11.2 Digital measuring error

11.2
Digital measuring error
Digital measuring errors for resistance thermometers
Input point
According to IEC
60751
Pt25
Pt50
Pt100 ... Pt200
Pt500
Pt1000
According to
JIS C1604-81
Pt25
Pt50
Pt100 ... Pt200
Pt500
Pt1000
Ni25 ... Ni1000
Digital measuring errors for resistance-type transmitters
Input point
Resistance
Resistance
78
Measuring range in °C (°F)
-200 ... +850 (-328 ... +1562)
-200 ... +850 (-328 ... +1562)
-200 ... +850 (-328 ... +1562)
-200 ... +850 (-328 ... +1562)
-200 ... +350 (-328 ... +662)
-200 ... +649 (-328 ... +1200)
-200 ... +649 (-328 ... +1200)
-200 ... +649 (-328 ... +1200)
-200 ... +649 (-328 ... +1200)
-200 ... +350 (-328 ... +662)
-60 ... +250 (-76 ... +482)
Measuring range
[Ω]
0 ... 390
0 ... 2200
Minimum measuring
Digital accuracy in °C
span in °C (°F)
(°F)
10 (18)
10 (18)
10 (18)
10 (18)
10 (18)
10 (18)
10 (18)
10 (18)
10 (18)
10 (18)
10 (18)
10 (18)
Minimum span in [Ω]
Digital accuracy [Ω]
5
25
SITRANS TH200/TH300
Operating Instructions, 06/2010, A5E00393069-03
0,3 (0.54)
0,15 (0.27)
0,1 (0.18)
0,15 (0.27)
0,15 (0.27)
0,3 (0.54)
0,15 (0.27)
0,1 (0.18)
0,15 (0.27)
0,15 (0.27)
0,1 (0.18)
0,05
0,25