This Handbook describes in detail how to use the Control and Evaluation Software. System overview Instrumentation in the Biacore T100 system is described in full in the Biacore T100 Instrument Handbook. Important features relevant to software operation include: •...
) environments is provided in an optional package that adds appropriate functionality to the Biacore T100 software. Functions for GxP support are described in a separate Biacore T100 GxP Handbook. Descriptions of software in the current Handbook apply to installations both with and without the GxP package unless otherwise stated.
Page 11
The response may be absolute (above a fixed zero level determined by the detector) or relative to the response at another specified report point. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 12
1 Introduction 1.4 Biacore terminology Figure 1-2. Schematic illustration of a sensorgram. The bars below the sensorgram curve indicate the solutions that pass over the sensor surface. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Control Software – general features 2 Control Software – general features Operational modes Biacore T100 Control Software offers three modes of operation: • Manual run provides interactive control of the instrument operation, executing commands singly as they are issued. This mode is most useful for ad hoc experiments involving one or a few injections, such as testing the response obtained from injection of a single sample.
The content of the status bar varies between different situations: for wizard- and method-based runs, the elapsed run time and the estimated total run time are included. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Export exports the current results to a file in Microsoft Excel or XML format, or exports the contents of the report point table to a tab-separated file. See Appendix A for details of the export format. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 18
When you close the software with Exit while the instrument is still switched on, you may choose to shut down the instrument for a shorter or longer period if required. See the Biacore T100 Instrument Handbook or the on-line help for more details.
[Blank] for flow cells that have been prepared as a blank reference surface by activation and deactivation. The text [Incomplete results] indicates that the immobilization run was interrupted (by for instance user intervention or power failure) before it could be completed. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 20
When the reference line is displayed, choosing Baseline sets a baseline at the current reference line position, and the coordinates window shows the response relative to that baseline. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Standby puts the instrument in standby mode, which maintains a low buffer or water flow through the flow system for up to 4 days. Leaving the instrument in standby mode when not in use is generally recommended. The instrument is Biacore T100 Software Handbook BR-1006-48 Edition AE...
Insert Chip and Eject Chip are used for docking and undocking the sensor chip respectively. More details are given in Chapter 3 of the Biacore T100 Instrument Handbook. Set Temperature sets the sample compartment and analysis temperature.
Results are saved in files with the file name extension .blr. Result files from wizard- or method-based runs contain a copy of the wizard template or method as well as the results of the run. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 24
2 Control Software – general features 2.4 File storage Biacore T100 Software Handbook BR-1006-48 Edition AE...
You insert the samples as part of the starting procedure for the run. You can also add samples during the run. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Click Eject Rack to eject the rack tray so that you can load your samples. Click Start to start the run. You will be asked to specify a result file name before the run actually starts. Biacore T100 Software Handbook BR-1006-48 Edition AE...
You can also use the right-click menu to copy selected command or commands and paste them elsewhere in the queue. The Copy function works with both completed and pending commands. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 28
If you want to place the command later in the queue, use the right-click menu in the queue panel to insert the command at the appropriate place. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Stop Run command in the appropriate position. Choose Tools:Eject Rack to eject the rack tray and remove your samples and reagents. Choose Tools:Eject Chip to undock the chip if desired. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 30
3 Manual run 3.4 Ending a manual run Biacore T100 Software Handbook BR-1006-48 Edition AE...
Application wizards Application wizards guide you through the procedure of setting up common applications, with recommendations and settings based on GE Healthcare’s expertise in the field of SPR-based interaction studies. Wizards are an ideal starting point for inexperienced or infrequent users, since they offer a structured sequence of settings that covers all essential aspects of the assay in question.
4.2.1 Injection sequence This dialog determines the sequence of injections in the wizard analysis cycle. Some injections are not supported in certain wizards (e.g. the kinetics wizard does not support enhancement injections). Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 33
This is the sample to be analyzed. The solution used for the sample injection is normally different in different cycles, and is specified in the sample table at a later stage in the wizard. The sample injection is required in all wizards. Biacore T100 Software Handbook BR-1006-48 Edition AE...
For these chips (Sensor Chip C1, CM-series and SA) the option to run conditioning cycles is unchecked by default. You may want to use conditioning cycles in cases where Biacore T100 Software Handbook BR-1006-48 Edition AE...
The Injection parameters dialog specifies details of the injections selected in the Injection sequence. Details of this dialog box may vary according to the injections selected and the particular wizard. Some features may be generalized: Biacore T100 Software Handbook BR-1006-48 Edition AE...
Sample and Control Samples steps respectively. The details of these steps differs according to the wizard purpose, but the following general features may be noted. Further details are given in the respective wizard descriptions later in this chapter. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Sample details can be imported from an external file if this option is enabled under Tools:Preferences. See Appendix A for details of import formats and procedures. 4.2.5 System preparations This dialog box specifies how the system will be prepared before the first cycle. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 38
Note: Both analysis temperature and sample compartment temperature can be set in advance with the Set Temperature option from the main Tools menu (Section 2.3.6), to allow the temperature to equilibrate before setting up the assay wizard. Biacore T100 Software Handbook BR-1006-48 Edition AE...
This dialog box shows where samples and reagents are placed in the microplate and/or rack. Positions are color-coded by region according to sample and reagent categories: you can change the color-coding in the Automatic positioning dialog, accessed through the Menu button. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 40
(Ctrl-V) in another. Positions can also be reorganized using the Automatic positioning dialog (see below). Menu functions Use the Menu button to access additional functions for rack positioning. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 41
This option controls whether the samples and reagents will be placed in the reagent rack or the sample microplate. If Auto is chosen, placement is decided on the basis of number and volume of solutions in the region. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 42
Appendix A for details of the exported file format. Print Rack Positions Prints a copy of the rack positions diagram and table. Print Wizard Template/Print Method Prints a copy of the currently open wizard template or method. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Buffer requirements for standby after the run (approximately 65 ml/day) are not included. The Menu button provides options for saving and printing the wizard template (Section 4.2.6). Biacore T100 Software Handbook BR-1006-48 Edition AE...
Enter the buffers and pH values to be used for scouting. The default list covers sodium acetate buffers in the pH range 4 to 5.5, available as ready-to-use solutions from GE Healthcare. Buffers will be tested in the order listed. Note: The buffers listed here are buffers in which the ligand should be prepared.
Page 45
The same overlay plot is created when a saved run from Immobilization pH Scouting is opened in the Evaluation Software. Note: Opening a saved run in the Control Software does not generate the overlay plot automatically. Biacore T100 Software Handbook BR-1006-48 Edition AE...
See the Biacore Sensor Surface Handbook for more information about ligand immobilization. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 47
Choose the immobilization method. Predefined methods are provided for standard immobilization chemistries. Customized methods can be defined by clicking on the Custom Methods button (see below). Predefined methods are marked with a T100 icon ( ) in the selection lists. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 48
150 µl is reached. This option can be used to conserve valuable ligand without losing the benefits of aiming for a target immobilization level, and can be useful for sensor chips where preconcentration cannot be performed, e.g. Sensor Chip Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 49
The ligand injection is created automatically and cannot be deleted: solution and contact time for the ligand injection are specified in the main wizard dialog. A method may only contain one ligand injection. Other injections have the following functions: Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 50
Step 2. System preparations Check the System preparations options as required (see Section 4.2.5). Step 3. Rack positions Accept or change the rack positions for the various solutions required (see Section 4.2.6). Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 51
EDC/NHS. For low ligand levels, this value usually gives a better indication of the amount of ligand immobilized. Result files from immobilization can also be opened in the Evaluation Software if you want to prepare other sensorgram displays or plots (see Chapter 7). Biacore T100 Software Handbook BR-1006-48 Edition AE...
Regeneration scouting always includes one startup cycle with the same injection sequence as the scouting cycles but with injection of buffer for all injections. Note that this differs from the startup cycle construction in the other application wizards. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 53
Check High viscosity solution if any of the regeneration solutions tested has a relative viscosity higher than about 3 (corresponding to about 35% glycerol or 40% ethylene glycol at 20°C). This will modify the injection procedure for better handling of viscous solutions. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 54
Note: Report points are set before the sample injection for baseline and shortly after the sample injection for sample response. Thus the points for the first cycle indicate the starting values, while those for subsequent cycles each indicate the effect of the previous cycle. Biacore T100 Software Handbook BR-1006-48 Edition AE...
(in which case a single flow cell will be used for each buffer, with flow cells 1, 2, 3 and 4 for buffers A, B, C and D respectively). Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 56
Ligands for capture, samples and enhancement reagents should be prepared in each of the buffers tested. Separate rack positions will be created for samples in each buffer (for example, running buffer scouting with 4 buffers and 5 samples will require 20 sample positions). Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 57
When the wizard run is completed, the results are opened automatically in the Evaluation Software. In addition to the general predefined plots (Section 6.4), plots of binding and stability against sample are created to visualize the behavior in the different buffers. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Step 3. Injection parameters Specify the injection parameters for each injection in the cycle (see Section 4.2.3). Step 4. System preparations Check the System preparations options as required (see Section 4.2.5). Biacore T100 Software Handbook BR-1006-48 Edition AE...
Specify the conditioning and start-up cycles (see Section 4.2.2). Step 3. Injection parameters Specify the injection parameters for each injection in the cycle (see Section 4.2.3). One sample injection panel will be created for each sample Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 60
Click Import to import the sample data from an external file. Import of sample information must be enabled in Tools:Preferences to use this function. See Appendix A for details of import functions and file formats. Click Control Samples to enter control samples for the run. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 61
Binding analysis results When the wizard run is completed, the results are opened automatically in the Evaluation Software. Predefined plots (Section 6.4) are created for each sample. Biacore T100 Software Handbook BR-1006-48 Edition AE...
The value specified for Fraction refers to the proportion of the fixed component in the final mixture: for example, a value of 30% will mix 7 parts of sample with 3 parts of the specified solution. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 63
To run replicate concentrations, enter the same concentration on multiple rows. Calibration points will be run in the order entered. You can choose a different concentration unit if required from the pull- down list in the table header. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 64
Control samples are specified in terms of sample ID and expected concentration. The expected concentrations should lie within the range covered by the calibration curve. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 65
Save the wizard template if required and start the run. Evaluation of concentration assays Chapter 8 describes how to evaluate concentration assays. Biacore T100 Software Handbook BR-1006-48 Edition AE...
30-second injection at 40 µl/minute, regardless of other settings in the wizard. Report points are set automatically for the carry-over injection to allow evaluation of potential carry-over problems. Step 2. Setup Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 67
Section 4.2.3). The sample injection for kinetics measurement has an additional setting for dissociation time. This is the time for which dissociation will be monitored after the end of the injection without disturbances from flow system washing procedures. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 68
A, B, A, A, B, A, B, B… will be run in the order A, A, A, A, B, B, B, B… The Run order setting applies within each concentration series. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 69
Save the wizard template if required and start the run. Evaluation of kinetics/affinity assays Chapter 9 describes how to evaluate kinetics/affinity assays. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Specify the conditioning and start-up cycles (see Section 4.2.2). Start-up cycles are recommended for thermodynamics experiments to ensure that the response drift that may occur with a new chip or when the temperature is Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 71
Enter the details of the samples. This dialog box is equivalent to the concentration series dialog for kinetics determination (see Section 4.5.3, step 4). Separate rack positions will be created for each sample at each temperature. Biacore T100 Software Handbook BR-1006-48 Edition AE...
The control experiment wizards have the same basic structure as the kinetic analysis wizard, but some of the analysis settings are fixed. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Note: The Injection Parameters step allows you to enter a stabilization time after each sample injection, but all other settings in this dialog are fixed in the wizard. Biacore T100 Software Handbook BR-1006-48 Edition AE...
This example shows a clear dependence of binding rate on flow rate, indicating mass transfer limitations in the observed binding. It may however still be possible to extract kinetic information from the sensorgrams. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 75
This example shows a clear dependence of dissociation rate on contact time, indicating linked reactions in the interaction model. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Methods 5 Methods Methods in Biacore T100 offer flexibility in instrument control, providing support for applications that cannot conveniently be handled with wizards. Methods are constructed with the Method Builder tool as described in this chapter. Templates from application wizards can be opened in Method Builder (Section 5.1) to provide a starting point for development of customized applications.
+- marking or use the Expand All button to expand all commands in the panel. This screen is for information only: settings cannot be changed here. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Check the Vary with analysis temperature box to set the sample compartment temperature automatically to the same value as the analysis temperature. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Recurrence. To create a new assay step, click New Step. The step will be created with default settings at the end of the current method. Move the step in the method with the Biacore T100 Software Handbook BR-1006-48 Edition AE...
Assay steps are assigned a purpose, used to identify cycles in the evaluation software. The choice of purpose can help to document the method structure, and also determines the way the data is treated in evaluation. Choose the purpose from the list. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 83
Sample step. Undefined Used for assay steps that do not fit the predefined purposes. Assay steps with Undefined purpose will not be included in application-specific evaluation. Biacore T100 Software Handbook BR-1006-48 Edition AE...
If the higher level step is run in replicate, a recurring step is distributed among the total number of cycles including replicates. This is illustrated by the calculations below for a recurring step set to Every 5 cycles: Biacore T100 Software Handbook BR-1006-48 Edition AE...
Select the running buffer to be used for the assay step. The default buffer is A (corresponding to buffer bottle and tubing A on the instrument). Cycle types Cycle types define the detailed sequence of operations to be performed in each assay step. Biacore T100 Software Handbook BR-1006-48 Edition AE...
) buttons to change the position of the selected command in the cycle, and the Remove button to remove the selected command from the cycle definition. Commands are executed from top to bottom in the cycle definition. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 87
The illustration below shows a cycle which will perform an additional regeneration if the relative response after the first regeneration exceeds a specified value: Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 88
If…then construction. You can use the Move up and Move down buttons to rearrange the order of commands within a branch, but you cannot move commands outside the branch in which they are placed. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 89
Biacore analysis with mass spectrometry (MS). Most of the parameters for the command can be set as variables. This command can only be used for injection over all four flow cells: normally, the same ligand should be immobilized in each cell. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 90
The same target position is used for recovered analyte from all repetitions. Note that only one aliquot of deposition solution is used, regardless of the number of repetitions. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 91
Notes: Methods that include the InjectAndRecover command require a sample and reagent rack and cannot be used with microplates (see the Biacore T100 Instrument Handbook for rack details). The contact time for sample, flow rate and number of repetitions determine the total injected volume for both sample and recovery solution.
(for example, kinetic evaluation requires a variable called Conc which holds the analyte concentration). These are selected from a predefined list. Other evaluation variables may be freely defined by the user, to hold Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 93
Analyte molecular weight: required for molecular weight adjustment of report points, and for kinetic evaluation when the concentration is entered in weight-based units. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 94
(the concentration of B is kept constant) ConcA-variable Concentration of interactant A in the sample mixture (the concentration of A is varied) *See Section 11.1.1 for details of how this assay is set up. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Columns in the table correspond to variables for the cycle type used in the assay step, and are grouped according to commands in the cycle type definition. Use the right Biacore T100 Software Handbook BR-1006-48 Edition AE...
5.9.1 Detection Set the flow path for the method in the Detection dialog. You can only choose a flow path that is consistent with the Detection setting for the method (see Section 5.4). Biacore T100 Software Handbook BR-1006-48 Edition AE...
Method variable values must be entered in all tables before you can continue to the next step. Evaluation variables may be left blank if desired at this step and values entered in the Evaluation Software. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Check through the cycle list to confirm that the variable tables are correctly filled in. Click Overview to display the method overview (Section 5.3). 5.9.4 System preparations Choose which preparation steps should be executed before the method starts. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Conc. At least two different concentrations are required for linear calibration curves and at least four for 4-parameter fitting. • Samples in the Sample step will normally not have specified concentrations. If concentrations are specified, they will be ignored. Biacore T100 Software Handbook BR-1006-48 Edition AE...
At least one assay step is required with purpose Calibration and one with purpose Sample. Both assay steps must be connected to a cycle type that includes one Sample command. The two assay steps will normally be connected to the same cycle type. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Other requirements Application of solvent correction (see Section 6.7) requires an assay step with purpose Solvent Correction, connected to a cycle type that includes at least four Solvent Correction commands for different solvent concentrations. Biacore T100 Software Handbook BR-1006-48 Edition AE...
This chapter describes the organization of the evaluation software. The various evaluation functions are described in detail in the following chapters. User interface 6.1.1 Organization The Biacore T100 Evaluation Software screen is divided into three main regions: Menu and toolbar Evaluation explorer Work area Status bar •...
To open multiple result files in the same session, use the File:Append file option. You can only append result files, not evaluation sessions. Appending a file to a session will delete all the user-defined evaluation items in the session. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Displays the sensorgram relating to a point in the plot. This option is only available when you right-click on a point: the sensorgram is displayed in a separate window that must be closed before you can continue with the evaluation. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 108
Exports the curves in the current display to a tab-separated text file, for import to third-party software. Complete data is exported regardless of the scale setting of the screen display. Unzoom Restores the previous zoom setting. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Binding to reference Relative response for report point stability against cycle number for the reference flow cell. Capture Relative response for report point capture_level against cycle number for the capture injection. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Control Software (Section 2.3.3), are not listed in the Custom Report Points dialog and cannot be edited or deleted in the Evaluation Software. Choose Tools:Custom Report Points to add and edit custom report points. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Click Add to add a new report point. Enter the id for the report point (maximum 30 characters) in the Id field. The report point id must be unique within an evaluation session. Biacore T100 Software Handbook BR-1006-48 Edition AE...
If you delete a custom report point that is used as baseline for other report points, the relative response can no longer be calculated for the latter points. You will be warned if this situation arises. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Click in the filter row (directly below the column header) and select a value to display only rows with that value for the chosen column. Click Reset All Filters to restore all filters to the [All] setting. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 114
Note: All user-defined evaluation items are deleted from the session when changes are made to the keyword table, since the changes may affect existing evaluation items. To avoid losing work, save your evaluation session before editing or deleting custom report points. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Such variations are unavoidable in the preparation of diverse samples such as drug candidates for screening applications. The solvent correction procedure corrects for the variations arising from the excluded volume effect in these cases. Biacore T100 Software Handbook BR-1006-48 Edition AE...
–150 RU in the reference sensorgram corresponds to a solvent error of +5 RU in the reference-subtracted sensorgram. 3. The reference-subtracted sensorgram is corrected by subtracting the solvent error. This procedure is applied to every point during sample injection. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Statistical fitting parameters (chi- squared) are shown for each correction curve in the cycles list. Right-click on a point or curve in the right panel and choose Show sensorgrams if you want to Biacore T100 Software Handbook BR-1006-48 Edition AE...
Click Skip to create the method with no kinetics/affinity item included. Kinetic/affinity items that use data sets with multiple R cannot be saved in evaluation methods. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Click Apply to apply the method. Items in the method will be created where possible. Items that cannot be created will be reported in a dialog box. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Hold the cursor over a sensorgram to display a tool tip identifying that particular curve. The sensorgram coloring can be changed if desired with the Tools:Color by option. The following sections describe display functions specific to sensorgram items. General display functions are described in Section 6.3. Biacore T100 Software Handbook BR-1006-48 Edition AE...
This function can be useful for removing injections with high bulk contributions (such as regeneration injections) or other visual disturbances from the display. Choose Undo Cut from the right-click menu to restore the removed data. Biacore T100 Software Handbook BR-1006-48 Edition AE...
(e.g. adjusting analyte response for varying capture levels, by adjusting the baseline to 0 and the capture level to 100). Biacore T100 Software Handbook BR-1006-48 Edition AE...
Report points and Event markers options respectively under the Tools button. Report points are displayed on the curve and event markers on the x-axis. Figure 7-1. Sensorgram display with labels and markers shown for both report points and events. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Tool tips identify the data points (place the cursor on a point for a couple of seconds to display the tool tip). Biacore T100 Software Handbook BR-1006-48 Edition AE...
Selection operates in the same way as in the sensorgram window (Section 7.1.1). Other general display functions are described in Section 6.3. Biacore T100 Software Handbook BR-1006-48 Edition AE...
You can also change the order in which columns will be displayed using the Move up and Move down buttons (the top of the column list represents the left-hand column in the table). Biacore T100 Software Handbook BR-1006-48 Edition AE...
The numerical fitting results are displayed in an extra panel below the plot. Choose Curve Fitting:Curve parameters under the Tools button to toggle display of this extra panel. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 129
The closeness of fit is reported for linear fitting as the correlation coefficient R and for 4-parameter fitting as the chi-squared value (see Section 9.7.1). Curve fitting cannot be applied to plots that have been sorted. Biacore T100 Software Handbook BR-1006-48 Edition AE...
If no sample is chosen as a negative control, 0 RU is used as the negative control response. • The fitted line(s) are transformed to straight horizontal lines with values 100 for the positive control and 0 for the negative control. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 131
(Figure 7-2). Figure 7-2. Polynomial function fitted to 3 control points (highlighted) with the resultant adjustment for controls. Biacore T100 Software Handbook BR-1006-48 Edition AE...
This selection applies to all report points shown in the chart. Points for which the selection is invalid are not shown (for example, the report point baseline will normally not be shown on a bar chart of relative response). Biacore T100 Software Handbook BR-1006-48 Edition AE...
The report point table is created automatically as an evaluation item. Each evaluation session can only have one report point table item: the table is updated automatically if you add custom report points or apply solvent correction. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 135
Report point time in seconds from the start of the cycle. Window (s) Report point time window in seconds. AbsResp (RU) Absolute response in RU, calculated as the mean value over the time window. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 136
The report point table can be sorted by any column in ascending or descending order for any column. Click in the header for a column to sort the table by that column. Each click in the same header toggles the sort order. Header row Filter row Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 137
Choose Copy Table to copy the whole table. The contents will be copied in tab- separated text format to the Windows clipboard, and can be pasted from there into other programs. All selected cells will be copied, including header cells and filter settings. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 138
7 Data presentation tools 7.4 Report point table Biacore T100 Software Handbook BR-1006-48 Edition AE...
Concentration analysis 8 Concentration analysis Biacore T100 software supports two approaches to concentration analysis: • Using a calibration curve determined by measurement of known samples (referred to as calibrated measurements). • Calibration-free measurements that rely on the relationship between the analyte concentration and the rate of mass transfer of analyte to the sensor surface.
Settings on the Calibration tab define the report point and fitting function that are used to create the calibration curve. The settings apply to all calibration curves in the evaluation. Choose the settings appropriate for your analysis: Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 141
Changing sample names in calibration cycles only changes the title in the calibration curve panel and has no effect on the curve itself. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Right-click on a sample row in the table or a sample point in the plot and choose Exclude Cycle to exclude that sample from the sample evaluation. Excluded cycles are shown as open symbols and are marked in the table with red strikethrough text. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Right-click on a sample row in the table or a sample point on the curve and choose Exclude Cycle to exclude that sample from the average sample calculation. Excluded cycles are shown as open symbols and are marked in the table with red strikethrough text. Biacore T100 Software Handbook BR-1006-48 Edition AE...
It is for example in principle possible to append a kinetic analysis result file to a concentration analysis, and then apply concentration analysis evaluation: Biacore T100 Software Handbook BR-1006-48 Edition AE...
To evaluate calibration-free concentration analysis, choose Concentration Analysis:Calibration-free from the evaluation toolbar, or choose Add Concentration Analysis:Calibration-free from the Evaluation menu. Biacore T100 Software Handbook BR-1006-48 Edition AE...
The option Use reference subtracted data is checked by default if a reference surface is included in the run setup. You should generally perform the evaluation using reference-subtracted data, but you can uncheck this option to use unsubtracted data if you wish. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 147
(so the binding rates should in theory be equal and the QC ratio should be 0). Samples where the QC ratio is negative do not show sufficient mass transport limitation for calibration-free concentration measurements. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 148
Click Included blanks to display an overlay plot of blank cycles in the run. Remove the check-mark from the Include column to exclude a blank cycle from the evaluation (for example if the sensorgram shows unreasonable response levels or unacceptable disturbances). Biacore T100 Software Handbook BR-1006-48 Edition AE...
When you are satisfied with the data selection, click Next to start the evaluation. Results appear in the table as each sample series is completed. The evaluation progress is shown in the sensorgram panel. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Samples where the QC ratio of the i is less than about 0.13 should be discarded (see Section 8.3.1). Results for samples where the initial binding rate at 5 µl/min is outside the range 0.3–15 RU/s should be treated with caution. Biacore T100 Software Handbook BR-1006-48 Edition AE...
20°C. The value at 20°C is provided as an evaluation variable when the assay is run (Section B.2). Parameters for calculating the mass transport coefficient from the diffusion coefficient are fixed in the software. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 153
Kinetics and affinity analysis 9 Kinetics and affinity analysis Biacore T100 offers three main functions for analysis of interaction kinetics and affinity: • Kinetics and affinity measurements on the sensor surface, which determine the interaction characteristics between ligand and analyte.
• for multi-cycle kinetics, duplicate determinations for at least one non-zero concentration. These conditions are recommended but not mandatory in the Kinetics/Affinity wizard. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Ligand name. Multiple choices may be available if the ligand is captured and varies between cycles. Result files from versions of Biacore T100 Software earlier than 2.0 do not contain a keyword for Ligand name, and the name will be shown as N/A.
Page 156
The adjustment compensates for small systematic discrepancies in the interval between the start of the injection as recorded in the event log and the time that the Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 157
Click Next> when you have selected the curves to be evaluated. The second dialog shows the blank subtracted curve set and allows you to delete selected regions from all or selected curves, for example to eliminate spikes or other disturbances. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 158
The top panel shows the plot of R against C, based on average response values over the region marked on the sensorgrams. Click Settings to adjust the region used for calculation of R values. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 159
Clicking on Accept will stop the Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 160
The quality control criteria are discussed in Section 9.4.1. This tab is only shown for kinetic evaluations using the predefined 1:1 model. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 161
Report tab are defined in the model. Global parameters are listed on a single row at the top of the table, and local parameters are listed on one row for each curve. • Parameters shows the values for all parameters in the fitting equations. Biacore T100 Software Handbook BR-1006-48 Edition AE...
The same sample name and analysis temperature apply to all subsets. Click on a subset to manage the contents of the subset. You cannot mix subsets that use different samples or different temperatures. Biacore T100 Software Handbook BR-1006-48 Edition AE...
To perform evaluation in batch mode, start the kinetics/affinity evaluation (Section 9.2.1) and check the Batch mode option in the first dialog. Biacore T100 Software Handbook BR-1006-48 Edition AE...
1:1 model only) gives a brief overview of the reliability of the results. If you prefer, you can hide the quality control tab by setting the appropriate option in Tools:Preferences on the main menu. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 165
Fail: quality assessment unacceptable (Blue) Neutral or user assessment required The quality of the fitting is assessed in five areas: Magnitude of kinetic constants Biacore T100 can reasonably measure kinetic constants in the range 10 –10 for k and 10 – 0.5 s for k .
Page 166
Bottom right: Slow association and dissociation, giving insufficient curvature in both association and dissociation phases. Evaluation is possible but will not be very reliable. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 167
On the other hand, Fail status in any of the parameters is a reliable warning indicator. Base your assessment on the overall quality of the results and the fitting, taking all quality control parameters into account. Biacore T100 Software Handbook BR-1006-48 Edition AE...
For ease of comparison between parameters with widely different absolute values (e.g. k and k ), the standard error may be expressed as a T-value, which is obtained by dividing the value of the parameter by the standard error. A high Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 169
U-values above about 25 indicate that absolute values for two or more of the parameters (rate constants and R ) are correlated and cannot be determined. If the U-value is below about 15 the parameter values are not significantly correlated. Biacore T100 Software Handbook BR-1006-48 Edition AE...
This tool is only available for results obtained with the 1:1 fitting model. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 171
The tool display allows the simulated difference curves to be compared to the experimental residuals Biacore T100 Software Handbook BR-1006-48 Edition AE...
To help in this assessment, the calculated K value is indicated as a vertical line at the corresponding analyte concentration. The line is red and broken if the value is greater than half the highest concentration. Biacore T100 Software Handbook BR-1006-48 Edition AE...
The summarized kinetic and affinity results are presented as thumbnail plots of the kinetics and evaluation items and as a table of result data. Multiple fits in the same item in the Evaluation Software are presented as separate fits in the summary. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 174
Extended Shows more detailed identification of each thumbnail than the standard view. Thumbnail display settings Choose Display Settings from the View menu or the right-click menu for display settings for the thumbnails. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 175
(Copying a graph and a thumbnail is equivalent when small thumbnails are displayed, since there is no additional information in this mode.) Copying all items copies all the included items in the summary (regardless of Biacore T100 Software Handbook BR-1006-48 Edition AE...
Click on a point in the on-off rate map or select a row in the table below the map to display the thumbnail for the evaluation. Biacore T100 Software Handbook BR-1006-48 Edition AE...
The fitting process begins with initial values for the parameters in the equation set, and optimizes the parameter values according to an algorithm that minimizes the chi-squared value (Section 9.4.2) for the fitting. Biacore T100 Software Handbook BR-1006-48 Edition AE...
It is however justified to use a local R if there is reason to believe that the ligand activity may vary between cycles (e.g. in a capture assay, if the capture level varies between cycles). Biacore T100 Software Handbook BR-1006-48 Edition AE...
(m h, w, l are the flow cell dimensions (height, width, length in m) One form used in fitting models in Biacore T100 is referred to as the mass ⋅m⋅s transfer constant k (units RU⋅M...
Analyte binding capacity of the surface (RU) Rmax Conc Analyte concentration (M) Conc Flow rate-independent component of the mass transfer constant Flow Flow rate (µl/min) tc × f Mass transfer constant Bulk refractive index contribution in the sample Biacore T100 Software Handbook BR-1006-48 Edition AE...
Flow rate (µl/min) Provided as input Sample injection start time (s) Provided as input tOff Sample injection end time (s) Provided as input Bulk refractive index contribution in the sample Fitted Biacore T100 Software Handbook BR-1006-48 Edition AE...
The model cannot evaluate interactions where the proportions and relative sizes of the analytes are unknown. A1 + B = A1B A2 + B = A2B Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 183
(Rmax1/Rmax2) / (MW1/MW2). Flow rate (µl/min) Provided as input Sample injection start time (s) Provided as input tOff Sample injection end time (s) Provided as input Bulk refractive index contribution in the Fitted sample Biacore T100 Software Handbook BR-1006-48 Edition AE...
In cases where the heterogeneous ligand model is found to give the best fit to the observed sensorgrams, further experimental efforts to reduce the heterogeneity are recommended where possible. A + B1 = AB1 A + B2 = AB2 Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 185
Association rate constant for the first and second ligands (M Dissociation rate constant for the first and second ligands (s Equilibrium dissociation constants (M) kd1/ka1 kd2/ka2 Rmax1 Analyte binding capacity of the first ligand Rmax1 (RU) Biacore T100 Software Handbook BR-1006-48 Edition AE...
Reverse rate constant for the conformational Fitted change (s Rmax Analyte binding capacity of the surface (RU) Fitted Conc Analyte concentration (M) Provided as input Flow rate-independent component of the Fitted mass transfer constant Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 187
Analyte binding capacity of the surface (RU) Rmax Conc Analyte concentration (M) Conc Flow rate-independent component of the mass transfer constant Flow Flow rate (µl/min) tc × f1/3 Mass transfer constant Bulk refractive index contribution in the sample Biacore T100 Software Handbook BR-1006-48 Edition AE...
Note: Custom models defined in software versions earlier than 2.0 can be imported and viewed but cannot be edited.These models do not support evaluation of single-cycle kinetics. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Note: Numbers are used as part of the identifier, not in the conventional chemical sense of stoichiometry. Thus a complex named AB2 does not imply two molecules of B binding to one of A. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 190
Check the box and enter the molecular weight parameter for the analyte to which the binding capacity parameter refers. Binding capacity for the other analyte will be calculated using the molecular weight values. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 191
Enter parameter names for the forward and backward rate constants for each line in the reaction scheme. (The terms k-forward and k-backward apply to the reaction as entered in the scheme, reading from Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 192
In the Report panel, define the parameters you want to appear in the Report tab of the results. Report parameters are defined by a name that may be chosen freely and a value that is entered as a parameter or expression containing parameters. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 193
You can select the equations in the display and click Copy to copy the equations to the Windows clipboard. Use this function and paste the equations in to e.g. Wordpad to print a copy of the rate equations. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Select Equation model in the subsequent dialog. Parameters and report parameters are defined in the same way as for kinetic models. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Note: Beware of trying to define and use complex models for steady state affinity. Because of the relatively few points available for fitting to steady state affinity models (typically about five concentrations in duplicate), complex models tend to give unstable fitting behavior. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 196
9 Kinetics and affinity analysis 9.9 Creating and editing models Biacore T100 Software Handbook BR-1006-48 Edition AE...
Thermodynamic analysis 10 10 Thermodynamic analysis Biacore T100 supports automated measurement of kinetics or affinity at a series of temperatures using the Thermodynamics wizard (Section 4.5.4). In addition to displaying the variation of kinetic and affinity constants with temperature, the evaluation software extracts standard thermodynamic parameters from the data.
Note that the Eyring equation does not have a corresponding non-linear form that takes account of the heat capacity change for transition state formation. Non-linear fitting to obtain values for ΔC ° can only be applied to equilibrium thermodynamic analysis. Biacore T100 Software Handbook BR-1006-48 Edition AE...
In some cases it may be necessary to examine the kinetic or affinity evaluation items to determine which constants belong together. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 200
Click Next> to display the van’t Hoff and Eyring plots together with a table of thermodynamic constants for the equilibrium and transition state formation. In any of the plots, right-click on a point to exclude the point from the line fitting. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 201
This may be evident from direct comparison of the kinetic fits, but will not be immediately apparent in the thermodynamic analysis. Biacore T100 Software Handbook BR-1006-48 Edition AE...
B determined over the same sensor surface, in order to calculate the free B concentrations in the samples. 11.1.2 Evaluation principles The equilibrium constant for a 1:1 interaction is given by ⋅ free free ---------------------------- - – – ------------------------------------------------------- - Biacore T100 Software Handbook BR-1006-48 Edition AE...
Refer to Chapter 5 for details of how to construct methods in Method Builder. Determination of affinity in solution is not supported by a wizard. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Samples that lie outside the range of the calibration curve are not shown. Right click on calibration points to exclude the points from the curve. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 206
ConcB. This value should be the same as or close to the value entered for the variable ConcB in the method. The calculated K value is shown in the panel below the table. Biacore T100 Software Handbook BR-1006-48 Edition AE...
.xls) containing separate worksheets for the file properties and report point table. The audit trail is also exported to a separate worksheet if it is present when the GxP module is installed (see the separate Biacore T100 GxP Handbook). Export from the Evaluation software an Excel spreadsheet file (extension .xls) containing separate worksheets for the file properties and for tabulated data for all evaluation items where appropriate (i.e.
Import programs and data files for use with import are the responsibility of the user. For each import function, a check box allows the import program to be started automatically without user intervention. If the respective box is not checked, the Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 211
Figure A-1. Part of an XML import file from a method for heterogeneous analyte kinetics displayed in an XML-compatible editor. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 212
Position specification and must be placed in the microplate or reagent rack before the run can be started. Details of the required import file format can be investigated further by examining a file created with the Menu:Export Positions command. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Evaluation Software The Evaluation Software supports import of model definitions for kinetics and affinity evaluation. Model files for import should be obtained from GE Healthcare or created by exporting models from another installation. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 214
Appendix A Data import and export A.2 Importing data Biacore T100 Software Handbook BR-1006-48 Edition AE...
Predefined evaluation variables are included for the evaluation purpose Affinity in solution: The Variable Settings is different for the two assay steps, so that only the relevant variables are entered at run-time for each assay step. Biacore T100 Software Handbook BR-1006-48 Edition AE...
µl) may differ significantly at the lowest flow rate. Sample solution and flow rate are set as method variables. For evaluation variables, the evaluation purpose is set to Calibration-free conc and predefined variables MW, D(20°C), Blank and Dilution are required. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Method examples and recommendations Appendix B B.2.3 Variable settings Sample cycles All the variables for assay step Sample are set at run time, except for the method variable Blank which is set to No in the method. Biacore T100 Software Handbook BR-1006-48 Edition AE...
(MW), diffusion coefficient at 20°C (D(20°C) and dilution factor (Dilution). Only flow rates and dilution factor are required for assay step Blank: these values must be the same as the flow rates used for samples. Biacore T100 Software Handbook BR-1006-48 Edition AE...
General Settings. Detection settings Multi and Single are not recommended. The method involves 5 conditioning cycles with injection of recombinant GST (provided in the GST Capture Kit from GE Healthcare) over both flow cells followed by regeneration: The sample analysis cycle uses two capture injections, one to capture...
Page 220
Note: If you change the detection setting to Multi, you will need to include additional capture injections for the active surfaces in flow cells 3 and 4. Do not use detection setting Single without removing the first capture injection. Biacore T100 Software Handbook BR-1006-48 Edition AE...
The cycle for recovery of bound analyte contains just one single InjectAndRecover command. Additional commands are usually not needed, since the recovery component of the InjectAndRecover command serves as a regeneration step (see Section 5.6.1). Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 222
B.4 Inject and Recover All parameters in the InjectAndRecover command are fixed: check the appropriate boxes in the Method Variables list if you want to use variable parameters. There are no evaluation variables for this command. Biacore T100 Software Handbook BR-1006-48 Edition AE...
The recommended method for liposome capture on Sensor Chip L1 uses one conditioning cycle consisting of two injections of 40 mM octylglucoside, followed by three start-up cycles using a dummy sample before the actual samples are run. Biacore T100 Software Handbook BR-1006-48 Edition AE...
This assay step is connected to a cycle type that includes 8 Solvent Correction commands for injection of 8 different solvent concentrations (see Section 6.7). Note: In Biacore T100 Control Software version 2.0 and later, solvent correction is included in the Kinetics/Affinity and Thermodynamics wizards.
Page 225
If you require a regeneration step, it is advisable to include regeneration after both the sample and the carry-over injections. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Ni and one for the poly-His tagged ligand. The Biacore method provided is designed for kinetic or affinity determinations, but may be readily modified for other applications that use Sensor Chip NTA. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Single cycle kinetics with 5 sample concentrations per cycle. Note that with this command type, the evaluation purpose is fixed as Kinetics/Affinity and the number of predefined variables for sample concentration corresponds automatically to the number of sample concentrations per cycle. Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 228
Appendix B Method examples and recommendations B.9 Single-cycle kinetics Biacore T100 Software Handbook BR-1006-48 Edition AE...
Page 238
Björkgatan 30 All goods and services are sold subject to the terms and conditions of sale of the company within GE Healthcare which supplies them. A copy of these 751 84 Uppsala terms and conditions is available on request. Contact your local GE Healthcare representative for the most current information.