INSTRUCTION MANUAL
Transmission-Type Digital Displacement Sensor / Sensor Head
HG-T1110
MJE-HGT1110 No.0094-67V
Thank you very much for purchasing Panasonic products. Read this Instruction Man-
ual carefully and thoroughly for the correct and optimum use of this product. Kindly
keep this manual in a convenient place for quick reference.
WARNING
Never use this product as a device for personnel protection.
When using devices for personnel protection, use products that meet the laws
and standards for personnel protection that apply in each region or country, such
as OSHA, ANSI and IEC.
This product inspects (judges / measures) objects. Do not use it to ensure safety,
Avoid observing beams in a dark surrounding environment.
Do not look at beams using an optical device such as an optical telephoto system.
Never attempt to disassemble, repair, or modify this product.
Control or adjustment according to procedures other than those provided in this Instal-
lation Instructions Manual may cause exposure to hazardous emitted laser beams.
This document provides a brief summary of mounting and other related information.
For detailed information, refer to the
HG-T Series User's Manual
(our Website:
"
"
https://industry.panasonic.com/ ).
1
REGULATIONS AND STANDARDS
This product conforms to the regulations and standards below.
<Conformity Directives/Conforming Regulations>
EU Law : EMC Directives 2014/30/EU
British Legislation : EMC Regulations 2016/1091
- Applicable Standards
EN 61000-6-4, EN 61000-6-2
2
CONTENTS OF PACKAGE
Sensor Head (Emitter, Receiver)
Instruction Manual (English / Japanese, Chinese / Korean)
General Information for Safety, Compliance, and Instructions
3
FOR SAFE USE OF LASER PRODUCTS
IEC/EN/JIS/GB/KS
In order to prevent user injury caused by a laser product, the following standards
have been established in the IEC Standards, EN Standards, JIS Standards, GB
Standards, KS Standards and FDA Regulations.
IEC : IEC 60825-1:2014
EN : EN 60825-1:2014/A11:2021
JIS : JIS C 6802:2014
GB : GB 7247.1-2012
KS : KS C IEC 60825-1:2014
FDA : PART 1040.10, 1040.11(Laser Notice No.56 applied)
These standards classify laser products by level of laser hazard, and specify safe-
ty preventive measures to be implemented for each class.
This product is equivalent to a "Class 1 Laser Product" based on IEC 60825-
1:2014(EN 60825-1:2014/A11:2021) "Radiation Safety Standards for Class 1 La-
ser Products".
Explanation of hazard levels
Summary of hazard evaluation
Class 1
A laser that is safe when operated under operating conditions that can be reasonably foreseen.
When an unexpected failure occurs, dangerous radiation may be generated. Therefore, pay special attention to safety.
dards for laser products.
Explanation label
<Label position>
Emitter side
FDA
Export to the USA.
When included in equipment for export to the USA, the product is subject to the
regulations of the laser standards of the FDA (Food and Drug Administration). In
order to prevent user injury caused by a laser product, Part 1040 (Performance
Standards for Light-Emitting Products) of the FDA standards has been estab-
lished. These standards classify laser products by level of laser hazard, and spec-
ify safety preventive measures to be implemented for each class.
This product complies with the FDA regulations (FDA 21 CFR 1040.10 and
1040.11) in accordance with FDA Laser Notice No. 56, except for complying with
IEC 60825-1 Ed. 3.(Class 1 Laser Product)
4
DESCRIPTION OF PARTS
<Emitter>
3
2
1
1 pc.
1 pc. each
1 pc.
5
5
Name
1 Laser emission indicator (Green)
Lights while laser light is emitted.
2 Light-emitting face
Face that emits the laser.
3 Light-receiving face
Face that receives the laser.
• Lights green: Controller judgment output is within the set range.
Judgment output indicator
4
• Lights orange: Controller judgment output is outside the set range.
(Orange / Green)
• OFF: Beam alignment function is in use.
5 Sensor head connection cable connector Connects to the sensor head connection cable.
5
MOUNTING
Mounting the sensor head
Do not allow contaminants that will refract light such as water, oil or finger-
prints, or contaminants that will block light such as dust or dirt, to collect on the
light-emitting and light-receiving faces of the sensor head If there is dirt or other
contamination on a surface, wipe with lens paper or a soft cloth that will not leave
that has been moistened with absolute alcohol.
A serial number is indicated on the reverse sides of the light-emitting and light-re-
ceiving faces of the product. Always combine an emitter and receiver with identi-
cal serial numbers.
Use M3 screw with washer (obtain separately) to mount the product.
-
8mm
M3 screws with washers
(Please obtain separately)
Serial number
Attaching the sensor head and Sensor Head Connection Cable CN-HT-C
tional)
Always use Sensor Head Connection Cable CN-HT-C
product to the controller.
the controller power before inserting or removing the connectors.
Do not grasp the cable to pull out or insert the connectors. Wires may break if
excessive stress is applied to the cable.
nector may fall out and cause an error.
The sensor head connection cable branches into two cables. The cables are
identical and can be connected to either the emitter or receiver
Mounting procedure
1. Insert the sensor head connecting cable to the sensor head connection cable
connector of the sensor head with the mark aligned.
Mark
1
2. Turn the fastening ring on each sensor head connection cable connector in the
<Label position>
Emitter side
2
Removal procedure
1. Turn the fastening ring on each sensor head connection cable connector in the
direction of the arrow to loosen.
<Receiver>
2. Grasp each connector on the sensor head connection cable and pull out to re-
4
move.
When disconnecting, always make sure that the fastening ring has been complete-
ly loosened before pulling out the cable. Risk of damage if you pull the cable with
Function
excessive force (15N or more) with the fastening ring tightened.
6
BEAM ALIGNMENT AND REFERENCE WAVEFORM REGISTRATION
When installing the product, always register a reference waveform. If a reference
waveform is not registered, correct measurement will not be possible.
Register the reference waveform without a measurement object.
prevent correct detection. In this event, mount the emitter and receiver pair at an
justing the angle, always perform beam alignment.
<Beam alignment>
Check the digital display of the controller as you align the beam.
Mounting hole dimensions
<On emitter>
<On receiver>
<Reference waveform registration>
You can perform reference waveform registration by pressing the ENTER key on
the controller with the beam correctly aligned. The reference waveform is stored
in EEPROM in the receiver.
24mm
14mm
-
-
7
CAUTIONS
The HG-T
tions will not be met and product malfunctioning or damage may occur.
Mark
This product has been developed / manufactured for industrial use only.
The receiver in this product uses an EEPROM. The EEPROM has a service life of
one million setting operations.
1
Sensor head connection cable
CN-HT-C
and product damage. There is also a risk of a noticeable reduction of service life.
To ensure performance, do not use the product after turning ON the power until it
Sensor head connection cable connector
has warmed up for at least 30 minutes.
Do not wire in parallel with a high-voltage line or power line, or run through the
same conduit. Risk of malfunctioning due to induction.
Do not apply stress such as excessive bending or pulling to the extracted part of a
cable.
Do not touch or allow any foreign objects to enter the terminal inside the connector.
Do not allow ambient light such as sunlight to directly enter the sensor head re-
2
ceiver. When precision is particularly required, take measures such as attaching a
shield to the sensor head.
Avoid use in a location subject to sudden temperature changes that cause con-
Fastening ring
densation.
This product is suitable for indoor use only.
Avoid dust, dirt, and steam.
Do not use in a location where there are corrosive or other harmful gases.
Ensure that the product does not come into contact with organic solvents such as
thinner.
Ensure that the product does not come into contact with strong acid or alkaline.
1
Ensure that the product does not come into contact with oil or grease.
1
sive gases.
Do not use the product in a location subject to intense vibration or shock.
The connectors are not structurally dustproof, waterproof, or corrosion-resistant.
Do not use the product underwater or in the rain. Take care of the usage environment.
This product is a precision device. Do not drop or otherwise subject to shock. Risk
of product damage.
2
When the product becomes unusable or unneeded, dispose of the product appro-
2
priately as industrial waste.
2
2
8
SPECIFICATIONS
Type
Model No.
Combined controller
Position detection method
Installation distance
Measurement width
Light source
Repeatability (Note 2)
Linearity (Note 3)
Minimum sensing object (Note 4)
Temperature characteristic
-
(Note 5)
Protection
Pollution degree
Ambient temperature
Ambient humidity
Ambient illuminance
Operating altitude
Grounding method
Material
Cable
Weight
• Ambient temperature : +20°C
• Average controller count setting: 16 times
• Measurement object : Opaque knife edge
• Measurement object position condition: Light half-blocked at the center of 100-mm installation distance
Notes: 1) This product complies with the FDA regulations (FDA 21 CFR 1040.10 and 1040.11) in accordance with FDA
Laser Notice No. 56, except for complying with IEC 60825-1 Ed. 3.
2) P-P value of deviations in digital measurements obtained with light half-blocked at the center of installation
distance
3) Indicates the error from the ideal line formed by digital measurements.
4) When light is light blocked at the center of 500-mm installation distance.
5) When light is light half-blocked at the center of 100-mm installation distance.
6) When the sampling cycle of the controller is set to "Standard Sampling". For detailed information, refer to the
"HG-T Series User's Manual" (our Website: https://industry.panasonic.com/ ).
7) Do not use or store in an environment that has been pressurized to an air pressure higher than the atmo-
spheric pressure at 0m.
Panasonic Industry Co., Ltd.
1006, Oaza Kadoma, Kadoma-shi, Osaka 571-8506, Japan
https://industry.panasonic.com/
Please visit our website for inquiries and about our sales network.
Panasonic Industry Co., Ltd. 2024
April, 2024
HG-TC
-
-
Measurement width 10 mm / slim type
HG-T1110
CMOS method
0mm to 500mm
10mm
Red semiconductor laser Class 1 [IEC / EN / JIS / GB /KS Standards / FDA Regulations (Note 1)]
Maximum output: 0.3mW, Peak emission wavelength: 655nm
1μm (at an installation distance of 20mm)
2.5μm (at an installation distance of 100mm)
5μm (at an installation distance of 500mm)
±0.12% F.S. (at an installation distance of 20mm)
±0.28% F.S. (at an installation distance of 100mm)
ø0.5mm (at an installation distance of 500mm)
±0.03% F.S./°C
IP67 (IEC) (excluding the connector)
2
-10°C to +45°C (No condensation or icing), Storage: -20°C to +60°C
35% to 85% RH, Storage: 35% to 85% RH
Incandescent light: 5,000lx at the light-receiving face (Note 6)
2,000m or less (Note 7)
Capacitor grounding
Case: Aluminum die-casting, Light emitting / receiving faces: Glass
2
Attached 0.2m shielded cable with 0.08mm
4-conductor type, round connector
Emitter : 30 g approx., Receiver : 25 g approx.
PRINTED IN JAPAN

Advertisement

loading
Need help?

Need help?

Do you have a question about the HG-T1110 and is the answer not in the manual?

Questions and answers

Summary of Contents for Panasonic HG-T1110

  • Page 1 Thank you very much for purchasing Panasonic products. Read this Instruction Man- Sensor head connection cable This product complies with the FDA regulations (FDA 21 CFR 1040.10 and CN-HT-C ual carefully and thoroughly for the correct and optimum use of this product.
  • Page 2 0.08mm 0.2m HG-TC • 0.5N m • • • 100mm ENTER Laser Notice No.56 IEC 60825-1 Ed. 3 FDA 21 CFR 1040. EEPROM 1040.11 500mm 100mm HG-TC “ ” HG-T https://industry.panasonic.com/ 24mm 14mm CN-HT-C CN-HT-C Panasonic Industry Co., Ltd. 2024...