C
ALIBRATION
Calibration of the analyzer is performed exclusively in the software. During instrument
calibration, (see Section 4) the user enters expected values for zero and span via the front
panel touchscreen control and commands the instrument to make readings of calibrated
sample gases for both levels.
The readings taken are adjusted, linearized and compared to the expected values.
With this information, the software computes values for instrument slope and offset
and stores these values in memory for use in calculating the NO
concentrations of the sample gas.
The instrument slope and offset values recorded during the last calibration can be viewed
in the Dashboard.
220
Teledyne API T200 NO/NO
- S
O
LOPE AND
FFSET
/NO
Analyzer with NumaView™ Software
2
x
, NO and NO
x
2
083730200A DCN7962