Preface Thank you for your support. Please read this manual carefully before use.The correct use will maximize the performance and advantages of the product, and bring you a good experience. When receiving the instrument, please open the package carefully, check whether the instrument and accessories are damaged by transportation and whether the accessories are complete.
The analyzer must be installed by professionals in accordance with relevant local specifications, and instructions are included in the operation manual. Comply with the technical specifications and input requirements of the analyzer. Features Online Turbidity meter is a microprocessor-based water quality online monitoring control instrument.It is widely used in drinking water treatment plants, drinking water distribution networks, swimming pools, water treatment projects, sewage treatment, water turbidity...
Technical Specifications Measurement range 0-99999NTU NTU, mg/L Measurement unit 0.001; 0.01; 0.1; 1 Resolution Basic error ±1%F.S -10~150℃ Temperature Temperature Resolution 0.1℃ Temperature Basic error ±0.3℃ Current outputs Two 4~20mA,20~4mA,0~20mA Signal output RS485 MODBUS RTU Other functions Data record &Curve display Relay control contacts 3 Group: 5A 250VAC,5A 30VDC Optional power supply...
Installations Installation size 235 mm 120 mm 190 mm Hole:8*12.5mm Instrument shape, wall - mounted fixed piece size Fix the fixed piece of the instrument Installation completion drawing...
Instrument connection EC1 EC2 EC3 EC4 T+ T- I N PUT REF S+ S- WE RE CE I1 G I2 I3 G I4 RLY1 RLY2 RLY3 G N L COND TEMP pH DO/FCL FCL/DOZ mA1 mA2 mA3 mA4 AC220V V+ V- RX TX V+ V- A B V+ V- A B A1 B1 RS232...
Keypad descriptions Keypad operation tips: Short Press: Short Press means to release the key immediately after pressing. ((Default to short presses if not indicated below) Long Press: Long Press is to press the button for 3 seconds and then release it. Enter standard solution calibration mode Menu setting mode: Calibrated: check the calibration status Re-...
Display descriptions All pipe connections and electrical connections should be checked before use. After the power is switched on, the meter will display as follows. Main Value+ Unit Year/Month/Day Time:Minutes:Senconds 2019-01-09 12:53:17 4.001 Abnormal alarm of electrode communication The main measurement corresponds to the Temp.
Menu structure The following is the menu structure of this instrument Turbidity Configure Type Sensor mg/L Unit NTC2.252 kΩ Temperature Sensor Temp. NTC10 kΩ Pt100 Pt1000 0.0000 Temperature Offset Automatic Temperature Input Manual °C Temperature Unit °F Calibration Point 1 0.3 NTU (Default, can be modified) Calibration Standard...
Page 12
Output Current 1 Channel Main Temperature Output Option 4-20mA 0-20mA 20-4mA Upper Limit Lower Limit Channel Current 2 Main Temperature Output Option 4-20mA 0-20mA 20-4mA Upper Limit Lower Limit RS485 Baud Rate 4800BPS 9600BPS 19200BPS Parity Check None Even Stop Bit 1 Bit 2 Bit Network Node...
Calibration Press [MENU] to enter the setting mode and select the calibration Standard Calibration Calibration Point 1 Enter given standard liquid value(Example:0.3 NTU) Solution Calibration Point 2 Enter given standard liquid value(Example:2 NTU) Calibration Calibration Point 3 Enter given standard liquid value(Example:10 NTU) Calibration Point 4 Enter given standard liquid value(Example:20 NTU) Calibration Point 5...
Page 14
Field Calibration Select on-site calibration methods: L inear calibration , O ffset adjustment , l inear adjustment . Field Calibration When the data from laboratory or portable instrument are input into this item, the instrument will automatically correct the data. 2019-01-09 12:53:17 Field Calibration pH / ORP...
Page 15
Graphic Trend(Trend Chart) Press M enu k ey to enter the setting mode,set the recording interval, and the instrument will Interval/point Curve 400 points per screen,displays the most recent data trend graph Data Log according to interval Settings query 1h/point (trend chart) 400 points per screen, display trend chart of the last 16 days of data...
MODBUS RTU General Information Overview The hardware version number of this document is V2.0; the software version number is V5.9 and above. This document describes the MODBUS RTU interface in details and the target object is a software programmer. MODBUS command structure Data format description in this document;...
Page 17
MODBUS RTU Transmission Mode When the instrument uses RTU (Remote Terminal Unit) mode for MODBUS serial communication, each 8-bit byte of information contains two 4-bit hexadecimal characters. The main advantages of this mode are greater character density and better data throughput than the ASCII mode with the same baud rate.
Page 18
Implementation of MODBUS RTU in Instrument According to the official MODBUS definition, the command starts with a 3.5 character interval triggering command, and the end of the command is also represented by a 3.5 character interval. The device address and MODBUS function code have 8 bits. The data string contains n*8 bits, and the data string contains the starting address of the register and the number of read/write registers.
Page 19
MODBUS function code 0x10: write multiple registers This function code is used to write continuous registers to remote devices (1... 123 registers) block that specifies the value of the registers written in the request data frame. Data is packaged in two bytes per register.
Page 20
Data format in instrument Overview Floating Point Definition: Floating point, conforming to IEEE 754 (single precision) Description Symbol Index Mantissa 30…23 22…0 22…0 Index Deviation Figure 14: floating point single-precision definition (4 bytes, 2 MODBUS registers) Example: Compile decimal 17.625 to binary Step 1: Converting 17.625 in decimal form to a floating-point number in binary form, first finding the binary representation of the integer part 17decimal= 16 + 1 = 1×2...
Page 21
If it is large-end storage mode, after executing the above statement, the data stored in outdata of address unit is 0x41 Outdata + 1 stores data as 0x8D address unit (outdata + 2) stores data as 0x00 address unit (outdata + 3) stores data as 0x00 2.
Page 22
Read instruction mode The communication protocol adopts MODBUS (RTU) protocol. The content and address of the communication can be changed according to the needs of customers. The default configuration is network address 01, baud rate 9600, even check, one stop bit, users can set their own changes;...
5. Check whether the wiring on the terminal of the instrument is firm. Pay attention to disconnect the AC power before removing the wiring cover. Package Set Product Description Quantity 1) T6570 Online Turbidity Meter 2) Instrument Installation Accessories 3) Operating Manual 4) Qualification Certificate Note: Please check the complete set of instruments before use.
1.LCD display is not bright Possible causes:Instrument or LCD Screen power supply failure. Solutions:Check whether the power supply is connected or not, and check whether the power supply wire of the sensor is connected in the wrong direction. 2.No current output Possible causes:It could be a fault in the current module or a wiring fault.
Warranty We e Instruments warrants this product to be free from significant deviations in material an workmanship for a period of one year from the date of purchase. If repair is necessary and has not been the result of abuse or misuse within the warranty period, please return to We e Instruments and amendment will be made without any charg .
Notes Distinguished users, please pay attention to the following points when using the instrument, in order to ensure the life and accuracy of the instrument. . Careful handling to avoid collision and falling instruments in use. Avoid contact with water or other liquids during use. Don't put the instrument in the sunshine for a long time.