  • Page 2: Specifications

    Span are adjustable over -50 to 212°F and 30 to 320°F respectively allowing for re-ranging by means of a simple calibration process. The gasketed metal housing provides excellent weather resistance and is easily installed using the attached mounting brackets. Request Minco Application Aid 15 for more general information on transmitters. Specifications Sensing Element: 2-wire 1000 ohm @ 0°C platinum RTD, 0.00375 or 0.00385 ohm/ohm/°C TCR.
  • Page 3: Installation

    Totaling these, we get a minimum power supply requirement of 14 VDC. Using a 24 VDC power supply will take care of nearly all installations, but the TT809 will operate with as much as 35 VDC. Minco Tel: 763. 571.3121 • Fax: 763. 571.0927 • www.minco.com...
  • Page 4: Calibration Procedure

    Zero switch settings should then be changed to 38°F (ON, OFF, ON and OFF) which is the next lower range. Likewise if the Span trim pot does not have sufficient adjustment, then the Span switch settings should be changed. Minco Tel: 763. 571.3121 • Fax: 763. 571.0927 • www.minco.com...
  • Page 5 (Upper Minus Lower Temperature Limits) S P A N * Switch Settings Center Point °F °C Table 2 * Span = Upper - Lower Temperature. # Zero ranges overlap next adjacent range. Minco Tel: 763. 571.3121 • Fax: 763. 571.0927 • www.minco.com...
  • Page 6 1758.56 1776.93 1795.28 1813.59 1831.88 1850.13 1868.36 1886.56 1904.73 1922.87 1914.93 1932.46 1949.96 1967.44 1984.88 2002.29 2019.67 ------ ------ ------ 1940.98 1959.06 1977.12 1995.14 2013.14 2031.11 2049.05 ------ ------ ------ Table 4 Minco Tel: 763. 571.3121 • Fax: 763. 571.0927 • www.minco.com...
  • Page 7: How To Order

    1 = No calibration data, sensor or transmitter 2 = Sensor/Transmitter matched at 0°C with NIST cert 3 = Sensor/Transmitter matched at 0, 100, & 260°C with NIST cert TT809PW080E1S1  Sample Part Number Minco Tel: 763. 571.3121 • Fax: 763. 571.0927 • www.minco.com...
  • Page 8 Dimensions: All dimensions are in inches (millimeters) HOUSING 1/2" AND COVER NPT (2) SPAN 4.50 TRANSMITTER (1143,0) RANGING DIP SWITCHES ZERO 2.75 (69,9) (50,8) Figure 3 Minco Tel: 763. 571.3121 • Fax: 763. 571.0927 • www.minco.com...
  • Page 9: Warranty

    Warranty Items returned within one year from the date of sale, transportation prepaid, which Minco Products, Inc. (The "Seller") reasonably determines to be faulty by reason of defective materials or faulty workmanship will be replaced or repaired at the Seller's discretion, free of charge. This remedy is to be the sole and exclusive remedy available to the buyer in the event of a breach by the Seller.