Twinno T4030 Operating Manual

Twinno T4030 Operating Manual

Online conductivity meter conductivity/resistivity/tds/salinity

Advertisement

Quick Links

T4030 Online Conductivity Meter
Conductivity/Resistivity/TDS/Salinity
Operating Manual
上海淳业仪表科技有限公司
Shanghai Chunye Instrument Technology Co.,Ltd

Advertisement

Table of Contents
loading
Need help?

Need help?

Do you have a question about the T4030 and is the answer not in the manual?

Questions and answers

Subscribe to Our Youtube Channel

Summary of Contents for Twinno T4030

  • Page 1 T4030 Online Conductivity Meter Conductivity/Resistivity/TDS/Salinity Operating Manual 上海淳业仪表科技有限公司 Shanghai Chunye Instrument Technology Co.,Ltd...
  • Page 3: Table Of Contents

    Table of Contents Preface Features Technical Specifications Installation Installation size Embedded installation Wall mounted installation Electrical connection Keypad descriptions Display descriptions Menu Calibration MODBUS RTU Daily Maintenance Package Set Warranty Notes...
  • Page 4: Preface

    Preface Thank you for your support. Please read this manual carefully before use.The correct use will maximize the performance and advantages of the product, and bring you a good experience. When receiving the instrument, please open the package carefully, check whether the instrument and accessories are damaged by transportation and whether the accessories are complete.
  • Page 5: Technical Specifications

    Technical Specifications Conductivity 0~500mS/cm Resolution 0.01μS/cm;0.01mS/cm Basic error ±0.5%F.S Resistivity 0~18.25MΩ/cm Resolution 0.01KΩ/cm;0.01MΩ/cm 0~9999mg/L;0~250g/L Resolution 0.01mg/L;0.01g/L Salinity 0~700ppt Resolution 0.01ppt Stability pH:≤0.02pH/24h;ORP: ≤2mV/24h Temperature 0~50.0°C Temperature resolution 0.1°C Temperature accuracy ±0.3°C Temperature compensation 0~60.0°C Temperature compensation Manual or automatic Electrode residual signal <1‰...
  • Page 6: Installation

    Instrument installation 10 mm 98 mm 120 mm MENU Instrument dimensions Instrument dimensions(Side view) Embedded installation Installation of back cover with waterproofing seal 93 mm (± 0.5 mm) Insert mounting hole size Insert the instrument into the square hole and fix it with the collocated clip.
  • Page 7: Wall Mounted Installation

    Wall mounting a. Fixed bracket (M4 screw) is installed on the wall first. b. Insert the instrument into the fixture along the chute 53 mm Fixture Dimension Fixed bracket (M4 screw) is Insert the instrument into the installed on the wall first. fixture along the chute...
  • Page 8: Electrical Connection

    Instrument connection CELL1 NO 1 CELL2 COM 1 CELL3 NC 1 CELL4 NO 2 Temp COM 2 Temp NC 2 Pow ACL Pow DC+ (9~36V) (85~265V) Cur 1+ Pow DC- Pow AC N Pow PE Com - Pow PE Cur 2+ RS485 B RS485 A Terminal...
  • Page 9: Keypad Descriptions

    Keypad descriptions Keypad operation tips: Short Press: Short Press means to release the key immediately after pressing. ((Default to short presses if not indicated below) Long Press: Long Press is to press the button for 3 seconds and then release it. Press and hold: Press and hold means to press the button, and accelerate after a certain time until the data is adjusted to the user's required value before releasing the button.
  • Page 10: Display Descriptions

    Display descriptions All pipe connections and electrical connections should be checked before use. After the power is switched on, the meter will display as follows. Main measurment+ Unit Instrument type Conductivity / TDS / Salinity / Resistivity 83.01 μS/cm The percentage 25.0 Temp.
  • Page 11: Menu

    Menu structure The following is the menu structure of this instrument,press MENU key to enter menu setting mode: If the monitor prompts you to enter the calibration security password, press the【▼】 key or 【 】key to set the calibration security password, and then press the 【ENT】key to confirm the calibration security password.No initial password here,please enter directly by press【ENT】key.
  • Page 12 Alarm Relay 2 Status High Alarm High/Low Setpoint Low Alarm Limit Value Hysteresis Main Output Current 1 Channel Temperature 4-20mA Output Option 0-20mA 20-4mA Upper Limit Lower Limit Main Current 2 Channel Temperature 4-20mA Output Option 0-20mA 20-4mA Upper Limit Lower Limit 4800BPS RS485...
  • Page 13: Calibration

    Calibration Press [MENU] to enter the setting mode and select the calibration Calibration Standard Calibration Point 1 Enter given standard liquid value( Example: 0 Solution Calibration Point 2 Enter given standard liquid value( Example:1413μS/cm Calibration Calibration Point 3 Enter given standard liquid value( Example:12.43mS/cm Calibration Point 4 Enter given standard liquid value(...
  • Page 14 Field Calibration Select on-site calibration methods: Linear calibration , Offset adjustment , linear adjustment . Field Calibration When the data from laboratory or portable instrument are input into this item, the instrument will automatically correct the data. Conductivity / TDS / Salinity / Resistivity Field Calibration pH / ORP µS/cm: 07.00...
  • Page 15: Modbus Rtu

    MODBUS RTU General Information The hardware version number of this document is V2.0; the software version number is V5.9 and above. This document describes the MODBUS RTU interface in details and the target object is a software programmer. MODBUS command structure Data format description in this document;...
  • Page 16 MODBUS RTU Transmission Mode When the instrument uses RTU (Remote Terminal Unit) mode for MODBUS serial communication, each 8-bit byte of information contains two 4-bit hexadecimal characters. The main advantages of this mode are greater character density and better data throughput than the ASCII mode with the same baud rate.
  • Page 17 Implementation of MODBUS RTU in Instrument According to the official MODBUS definition, the command starts with a 3.5 character interval triggering command, and the end of the command is also represented by a 3.5 character interval. The device address and MODBUS function code have 8 bits. The data string contains n*8 bits, and the data string contains the starting address of the register and the number of read/write registers.
  • Page 18 MODBUS function code 0x10: write multiple registers This function code is used to write continuous registers to remote devices (1... 123 registers) block that specifies the value of the registers written in the request data frame. Data is packaged in two bytes per register.
  • Page 19 Data format in instrument Overview Floating Point Definition: Floating point, conforming to IEEE 754 (single precision) Description Symbol Index Mantissa 30…23 22…0 22…0 Index Deviation Figure 14: floating point single-precision definition (4 bytes, 2 MODBUS registers) Example: Compile decimal 17.625 to binary Step 1: Converting 17.625 in decimal form to a floating-point number in binary form, first finding the binary representation of the integer part 17decimal= 16 + 1 = 1×2...
  • Page 20 If it is large-end storage mode, after executing the above statement, the data stored in outdata of address unit is 0x41 Outdata + 1 stores data as 0x8D address unit (outdata + 2) stores data as 0x00 address unit (outdata + 3) stores data as 0x00 2.
  • Page 21 Read instruction mode The communication protocol adopts MODBUS (RTU) protocol. The content and address of the communication can be changed according to the needs of customers. The default configuration is network address 01, baud rate 9600, even check, one stop bit, users can set their own changes;...
  • Page 22: Daily Maintenance

    Daily maintenance According to the use of the transmitter requirements, the installation location and working conditions are relatively complex, in order to make the transmitter work normally, maintenance personnel need to carry out regular maintenance of the transmitter, please pay attention to the following matters during maintenance: 1)Installed in the outdoor, please check whether the transmitter installation box leakage phenomenon;...
  • Page 23: Package Set

    Complete Set Product Description Quantity 1) T4030 Online Conductivity Meter 2) Instrument Installation Accessories 3) Operating Manual 4) Qualification Certificate Note: Please check the complete set of instruments before use. The company's other series of analytical instruments, please login to our website for enquiries.
  • Page 24: Faq

    1.LCD display is not bright Possible causes:Instrument or LCD Screen power supply failure. Solutions:Check whether the power supply is connected or not, and check whether the power supply wire of the sensor is connected in the wrong direction. 2.No current output Possible causes:It could be a fault in the current module or a wiring fault.
  • Page 25: Warranty

    Warranty We Instruments warrants this product to be free from significant deviations in material and workmanship for a period of one year from the date of purchase. If repair is necessary and has not been the result of abuse or misuse within the warranty period, please return to We Instruments and amendment will be made without any charge.
  • Page 26: Notes

    Notes Distinguished users, please pay attention to the following points when using the instrument, in order to ensure the life and accuracy of the instrument. Careful handling to avoid collision and falling instruments in use. Avoid contact with water or other liquids during use. Don't put the instrument in the sunshine for a long time.
  • Page 28 上海淳业仪表科技有限公司 Shanghai Chunye Instrument Technology Co.,Ltd Tel: 021-61621082 Fax: 021-61621099 Web: www.twinno.cn Add: Building 6, No.166 Mindong Road, Pudong New District, Shanghai, China P.C.: 201209...