TT859 Temptran 2-wire Temperature Transmitter for RTD Thermometers Installation and Operating Instructions Tel: 763. 571.3121 • Fax: 763. 571.0927 • www.minco.com...
Page 2
20 mA at the top of the range. The leads that supply power also carry the current signal. The TT859 is FIELD RANGEABLE; the Zero and Span are adjustable over -10 to 60°F and 35 to 150°F respectively allowing for re-ranging by means of a simple calibration process.
Totaling these, we get a minimum power supply requirement of 14.9 VDC. Using a 24 VDC power supply will take care of nearly all installations, but the TT859 will operate at voltages up to 35 VDC.
Tables 3 and 4 (page 6). Using “RTD Calc” a more complete resistance vs. temperature chart can be printed; Download it from Minco’s web site, www.minco.com. 3. Set the decade box to the resistance value corresponding to the lower temperature. Adjust the Zero pot until the milliammeter reads 4.0 mA +/- 0.016 mA.
Page 6
37°F (OFF, ON, OFF and ON) which is the next higher range. Likewise if the Span trim pot does not have sufficient adjustment, then the Span switch settings should be changed. Minco Tel: 763. 571.3121 • Fax: 763. 571.0927 • www.minco.com...
Page 7
Span Switch Settings (Upper Minus Lower Temperature Limits) SPAN * Switch Settings Center Point °F °C Table 2 * Span = Upper - Lower Temperature. Zero ranges overlap next adjacent range. Minco Tel: 763. 571.3121 • Fax: 763. 571.0927 • www.minco.com...
Warranty Items returned within one year from the date of sale, transportation prepaid, which Minco Products, Inc. (the “seller”) reasonably determines to be faulty by reason of defective materials or faulty workmanship will be replaced or repaired at the seller’s discretion, free of charge.
Page 11
Dimensions Figure 3 Minco (Main Office) Customer Service/ Minco S.A. 7300 Commerce Lane Order Desk: Usine et Service Minneapolis, MN Tel: 1.763.571.3123 Commercial, Z.I. 55432 Fax: 1.763.571.0942 09310 Aston, France custserv@minco.com Tel: (33) 5 61 03 24 01 Tel: 1.763.571.3121 www.minco.com...