Instruction Manual T3111 T3111L T3111P T3111Ex Programmable transmitter of temperature, relative humidity and other calculated humidity values with external probe and 4 - 20 mA outputs...
Page 2
It is prohibited to copy and make any changes in this manual, without explicit agreement of company COMET System, Ltd. All rights reserved. COMET System, Ltd. makes constant development and improvement of their products. Manufacturer reserves the right to make technical changes to the device without previous notice.
TECHNICAL SUPPORT AND SERVICE ..........11 TECHNICAL DATA .................. 12 MEASURED VALUES ................12 CALCULATED HUMIDITY VALUES ............13 GENERAL ....................14 OPERATING CONDITIONS ..............16 END OF OPERATION ................16 DIMENSIONS..................17 TYPICAL APPLICATION WIRING ............19 APPENDIX A ................... 21 IE-SNC-T3111-13...
General description The transmitters T3111 are designed for measurement of ambient temperature and relative humidity of air without aggressive substances. Measured values are converted to other humidity interpretation: dew point temperature, absolute humidity, specific humidity, mixing ratio and specific enthalpy. The transmitters are equipped with an external humidity and temperature probe on a cable.
European Standards EN 60079-0:2018 and EN 60079-11:2012. Transmitter is suitable for use in potentially explosive atmospheres (Apparatus Groups llC) in applications requiring devices category 3G. Type Examination Certificate number FTZÚ 13 ATEX 0189X is available at www.cometsystem.com. IE-SNC-T3111-13...
I1: relative humidity range 4-20 mA corresponds 0 to 100 %RH value at output I2: temperature range 4-20 mA corresponds -30 to +105 ºC display: switched ON °C temperature unit: value displayed at line 2: relative humidity IE-SNC-T3111-13...
“Typical application wiring”. For device connection it is recommended to use a shielded cable with a maximum length of 1200m. External diameter of the cable for T3111, T3111P and T3111Ex connection is 4 to 8 mm. For T3311L connection use cable with respect to female connector parameters (do not connect shielding at connector side).
During info mode no measurement and no output current generation proceed. The transmitter stays at info mode 15 s, and then automatically goes back to measuring cycle. The verifying of settings of the T3111Ex transmitter is permitted only in non-hazardous area. IE-SNC-T3111-13...
Error 3 - there is a reading „Err3“ on LCD display upper line. Error of internal A/D converter appeared (converter does not respond, probably damage of A/D converter). No measurement and calculations of values are proceeded. It is a serious error, contact distributor of the instrument. IE-SNC-T3111-13...
Technical support and service Technical support and service is provided by distributor. For contact see warranty certificate. You can use discussion forum at web address www.forum.cometsystem.cz. IE-SNC-T3111-13...
Accuracy: 0 to +105 °C Measuring range: 0.1 °C Resolution: Response time: t90 < 6 min T3111 t90 < 6 min T3111L t90 < 6 min T3111Ex t90 < 16 min T3111P (temperature step 20 ºC) Relative humidity: ± 2.5 %RH from 5 to 95 %RH at 23 °C...
Accuracy: RH > 30 %, for more details see graphs -60 to +80 ºC Measuring range: Absolute humidity: ±1.5 g/m at ambient temperature T < 25 °C, Accuracy: for more details see graph Measuring range: 0 to 400 g/m IE-SNC-T3111-13...
9 to 30 Vdc Current output in case of error: < 3.8 mA or > 22 mA Recommended calibration interval: 1 year relative humidity 1 year, temperature 2 years Protection: housing with electronics IP65, sensors are located in cover with IP40 protection IE-SNC-T3111-13...
Page 15
The equipment shall only be cleaned with a damp cloth. Electromagnetic compatibility: EN 61326-1 Weight: approximately T3111, T3111L, T3111Ex probe 1m 210 g T3111, T3111L, T3111Ex probe 2m 250 g T3111, T3111L, T3111Ex probe 4m 330 g...
Operating conditions Temperature operating range of the housing with electronics: -30 to +80 ºC T3111 -30 to +80 ºC T3111L -30 to +80 ºC T3111L Ambient temperature range of the housing with electronics: -30 to +60 ºC T3111Ex Temperature operating range of the probe: -30 to +105 ºC...
Typical application wiring T3111 T3111P Loop resistance value Rc = Rm + resistance of wires shall fulfill the condition Rc[Ω] < 40 x Uss[V] – 360. T3111Ex Loop resistance value Rc = Rm + Rs + resistance of wires shall fulfill the condition Rc[Ω] < 40 x Uss[V] – 360.
Page 20
T3111L Loop resistance value Rc = Rm + resistance of wires shall fulfill the condition Rc[Ω] < 40 x Uss[V] – 360. IE-SNC-T3111-13...
Page 21
For the accurate operation of sampling system is important to ensure perfect tightness of all connections and to use corrosion-resistant materials. Tube inclination is chosen so as to avoid the accumulation of fluid in the system. IE-SNC-T3111-13...
Page 22
Technical specification – flow chamber SH-PP Material of flow chamber: stainless steel (DIN 1.4301) Inlet and outlet connection: G1/8 Probe connection: G1/2 Sample flow rate: 0.1 to 3 l/min Operating pressure: up to 25 bar Weight: 580 g Note: Screw-coupling not included IE-SNC-T3111-13...