6: Technical Parameters

    You can choice one of the next value: Dew point temperature Accuracy: ±1,5 °C at ambient temperature T < 25°C and RH >30% Range: -60 to +80 °C Absolute humidity Accuracy: ±3g/m at ambient temperature T < 40°C Range: 0 to 400 g/m IE-SNC-T0213-08...
  • Page 7 This value depends on the atmospheric pressure. For computing is used constant value stored inside device memory. Default value preset by manufacturer is 1013hPa and can be changed by user’s software. This maximum is reached under conditions about 70°C/100%RH or 80°C/70%RH IE-SNC-T0213-08...
  • Page 8: Optional Accessory

    Measured (calculated) value is over upper limit of allowed full scale range. There is a reading „Err1“ on LCD display. Output value is about 10.5 V. This state appears in case of: • Measured temperature is higher than approximately 600°C (i.e. high non-measurable resistance of temperature sensor, probably opened circuit). IE-SNC-T0213-08...
  • Page 9: End Of Operation

    It is a serious error, contact distributor of the instrument. End of operation Device itself (after its life) is necessary to liquidate ecologically! Technical support and service Technical support and service is provided by distributor. For contact see warranty certificate. Appendix A IE-SNC-T0213-08...