Description

Model TT269 is a 3-wire temperature transmitter for RTD (resistance temperature detector)
thermometers. The Temptran
current changes according to the range marked on the Temptran: 0 mA at the lowest
temperature of the range, rising to 1 mA at the top of the range.

Installation

Locate the Temptran near the RTD, in an area where the ambient temperature stays between -
40 and 85 °C (-40 and 185 °F). Mount with #8 machine screws using the two mounting holes
provided in the Temptran case.
Connect the Temptran as shown below, observing the +/- polarity of the power supply
connections. Maximum DC supply voltage = 35 VDC. The maximum load resistance for the 0
to 1 mA output is 1000 Ω . The RTD connections for the Temptran in the wiring diagram below
must be connected as shown or the transmitter will not function properly.

Wiring Diagram

7300 Commerce Lane, Minneapolis, MN 55432-3177 U.S.A.
Telephone: (763)571-3121
MINCO
TT269
3-wire Temperature Transmitter
Installation and Operating Instructions
TM
converts the RTD's signal into a 0 to 1 mA current. The
PRODUCTS, INC .
/
FAX: (763)571-0927

Advertisement

Table of Contents
loading
Need help?

Need help?

Do you have a question about the TT269 and is the answer not in the manual?

Questions and answers

Summary of Contents for Minco TT269

  • Page 1: Wiring Diagram

    3-wire Temperature Transmitter Installation and Operating Instructions Description Model TT269 is a 3-wire temperature transmitter for RTD (resistance temperature detector) thermometers. The Temptran converts the RTD's signal into a 0 to 1 mA current. The current changes according to the range marked on the Temptran: 0 mA at the lowest temperature of the range, rising to 1 mA at the top of the range.
  • Page 2 Calibration NOTE: Because the transmitter cannot output a negative current, the calibration of the zero cannot be performed at 0 °C (0 mA). The zero calibration should be performed at 6 °C (.05 mA) to ensure proper calibration of the transmitter. If the transmitter is calibrated at 0 °C, the output of the transmitter may not be linear.
  • Page 3: Specifications

    Output: 1 = 0 to 1 mA DC. RTD Element Code: PD = 100 Ω platinum at 0 °C ( .00385 TCR ) Model Number: TT269, RTD Temptran. Dimensions: All dimensions are in inches ( millimeters ) When quality and performance are as important as price, call...