ZERO CALIBRATION CHECK AND ACTUAL CALIBRATION

SPAN CALIBRATION CHECK AND ACTUAL CALIBRATION

1. While still in the Calibration>M-P menu, input NO Span gas through the
2. Either check or calibrate as follows:
3. Press the Stop button and return to Home screen.
4. In the Dashboard, check and record the Slope(s) and the Offset(s). (See
083730200A DCN7962
Go to the Calibration>M-P menu.
Input Zero air through the Sample port and press the Start button.
Either check or calibrate as follows:
CHECK ONLY:
a. Wait for reading to stabilize.
b. Press Stop and check the
reading.
Sample port and press the Start button.
CHECK ONLY:
a. Wait to reach stability,
then press Stop.
b. Record the reading(s).
Table
4-4
in
Section
expected/acceptable values).
Teledyne API T200 NO/NO
ACTUAL CALIBRATION:
a. Press the Zero button.
b. Press Stop and check the reading.
ACTUAL CALIBRATION:
a. Press the Set Span Target button and
enter the NO and NOx concentrations.
b. Verify the concentration reading is the
same as the NO concentration being
supplied.
c. If correct, wait to reach stability, then
press the Span button.
d. In the Cal Result window, press OK.
4.4,
Calibration
Quality
/NO
Analyzer with NumaView™ Software